    DNA degradation and size selection

    Hi,

    I used the Covaris to shear gDNA for Illumina libraries, but something went wrong because instead of the expected 500 bp fragments I got extensive (and pretty random) degradation. I am attaching the results from the bioanalyzer: sample 4 is a + control and looked fine, but I had to manually set high and low marker peaks for the remaining 3 samples. S1 and S2, in particular look really degraded. However, they seem to have enough 200-300 bp fragments that might be recovered with a size selection. Has anyone ever successfully built and sequenced illumina libraries from similarly over-fragmented gDNA?

    Thank you in advance!
    Attached Files
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    It depends on the kit that you are going to use and amount of DNA available. You can do size selection first and use at least half the recommended input amount for library prep. Library can be prepared with lower amount as well but that would affect library diversity and coverage uniformity. There are some kits for low input DNA below 10 ng.

    Comment

    • Hamid
      Senior Member
      • Sep 2009
      • 108

      #3
      Please check the integrity of your DNA samples by running some unsheard samples on an agarose gel. Also please note that bioanalyzer chips are very sensitive to contaminants. How were these DNA samples purified?

      Thank you

      Hamid

      Comment

      Latest Articles

      Collapse

      • SEQadmin2
        Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
        by SEQadmin2



        CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

        Despite this, “CRISPR helped turn genome editing from a specialized technique into
        ...
        07-31-2026, 11:01 AM
      • SEQadmin2
        Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
        by SEQadmin2


        Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

        The systematic characterization of the human proteome has
        ...
        07-20-2026, 11:48 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, Today, 12:22 PM
      0 responses
      8 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 08-11-2026, 10:35 AM
      0 responses
      11 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 08-06-2026, 07:41 AM
      0 responses
      30 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 08-03-2026, 10:13 AM
      0 responses
      48 views
      0 reactions
      Last Post SEQadmin2  
      Working...