  • saikumarkv
    replied
    Hi everyone,
    In a previous experiment I used some MIRA enriched DNA (methylated) for a single read library prep (using Illumina Chip-seq kit) and sequencing, which worked very well. Right now I have lots of ChIP samples to sequence of various amounts (2-20ng). We decided we will use PE adapters for all sample preps from here on. I'm testing the NEBNext DNA sample prep kit and Illumina PE sample prep kits with different amounts of DNA (1-10ng). The question I have is whether anyone adopted the PE sample prep kit (where they use lot more reagents and enzymes) to the ChiIP-sample prep kit. Of course I will change the adapter and primer concentrations. I'm also testing TUFTS custom adapters and primers they they were gracious to share. Thanks in advance.

    Leave a comment:


  • hon
    replied
    So you said you combined 2 ChIP assay DNA to make library which works, do you know that equal to how many cells? Dose your transcriptional factor bind to a lot of genomic targets? I found if a TF doesn't have many targets may require even more cells. Right now I don't even bother to measure the ChIP DNA as I think that can not guarantee the ChIP seq will work. Even you ChIP with IgG which can give several ng of ChIP DNA that will make a good library but not give you any significant peaks. Seems people more incline to use more than 10 millions cells for TF ChIP, while histone ChIP can be in million range.

    Leave a comment:


  • xenia.zhang
    replied
    Hi Hon,

    Thanks for your suggestion. Current my explanation regarding your question is still not enough ChIP DNA. For histone modification, you don't need lots of starting cell since its so abundant and widespread. But for transcription factor, it's much weaker and rare. I know people in my lab who use half a million cell to do transcription factor chipseq. Although he did get result, total peak number compared with those using at least 20million cells is very small, and signal is not strong. Current my way is to use every chromatin prep for 2 ChIP, instead of 5 as before, and pool 2 chiped DNA together for library construction. I consulted experienced lab and they said that they've been successfully constructed library with 1-3ng of ChIPed DNA, but it's not always successful.

    I suggest you can use picogreen (from invitrogen) to quantify your ChIPed DNA. I tried once, and although I haven't been able to get linear standard curve, my DNA is in the range, and after calculation, is around 3ng per ChIP (that's why I choose to combine two chips). The sequencing result seems to be fine now.

    yah. That's my answer. Hope it will help!

    good luck!
    Sherry




    Originally posted by hon View Post
    Hi Sherry,

    Have you figured your problem yet on low peak signal? Recently I have been in similar situations that couple of my histone ChIP and Anti-HA Tag ChIP are not working out quite well, and like you said resembling to those ChIP of inefficient antibody. I kind of in line with zhaoj who suggest that the ChIP DNA used for making library is not enough. Previously I chip-seq two samples, K4me3 and K27me3, both of them working out really fine, although I only use 1 millions cells for chip assay. I thought I could get similar success by using same number of cells for one transcriptional factor ChIP (HA tagged), and one histone ChIP (K79me2), however the sequencing results turned out are not good, e.g. not many peaks, and high FDR (30%). Similar to you, I am quite confident that the antibody should be ok, as I checked it by QPCR to show good enrichments on my known targets. While I am repeating the ChIP seq with more cells and waiting for the results, I wonder there could be other problems that cause the failure of ChIP seq?

    For your bioanalyzer questions, I have tried the high sensitivity kit that can not detect the ChIP DNA, I guess because the quantity of ChIP DNA is really low, plus that they are in smear (not a single band), so even let's say if you have 5ng total but the smear may have 1000 bands, each band end up has 5pg, is already reach the limit of detection. In other aspect I guess all people using bioanalyser should pay attention is the bioanalyser is very sensitive to salt that means if there's minimal salt in your ChIP DNA that will ruin the measurement, I have seen this many times already, especially for phenol purified ChIP DNA which is not very clean and always contains a lot of salt. Hope this helps.

    Leave a comment:


  • hon
    replied
    Hi Sherry,

    Have you figured your problem yet on low peak signal? Recently I have been in similar situations that couple of my histone ChIP and Anti-HA Tag ChIP are not working out quite well, and like you said resembling to those ChIP of inefficient antibody. I kind of in line with zhaoj who suggest that the ChIP DNA used for making library is not enough. Previously I chip-seq two samples, K4me3 and K27me3, both of them working out really fine, although I only use 1 millions cells for chip assay. I thought I could get similar success by using same number of cells for one transcriptional factor ChIP (HA tagged), and one histone ChIP (K79me2), however the sequencing results turned out are not good, e.g. not many peaks, and high FDR (30%). Similar to you, I am quite confident that the antibody should be ok, as I checked it by QPCR to show good enrichments on my known targets. While I am repeating the ChIP seq with more cells and waiting for the results, I wonder there could be other problems that cause the failure of ChIP seq?

