Originally posted by kerplunk412
View Post
Unconfigured Ad
Collapse
X
-
Mass of genome in one human cell is 6.6 pg, so in a DNA solution of 10 ng/ul we would have equivalent of DNA from 1515 cells which would be 45,450,000 fragments of 100kb. Most standard extraction methods will result in fragments less than 100kb. So, I do not see how one can justify that 45.5 million fragments in 1ul will aggregate in a solution to give 10x variation in consecutive reads. I think just a gentle flick would be enough to have a homogenous solution (if sample was frozen) and vortexing definitely would damage large DNA fragments.
-
-
The 500 reaction kit is cheap (60 cents/sample) and it takes a matter of minutes to add a few extra standards. You are correct regarding the linear detection range; however, I think you should review linear regression.Originally posted by nucacidhunter View PostLinear detection range of PicoGreen is four orders of magnitude in 1 ng/ml to 1000ng/ml DNA concentration. As far as one calibrates fluorometer at 0 and 1000 range there is no need to any other concentration in between or standard curve. It seems to be waste of money and time. With correct calibration one needs only to multiply the fluorescence value in dilution factor to calculate original concentration of DNA sample.
Comment
-
-
Your logic makes sense to me, so maybe the difference in size before and after vortexing does not explain my observations. However, I tested this fairly rigorously and a few of my colleagues have tried this as well, so I can say with confidence that with the gDNA samples I was working with a gentle flick was not enough to get a consistent reading, vortexing was required. As far as damaging the DNA, I am pretty sure 10 seconds of vortexing will not cause enough DNA fragmentation to matter for most NGS applications. If it was that easy to fragment DNA into small pieces no one would need to buy a Covaris!Originally posted by nucacidhunter View PostMass of genome in one human cell is 6.6 pg, so in a DNA solution of 10 ng/ul we would have equivalent of DNA from 1515 cells which would be 45,450,000 fragments of 100kb. Most standard extraction methods will result in fragments less than 100kb. So, I do not see how one can justify that 45.5 million fragments in 1ul will aggregate in a solution to give 10x variation in consecutive reads. I think just a gentle flick would be enough to have a homogenous solution (if sample was frozen) and vortexing definitely would damage large DNA fragments.
Edit: I should also mention that the variation seen before vortexing was at most about 2x. Variation after vortexing was ~1%.Last edited by kerplunk412; 03-05-2015, 04:52 PM.
Comment
-
Latest Articles
Collapse
-
by SEQadmin2
Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.
The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
...-
Channel: Articles
06-02-2026, 10:05 AM -
-
by SEQadmin2
With the launch of new single-cell sequencing platforms in 2026, the field stands at an exciting inflection point. This article surveys the most impactful advances in the field and discusses how they’re reshaping research in cancer, immunology, and beyond.
Introduction
Single-cell sequencing technologies have undergone remarkable advances over the past decade, transitioning from low-throughput experimental approaches to highly scalable platforms capable of...-
Channel: Articles
05-22-2026, 06:42 AM -
ad_right_rmr
Collapse
News
Collapse
| Topics | Statistics | Last Post | ||
|---|---|---|---|---|
|
Sequencing the Two-Toed Sloth Genome Reveals Jumping Genes Tied to Its Extreme Metabolism
by SEQadmin2
Started by SEQadmin2, 06-09-2026, 11:58 AM
|
0 responses
15 views
0 reactions
|
Last Post
by SEQadmin2
06-09-2026, 11:58 AM
|
||
|
Started by SEQadmin2, 06-05-2026, 10:09 AM
|
0 responses
26 views
0 reactions
|
Last Post
by SEQadmin2
06-05-2026, 10:09 AM
|
||
|
Started by SEQadmin2, 06-04-2026, 08:59 AM
|
0 responses
37 views
0 reactions
|
Last Post
by SEQadmin2
06-04-2026, 08:59 AM
|
||
|
Started by SEQadmin2, 06-02-2026, 12:03 PM
|
0 responses
61 views
0 reactions
|
Last Post
by SEQadmin2
06-02-2026, 12:03 PM
|
Comment