Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • JBKri
    Member
    • Jan 2014
    • 80

    #1

    How to deal with too dilute libraries?

    So my worst fear came true; we have a deadline, we spent the last of the library kit, and our PCR-free libraries all came out well below the minimum for loading to the instrument (MiSeq). The concentrations from qPCR range from 0.14 to 0.7 nM

    I remember reading about an alternative denaturing protocol, with HCl used to neutralize the NaOH instead of the large dilution after denaturation. But I did not see any specific instructions about the quantities and concentration of HCl. Does anyone have experience with this?
    Edit: I see in the NextSeq protocol they use Tris-HCl pH 7, which enables denaturation of libraries down to 0.5 nM with a final loading concentration of 20 pM. We will have about half of that, so we will probably end up with a rather low cluster density.

    Another option is to concentrate the libraries with speedvac, but I am not too eager about this. We tried it before on a library, and we somehow lost everything. Even if we reduce the volume to 1/10th, some samples will still be a bit too dilute.

    Another option is to do enrichment PCR, (see my post here, but I am not sure about the primers we have or the conditions.

    Any help will be greatly appreciated!
    Jon
    Last edited by JBKri; 10-09-2015, 08:31 AM.
  • GA-J
    Member
    • Jul 2015
    • 28

    #2
    I would just use this library(0.7nM), denature with 0.1N NaoH, then it comes to 7pM after 100x dilution with HT1 buffter. If your library's size is about 400bp, you may load at this concentration or adjust using your library's size adjust factor.

    If you still worry about the 0.1N NaoH,after denature, use same concentration and volume of HCl (0.1N and same volume that you used NaOH for denature), then you are good to go as usual.
    Last edited by GA-J; 10-09-2015, 07:01 PM.

    Comment

    • MU Core
      Member
      • Apr 2008
      • 60

      #3
      You may find these instructions helpful.
      Attached Files

      Comment

      • JBKri
        Member
        • Jan 2014
        • 80

        #4
        Thanks for the responses. I ended up using the NextSeq protocol, which gave me a final concentration of 13.5 pM of the pool, and it worked just fine.
        Jon

        Comment

        • ishap
          Junior Member
          • Mar 2014
          • 1

          #5
          HI, My library concentrations are between 0.2 to 0.8ng/ul. Could you please let me know how exactly you proceeded to pool and denature and load your samples when they were this low?
          Thanks alot in advance!

          Comment

          • JBKri
            Member
            • Jan 2014
            • 80

            #6
            Originally posted by ishap View Post
            HI, My library concentrations are between 0.2 to 0.8ng/ul. Could you please let me know how exactly you proceeded to pool and denature and load your samples when they were this low?
            Thanks alot in advance!
            What type of library is this? How did you measure the concentration? For many types of libraries it is best to measure the concentration by qPCR. The resulting concentrations are expressed in pM or nM, not ng/ul, and depend on the average fragment size of the library.

            Normally, the MiSeq denaturation protocol needs a minimum of 2 or 4 nM library concentration, which limits the final loading concentration to maximum 10 or 20 pM. Let's say you have a concentration of 0.2 ng/ul, with fragment size 350 bp, then the concentration is 0.87 nM, which is not enough if using the MiSeq denaturation protocol.

            The NextSeq protocol is here.
            In short, you mix 1 volume of library with 1 volume of 0.2N NaOH, incubate for 5 min at room temp, then add 1 volume of 200mM Tris-HCl pH7 (room temp). Then add Hyb buffer to make the final volume 600 ul.
            Let's say you want to load at a final concentration of 13 pM. Then you need to take a 0.87nM pool volume of 9 ul, to which you add 9 ul of 0.2 NaOH, incubate, then add 9 ul 200 mM Tris-HCl and 573 ul Hyb buffer. Final concentration should be 13.05 pM.

            Edit: By chance, yesterday I used this protocol for the first time in several years. Our pool was just 130 pM, but with the NextSeq denaturation protocol we made the final concentration around 15 pM, and it worked just fine.
            Last edited by JBKri; 06-14-2018, 04:01 AM.

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
              by SEQadmin2




              NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
              10-30-2026, 06:46 AM
            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-28-2026, 09:57 AM
            0 responses
            41 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-28-2026, 09:51 AM
            0 responses
            25 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            74 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            48 views
            0 reactions
            Last Post SEQadmin2  
            Working...