    For your bioanalyzer questions, I have tried the high sensitivity kit that can not detect the ChIP DNA, I guess because the quantity of ChIP DNA is really low, plus that they are in smear (not a single band), so even let's say if you have 5ng total but the smear may have 1000 bands, each band end up has 5pg, is already reach the limit of detection. In other aspect I guess all people using bioanalyser should pay attention is the bioanalyser is very sensitive to salt that means if there's minimal salt in your ChIP DNA that will ruin the measurement, I have seen this many times already, especially for phenol purified ChIP DNA which is not very clean and always contains a lot of salt. Hope this helps.

    Originally posted by xenia.zhang View Post
    Hi all,

    I have a couple chipseq libraries that show really good qPCR enrichment both before and after library construction, but sequencing results show very low signal, which resembles some other libraries where antibody is not optimal, though I know the Ab I use is a very good one. Then I start to suspect I have low ChIP enriched DNA for library construction, and tried to quantify concentration on high sensitivity bioanalyzer DNA kit. Unfortunately, for most of them, I can't detect anything, though the kit claims to be able to detect as low as 5pg/ul.

    I wonder whether anyone else has this problem of quantifying chip enriched DNA with bioanalyzer high sensitivity dna kit, even with good Chiped dna samples? Then I wouldn't worry too much. Or it's really problem of my sample?

    Or does any of you have any experience in this similar good qPCR, bad sequencing situation?
    Million thanks to all!



    -Sherry

    Leave a comment:


  • xenia.zhang
    replied
    captainentropy,

    Thanks for your reply!

    Actually I'm at the step before library construction. I figure I have to have decent amount of ChIP enriched DNA before I start doing library construction.

    'half of the peaks are real' means, for example, I got 300 peaks from peak calling software (MACS/Cisgenome, etc). I manually checked these peaks and compared with other lab members' similar result and only about half of these peaks seem real. The other peaks are just background peaks that also exist in a control sample, or are too short to be real.

    Hope this is clear?

    thanks.


    -Sherry

    Leave a comment:


  • xenia.zhang
    replied
    Hi zhaoj,

    Thanks for your answer! That's also what i suspect. By combining several ChIP, do you combine at the final elution step, or you actually use more chromatin prep?

    THanks.



    -Sherry

    Leave a comment:


  • zhaoj
    replied
    Hi Sherry
    I had that problem before, I suspect the quantity of DNA was low in my CHIP sample (hard to quantitate), so I combined several ChIP and worked fine in sequencing.

    Leave a comment:


  • captainentropy
    replied
    from what I've been told about the Bioanalyzer is that it's not very good for quantitation purposes. I use it just to see what my library looks like (is there a peak around 80 or 125 bp indicating adapter carryover). According to Illumina, even if you see nothing on the bioanalyzer you may be able to generate clusters for sequencing. Even when I've had a tiny peak on the bioanalyzer chromatogram the sequencing was successful. For quantitation I use either a Nanodrop or Qubit. Qubit is more sensitive at low DNA concentrations. I have not done qPCR on my samples but the guy who does my sequencing does qPCR (on my samples) and I've never been told there was a problem.

    What exactly do you mean when you say "About half of the peaks callled are real." Are you saying that the half that are "real" passed QC and uniquely mapped to the genome? I assume you have a bioanalyzer chromatogram? Was there a big peak around 70 or 124 bp? If so that is due to adapter carryover in the size selection step. I had that happen a couple of times and one time much less than half the reads mapped.

    Leave a comment:


  • xenia.zhang
    replied
    BTW, immunoprecipitation itself does work. About half of the peaks callled are real. The others are basically either background or 'peaks' around 30bp, which are false positives.

    Leave a comment:


  • xenia.zhang
    replied
    Hi all,

    I have a couple chipseq libraries that show really good qPCR enrichment both before and after library construction, but sequencing results show very low signal, which resembles some other libraries where antibody is not optimal, though I know the Ab I use is a very good one. Then I start to suspect I have low ChIP enriched DNA for library construction, and tried to quantify concentration on high sensitivity bioanalyzer DNA kit. Unfortunately, for most of them, I can't detect anything, though the kit claims to be able to detect as low as 5pg/ul.

    I wonder whether anyone else has this problem of quantifying chip enriched DNA with bioanalyzer high sensitivity dna kit, even with good Chiped dna samples? Then I wouldn't worry too much. Or it's really problem of my sample?

    Or does any of you have any experience in this similar good qPCR, bad sequencing situation?
    Million thanks to all!



