    mRNA fragmentation/weird Bioanalyzer profile

    Hello everybody
    I'm going to prepare RNAseq libraries from cassava. First I did was purify mRNA from my samples using polydT(25) beads (life technologies). Then using 5x First Strand Buffer (Invitrogen) I incubated the samples at 94C for different times (2, 5 and 7 minutes). I sent this samples to Bioanalyzer before first strand cDNA synthesis and the profile is really weird. There's a peak around 200 bp (so the fragmentation seems to be good, especially 5 and 7 min: 2 was not enough) but the peak itself and the baseline are very noisy (distorted). I've read that can be a problem with the chip itself, but also because of not enough washing after the fragmentation. The thing is that the protocol I'm following doesn't not incluse a washing after this fragmentation. What do you think about this profile?. Any suggestion of what to do with these samples?. Thanks!
    Attached Files
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    You would expect to see a smooth electropherogram if sample does not contain much salt. Samples containing first strand buffer (especially in sensitive Pico Chip) are expected to run as the profiles shown. Also, the actual sizes will be a bit larger than shown because salts accelerate nucleic acids migration in Chip.

    There is no need to do a clean-up if the protocol does not call for it because the buffer composition would be part of the next step in protocol. RT reaction will be primed with random hexamers and because they will prime from different regions from RNA the final library inserts will be shorter than RNA fragment length. I think you can go ahead following the kit protocol.

    Comment

    Latest Articles

    Collapse

    • SEQadmin2
      Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
      by SEQadmin2



      Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

      Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
      ...
      10-14-2026, 09:55 AM
    • SEQadmin2
      New Genomics Technologies Take Aim at Long-Standing Limits
      by SEQadmin2


      Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

      We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
      ...
      09-28-2026, 10:25 AM

    ad_right_rmr

    Collapse

    News

    Collapse

    Topics Statistics Last Post
    Started by SEQadmin2, 10-22-2026, 11:13 AM
    0 responses
    23 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 10-22-2026, 10:48 AM
    0 responses
    20 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 10-15-2026, 10:52 AM
    0 responses
    36 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 10-14-2026, 05:39 AM
    0 responses
    42 views
    0 reactions
    Last Post SEQadmin2  
    Working...