    Rn/cycles: qPCR decided ATAC-seq library amplification

    Hello, I have tried to set up a experiment following the ATAC-seq protocol. when I ran the qPCR side reaction to decide the cycle for library amplification, I met the problem.

    The protocol uses Y-axis Rn 5000 RF as the threshold to decide the # of cycle since it is corresponded to ¼ of maximum fluorescent intensity.
    In my experiment, I have tried cells # from 50,000 to 1,000,000 (the concentration of start PCR template DNA is 50 ng to 300 ng) but the maximum RF is around 200 RF(the amplification curve is fine). Should I only set 40 or 50 RF as threshold? (The correspond cycle is around 5.)
    Has any one have similar issue? Could anyone help me here?
  • obischof
    Member
    • Jan 2011
    • 11

    #2
    Originally posted by zhaolin View Post
    Hello, I have tried to set up a experiment following the ATAC-seq protocol. when I ran the qPCR side reaction to decide the cycle for library amplification, I met the problem.

    The protocol uses Y-axis Rn 5000 RF as the threshold to decide the # of cycle since it is corresponded to ¼ of maximum fluorescent intensity.
    In my experiment, I have tried cells # from 50,000 to 1,000,000 (the concentration of start PCR template DNA is 50 ng to 300 ng) but the maximum RF is around 200 RF(the amplification curve is fine). Should I only set 40 or 50 RF as threshold? (The correspond cycle is around 5.)
    Has any one have similar issue? Could anyone help me here?
    Very much depends on your PCR machine but we have similar data. On another note, if you put your amplified material on a gel or bioanalyzer do you observe a nuclesome-like pattern or rather a smear? We have trouble getting this defined nucleosomal pattern as shown in the protocol.

    Comment

    • zhaolin
      Junior Member
      • Nov 2014
      • 4

      #3
      I haven't start doing the gel checking. But I would change the transposase digestion time if I got a smear.

      Comment

      • obischof
        Member
        • Jan 2011
        • 11

        #4
        We did; always smear no matter how long or short you incubate.

        Comment

        • zhaolin
          Junior Member
          • Nov 2014
          • 4

          #5
          Originally posted by obischof View Post
          We did; always smear no matter how long or short you incubate.
          What is the size of your library smear? I have tried to lower the transposase amount in the reaction and found the size of PCR product changes. But it still a piece of smear. How much of PCR product did you check on gel? Is it too low amount in the protocol?

          Comment

          • Wonghe
            Junior Member
            • Sep 2014
            • 9

            #6
            I also got a smear for my PCR product between 150-2500bp. Has anybody sequenced these DNA does it enriched the open chomatin?

            Comment

            • zhengxiang he
              Junior Member
              • Dec 2014
              • 1

              #7
              Originally posted by zhaolin View Post
              Hello, I have tried to set up a experiment following the ATAC-seq protocol. when I ran the qPCR side reaction to decide the cycle for library amplification, I met the problem.

              The protocol uses Y-axis Rn 5000 RF as the threshold to decide the # of cycle since it is corresponded to ¼ of maximum fluorescent intensity.
              In my experiment, I have tried cells # from 50,000 to 1,000,000 (the concentration of start PCR template DNA is 50 ng to 300 ng) but the maximum RF is around 200 RF(the amplification curve is fine). Should I only set 40 or 50 RF as threshold? (The correspond cycle is around 5.)
              Has any one have similar issue? Could anyone help me here?
              How about your experiment now?
              We try several times. We start with 74k cells and can see clear cell pellet. But after we run the qPCR, we got very few signal on the qPCR and the curve is not good. Can you give us some suggestion?

              Comment

              • pritb
                Junior Member
                • Mar 2019
                • 2

                #8
                Hello All

                I am new to ATAC seq experiments. I am working on human fibroblast with 50k cells (counted through countess). Initially, I started following J Buenrostro (2015) protocol but seems not working. Now I shifted myself to Omni-ATAC protocol.