    -Sherry

    Leave a comment:


  • captainentropy
    replied
    athos, I agree with cjohns comment. I use Qubit for what I suspect will be really low concentration samples and Nanodrop for post-PCR quantitation. I've compared Qubit and Nanodrop on low [DNA] samples and they're usually close in numbers but percentage-wise they aren't even close (eg. Qubit says 6 ng/ul, Nanodrop says 4 ng/ul). Also, the tech in the Bioanalyzer facility says not to trust the Bioanalyzer for quantitation purposes as it has ~20% CV!

    jawdekar, yes, you can definitely amplify after the ligation step. However, be warned that you may get massive amplification of concatemerized adapters that weren't removed in the clean-up step (Qiagen minelute). Thus if you want to amplify at this point it's important to use the right amount of adapters in the ligation step. As the Illumina protocol is written, using a 1:10 dilution is typically waaaay too many adapters. I use 1:30 as does another person I know that gets beautiful libraries (and that's ChIPed DNA from flies so there is tons of DNA per ChIP to start with). Basically you don't want your sample DNA to be the limiting factor but too many unligated adapters are problematic as they often become the preferred template for amplification and ultimately sequencing.

    Leave a comment:


  • alere
    replied
    ChIP seq enrichment

    Hi everyone,
    I'm preparing the ChIP enriched DNA for the ChIP seq.
    I'm wondering if is possible (and better) do the size selection directly on the ChIP sample before the amplification step as the ILLUMINA protocol recommend. Is possible to see any band on a 2% gel loading the ChIP from 20x 10 6 cells?? In this case the adaptor ligation and the library amplification step will more specific leading to an effective 200bp long fragment enrichment.
    Thanks.

    Leave a comment:


  • tomv
    replied
    You can definitely gel purify after amplification. This is what we have done for a long time. We are just now testing whether its better to do it before, after or before AND after but the results are not in yet. I'd be interested to know if anyone else has tested this.

    Leave a comment:


  • jawdekar
    replied
    ChIP-Seq help

    hi,

    can anyone tell me if I can enrich the adapter-modified DNA fragments by PCR first followed by the size selection of the library. This is because I am starting will approximately 10ng of ChIP'ed DNA and it seems unlikely that I can visualize 10ng on a gel for me to excise DNA in the ~200bp range.

    thanks

    Leave a comment:


  • Chipper
    replied
    Hi,
    thanks for the heads up on the gel purification. How many cells do you generally use? I ment to say that contamination can be an issue if you dont have the adapters on when running the gel, but it only a few thousand such reads in our runs so not a real problem.

    Leave a comment:

Latest Articles

Collapse

  • seqadmin
    Advanced Tools Transforming the Field of Cytogenomics
    by seqadmin


    At the intersection of cytogenetics and genomics lies the exciting field of cytogenomics. It focuses on studying chromosomes at a molecular scale, involving techniques that analyze either the whole genome or particular DNA sequences to examine variations in structure and behavior at the chromosomal or subchromosomal level. By integrating cytogenetic techniques with genomic analysis, researchers can effectively investigate chromosomal abnormalities related to diseases, particularly...
    09-26-2023, 06:26 AM
  • seqadmin
    How RNA-Seq is Transforming Cancer Studies
    by seqadmin



    Cancer research has been transformed through numerous molecular techniques, with RNA sequencing (RNA-seq) playing a crucial role in understanding the complexity of the disease. Maša Ivin, Ph.D., Scientific Writer at Lexogen, and Yvonne Goepel Ph.D., Product Manager at Lexogen, remarked that “The high-throughput nature of RNA-seq allows for rapid profiling and deep exploration of the transcriptome.” They emphasized its indispensable role in cancer research, aiding in biomarker...
    09-07-2023, 11:15 PM
  • seqadmin
    Methods for Investigating the Transcriptome
    by seqadmin




    Ribonucleic acid (RNA) represents a range of diverse molecules that play a crucial role in many cellular processes. From serving as a protein template to regulating genes, the complex processes involving RNA make it a focal point of study for many scientists. This article will spotlight various methods scientists have developed to investigate different RNA subtypes and the broader transcriptome.

    Whole Transcriptome RNA-seq
    Whole transcriptome sequencing...
    08-31-2023, 11:07 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by seqadmin, Today, 09:38 AM
0 responses
8 views
0 likes
Last Post seqadmin  
Started by seqadmin, 09-27-2023, 06:57 AM
0 responses
11 views
0 likes
Last Post seqadmin  
Started by seqadmin, 09-26-2023, 07:53 AM
0 responses
13 views
0 likes
Last Post seqadmin  
Started by seqadmin, 09-25-2023, 07:42 AM
0 responses
17 views
0 likes
Last Post seqadmin  
Working...
X