Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • shay covo
    Junior Member
    • Jan 2015
    • 2

    #1

    Tapestation or Fragment analyzer

    I am now about to buy tapestation of agilent or fragment analyzer of AATI

    I will use both to QC libraries genomic DNA and RNA

    We will not process many samples a week but we need a reliable low maintenance machine
    What would you recommend? Do you have any clear bias towards any one of them? why
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    We only have the bioanalyzer. I think it is preferable for us because the high sensitivity chips are more sensitive than the tapestation.
    That said, the down side of a bioanalyzer is that you have to run samples 10-12 at a time. So if you only have a couple of samples to run the reagent costs become quite outrageous. And the kits themselves have a limited shelf life, so that can make the cost even worse if you don't run the instrument fairly heavily.

    --
    Phillip

    Comment

    • GagneUW1
      Junior Member
      • May 2015
      • 1

      #3
      I am making the same decision between a tape station and a AATI fragment analyzer. What did you end up deciding on this and why? Any information helps. Thank you!

      Comment

      • shay covo
        Junior Member
        • Jan 2015
        • 2

        #4
        I chose tapestation

        It is much easier to operate. Other members of the department were very negative about the fragment analyzer.

        Comment

        • bilyl
          Member
          • Aug 2013
          • 52

          #5
          Originally posted by shay covo View Post
          I am now about to buy tapestation of agilent or fragment analyzer of AATI

          I will use both to QC libraries genomic DNA and RNA

          We will not process many samples a week but we need a reliable low maintenance machine
          What would you recommend? Do you have any clear bias towards any one of them? why
          Why not a Labchip GX? It can run 96-well plates of samples and can perform regular NGS quant as well as genomic DNA assays.

          Comment

          • seqdx
            Junior Member
            • Jun 2015
            • 3

            #6
            Originally posted by shay covo View Post
            It is much easier to operate. Other members of the department were very negative about the fragment analyzer.
            Really? Did you run any samples? The TapeStation is easier to use than the Fragment Analyzer for small sample sets, but the benefits stop there. Data quality matters most. The lack of resolution, sensitivity, and dynamic range will limit applications and worse, force you to keep using slab gels and the BioAnalyzer you hoped to get away from.

            See the attached restriction digest comparison of the TapeStation versus the Fragment Analyzer. Data speaks volumes, resolution is critical for sizing accuracy.

            NGS read lengths are getting longer and longer. If you demo the two also explore sensitivity limitations for samples over 2 kb.
            Attached Files

            Comment

            • Ingeneious
              Member
              • Dec 2014
              • 30

              #7
              I have used the BA a lot in the past and recently switched to the TapeStation. For convenience, it is pretty simple and accuracy is good enough to use for NGS library prep. My previous lab was using it for hundreds of clinical samples, after calibrating it with a BA. It also gives you the option between running a few or many samples, which can lead to reduced costs against the BA. If you run without a ladder and use expired tapes (they only last 2 weeks after first use), you can really cut down costs.

              On the flip side, you can buy around 3 BA's for 1 TapeStation and the per sample cost may be higher (off the top of my head ~$41 for 16 sample D1000 tapes/reagents, $65 for HSD1000 tapes/reagents vs ~$30? for 1 x 12 sample BA D1000 chip/reagents). I also do notice the resolution loss between the BA.

              Someone else in our group was an advocate of the LabChip GX, but she said prep time was ~50 minutes vs 10 minutes for the BA and about 5 minutes for the TapeStation. You can have results for the TapeStation in as few as 10 minutes vs 30 for the BA. That can certainly help with development.

              No one that I know has experience with the Fragment Analyzer, but I do know we had one sitting idle, while BA's and TapeStations were being used frequently...

              Some of the major downsides have been that with only 25 - 1000bp resolution for the D1000, some samples may bleed into the ladder and give inaccurate concentrations. The BA offered a chip with up to 7500bp.

              We have used the gDNA tape only once and it was good enough to check the overall integrity, plus gave a DIN number for accuracy. Based on seqdx, if you need the resolution, it looks like that might be a problem.
              Last edited by Ingeneious; 06-18-2015, 10:46 AM.

              Comment

              • barrmur
                Member
                • Nov 2009
                • 26

                #8
                New tape station, 4200, to be released soon, likely shipping in September. May be wise to wait and see what the offer is....

                Looks like it will be capable of running 96 samples in "walk away" mode.

                Comment

                • Ingeneious
                  Member
                  • Dec 2014
                  • 30

                  #9
                  Yeah, my boss said a new one was coming out that would do 96 samples. Good to hear details.

                  Comment

                  • Smoki
                    Junior Member
                    • Nov 2011
                    • 8

                    #10
                    The comment about Labchip GX prep time is silly. I have used both the BA and the GX, and the latter requires less hands-on work and skill.

                    You would get to 50 min only if you take into account that you have to put the chip and reagents on room temp half an hour before using them.
                    Not to mention that you prep the chip once and then it's good for the whole day. You just add samples to the microplate. That certainly helps with development.

                    And I absolutely love the GX software.

                    Comment

                    • Ingeneious
                      Member
                      • Dec 2014
                      • 30

                      #11
                      So same as BA and TapeStation + 20min prep. What's the per sample cost of the LapChip GX with reagents/chips?

                      Comment

                      • Smoki
                        Junior Member
                        • Nov 2011
                        • 8

                        #12
                        Unless you will have a very low throughput in the lab (i.e. always less than 10 samples per day), the reagents cost should be considerably lower on the GX. But the initial investment in the instrument is higher.

                        Comment

                        • seqdx
                          Junior Member
                          • Jun 2015
                          • 3

                          #13
                          Originally posted by barrmur View Post
                          New tape station, 4200, to be released soon, likely shipping in September. May be wise to wait and see what the offer is....

                          Looks like it will be capable of running 96 samples in "walk away" mode.
                          Agilent should focus on the resolution and sensitivity limitations of the TapeStation. The instrument simply lacks the resolution needed for accurate sizing of large NGS libraries, and the lack of sensitivity for samples over 2 kb make it useless for single-cell applications such as Fluidigm C1 transcriptomics (Clonetech). Why spend hard earned capital funds on a system that can't handle the diverse assay requirements of a rapidly evolving field.

                          Comment

                          • Mini
                            Junior Member
                            • Jan 2016
                            • 1

                            #14
                            Originally posted by Ingeneious View Post
                            Yeah, my boss said a new and improved Winsol review was coming out that would do 96 samples. Good to hear details.
                            Oh really? Any idea when this is coming out?
                            Last edited by Mini; 02-15-2022, 04:36 AM.

                            Comment

                            • GenoMax
                              Senior Member
                              • Feb 2008
                              • 7142

                              #15
                              Originally posted by Mini View Post
                              Oh really? Any idea when this is coming out?
                              Looks like it is already available: http://www.genomics.agilent.com/en/p...questid=859015

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM
                              • SEQadmin2
                                New Genomics Technologies Take Aim at Long-Standing Limits
                                by SEQadmin2


                                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                ...
                                09-28-2026, 10:25 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              28 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              20 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-15-2026, 10:52 AM
                              0 responses
                              38 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-14-2026, 05:39 AM
                              0 responses
                              44 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...