                My Initial OD (after Tn5 treatment) using Nanodrop gave 46 ng/ul. I proceeded with Library prep by giving 5 cycles. Realtime PCR for library quantification gave me 20 additional cycles for my Libraries. After PCR cleanup, I got 96 ng/ul as library concentration. On gel check, I cant seen anything, except for few bands that is less than 100 bps. What could be reason?

                I have attached my gel picture here.
                Lane-1: 2 ul of sample after Tn5 reaction post cleanup
                Lane-2: 15 ul of sample after 35 cycles (Although I selected 20 cycles based on 1/4 fluorescence intensity)
                Lane-3: 10 ul of sample of 20 cycles (additional PCR cycles). This is purified. You can seen 2 bands less than 100 bps.
                Attached Files

                Comment

                • Rosmano
                  Member
                  • May 2012
                  • 28

                  #9
                  Those two bands with less than 100bps are adapter dimers.

                  I am having problems in the step of pre-amplification.
                  After pre-amplification I get a DNA quantity of 7.6ng/uL from 50,000 total thymocytes. However, when I try to run a qPCR I get no amplification in the 20 cycles. I ordered the TDE1 individually and directly from Illumina now that they discontinued the Nextera DNA kit, so I wasn't expecting to have problems with it.
                  The primers I am using are the ones from the Buenruostro paper. I'm at a loss for why I am not getting amplification in the qPCR to determine how many extra cycles I need. It seems my problem is with the primers.

                  What's the usual DNA concentration one has for 50,000 cells after TN5 treatment and after 5-cycle pre-amplification?

                  Comment

                  • Rosmano
                    Member
                    • May 2012
                    • 28

                    #10
                    I am doing the qPCR quantification with Power SYBR Green PCR Master Mix from Thermo Fisher instead of the NEB Next HF Master Mix.

                    Does anyone know if this might impact the qPCR quantification?

                    Comment

                    • salix
                      Junior Member
                      • Mar 2022
                      • 1

                      #11
                      Originally posted by Rosmano View Post
                      I am doing the qPCR quantification with Power SYBR Green PCR Master Mix from Thermo Fisher instead of the NEB Next HF Master Mix.

                      Does anyone know if this might impact the qPCR quantification?
                      Were you able to figure out what was going wrong? I am experiencing the same thing.

                      Comment

                      • Rosmano
                        Member
                        • May 2012
                        • 28

                        #12
                        Yes, I needed to use the NEB Next HF Master Mix and then the quantification worked accurately.

                        Comment

                        Latest Articles

                        Collapse

                        • SEQadmin2
                          From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                          by SEQadmin2


                          Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                          The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                          ...
                          Yesterday, 10:05 AM
                        • SEQadmin2
                          Single-Cell Sequencing at an Inflection Point: Early Impacts of New Platforms and Emerging Trends
                          by SEQadmin2


                          With the launch of new single-cell sequencing platforms in 2026, the field stands at an exciting inflection point. This article surveys the most impactful advances in the field and discusses how they’re reshaping research in cancer, immunology, and beyond.


                          Introduction

                          Single-cell sequencing technologies have undergone remarkable advances over the past decade, transitioning from low-throughput experimental approaches to highly scalable platforms capable of...
                          05-22-2026, 06:42 AM
                        • SEQadmin2
                          Environmental Genomics in the Age of NGS: From Microbes to Conservation Strategies
                          by SEQadmin2

                          Studying ecosystems means dealing with complex, multi-species communities that are hard to observe at scale. This complexity, however, hides many important questions to be answered, from how biogeochemical cycles work and how climate change can affect species distribution to how conservation strategies can work best.


                          Genomics, particularly since the expansion of NGS, has transformed ecosystem ecology. By sequencing environmental DNA, we can now assess biodiversity without direct...
                          05-06-2026, 09:04 AM

                        ad_right_rmr

                        Collapse

                        News

                        Collapse

                        Topics Statistics Last Post
                        Started by SEQadmin2, Yesterday, 12:03 PM
                        0 responses
                        17 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, Yesterday, 11:40 AM
                        0 responses
                        13 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 05-28-2026, 11:40 AM
                        0 responses
                        29 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 05-26-2026, 10:12 AM
                        0 responses
                        31 views
                        0 reactions
                        Last Post SEQadmin2  
                        Working...