  • Library quantification: opinions?

    Am curious to survey how groups quantify their libraries prior to putting them on the sequencer (or sending them to a vendor, as is our case).

    qPCR would generally seem to be the winner to me; very high sensitivity & accuracy and you are measuring only the active molecules

    However, it seems that Bioanalyzer and NanoDrop have their adherents. BioAnalyzer does get get you some size distribution information as well, but my impression is that with low concentration libraries one is pushing the sensitivity limits of both BioAnalyzer & NanoDrop (with a resultant hit in the precision of the measurement).

  • #2
    You're correct that qPCR is the most sensitive and accurate. However, we've gotten fairly reproducible results using PicoGreen dye fluorimetry for quantification. We switched from Nanodrop b/c, in addition to poor sensitivity, some libraries contained contaminants that absorbed in the UV range. The dye is specific for dsDNA, and accurate for samples down to at least 0.4 ng/ul. We follow up with BioAnalyzer high-sensitivity DNA chips for size determination and to detect adapter dimer contamination (which will affect qPCR measurements). However, its concentration values can vary 2-3X from reality.

    Comment


    • #3
      The Nanodrop is no good for exactly the reasons HESmith gave.
      The Bioanalyzer is good for sizing, but only gives ballpark quantification I've found.
      Fluorometry is acceptable with some variability. I've used the Qubit system.
      qPCR is the very best overall and is now what I use exclusively (with the BioA for sizing). I use the Kapa Library quantification kits on a Roche 480.

      Comment


      • #4
        We use Bioanalyzer for sizing only.

        Kapa qPCR for Illumina library quantification. As of August 1, Kapa kits are now compatible with the small RNA and GEX libraries. We consistently hit our target of 300-400K clusters/tile.

        SOLiD qPCR assay (from Life Tech) for SOLiD library quantification.

        Comment


        • #5
          We also use qPCR for quantification and the Bioanalyzer for sizing. We use the nanodrop for other parts of prep.

          Comment


          • #6
            Bioanalyzer for sizing, pico green for quantitation. We will test the qPCR once that kit arrives. I really would not miss titration if pPCR is reproducible and hits the right numbers...

            Comment


            • #7
              Originally posted by lee View Post
              We use Bioanalyzer for sizing only.

              Kapa qPCR for Illumina library quantification. As of August 1, Kapa kits are now compatible with the small RNA and GEX libraries. We consistently hit our target of 300-400K clusters/tile.

              SOLiD qPCR assay (from Life Tech) for SOLiD library quantification.
              I saw that Kapa also has a SOLiD library quantification kit, does anyone ever compared it with the ABI qPCR reagents?

              Comment


              • #8
                Originally posted by lee View Post
                We use Bioanalyzer for sizing only.

                Kapa qPCR for Illumina library quantification. As of August 1, Kapa kits are now compatible with the small RNA and GEX libraries. We consistently hit our target of 300-400K clusters/tile.

                SOLiD qPCR assay (from Life Tech) for SOLiD library quantification.

                Hi Lee,
                could you please tell me your experience with the Kappa standards? I am working with Kapa kits to perform the qPCR but not with their standards.
                Thanks

                Comment


                • #9
                  I'm trying to quantify a library produced using the DSN normalisation protocol, which looks to be very low concentration...

                  Is it necessary to use the kit from Kapa Biosystems or could I use the qPCR reagents we have in the lab (SYBR GreenER from Invitrogen on an Applied Biosystems 7500)?

                  Many thanks,

                  Adam

                  Comment


                  • #10
                    Our lab uses both Bioanalyzer and qPCR (the Taqman assay described in Quail et al, 2009 -- Current Protocols) for quality control and quantitation. As HESmith observed, the Bioanalyzer quantitation results are often 2-3X higher than qPCR values. I'm relieved to find that we aren't the only ones getting results of this type. However, we have occasionally had libraries where the Bioanalyzer quantitation was 10x higher than qPCR. Upon titration we found that the qPCR value gave a closer approximation of the actual concentration of cluster-forming fragments in the library.

                    However, I was wondering if anyone else has observed this (Bioanalyzer quantitation 10 times higher than qPCR) and if they can suggest any causes for this?

                    Thanks in advance,

                    KAT

                    Comment


                    • #11
                      I think the accuracy is also dependent on the library prep method. Generally, the more PCR, the more accurate it's going to be to quantify by Bioanalyser/Picogreen. I think this is because amplification only occurs on those DNAs that have the correct adapter sequences ligated. Unfortunately, more PCR also means more PCR duplicates.

                      We mainly run target enriched samples that have undergone quite a few cycles of PCR (sometimes over 20 cycles in total) and the Bioanalyser plus Picogreen has been fairly consistent. When we run ChIPSeq and MeDipSeq we've seen more variablility, probably because the protocols used have limited PCR (4 cycles).
                      Running 454 genomic libraries show the most variation as there is no PCR involved, so you're at the mercy of your ligation efficiency - we would always quantify using qPCR for these libraries as the variation is way too high. For example plotting the estimated concentration of using qPCR vs Bioanalyser for our last 20 libraries gave an R^2 of 0.43, with Bioanalyser quantitation between 2 and 6 times higher than qPCR. Alternatively 454 amplicon libraries can be quantified by Bioanalyser as in theory, all molecules in the library should contain the adpaters - we found these to be consistent when using Bioanalyser.

                      Comment


                      • #12
                        We have used the Kapa 454 kit for both Rapid libraries (gDNA & cDNA) and for amplicons for some time now, but its not working perfectly. Well, it work fine for RL, but amplicons is more of a challenge. As we are a core facility people send us amplicons produced with fusion primers, nested PCR and ligation products. Even if qPCR results and DNA chip results are in agreement, the emPCR often give very varying enrichment numbers, and we really have no good explanation for it, but we see a differense between fusion and nested generated products. I´m considering skipping qPCR for amplicons and stick to chip + qubit.

                        Cheers!
                        M.

                        Comment


                        • #13
                          I think that if you have access to a qPCR machine, there really is no reason not to do qPCR.

                          I prefer probe-based qPCR to SYBR Green qPCR, though, since with a probe the signal is not affected by the amplicon length. This allows you to use the same standard curve for all of your libraries.

                          Comment


                          • #14
                            I am wondering is any one observed that the qPCR reading is higher than the Qubit? I mean this is impossible right as the qPCR only measure the molecule which ligate with adapter.

                            Comment


                            • #15
                              Originally posted by anant View Post
                              I am wondering is any one observed that the qPCR reading is higher than the Qubit? I mean this is impossible right as the qPCR only measure the molecule which ligate with adapter.
                              I think if you're using the stock Illumina adapters and primers, qPCR should be accurate. It has worked very well so far for us. Are you sure you are considering the correct total size for your library and not just insert size? Also, what are you using for a qPCR standard?

                              I have gotten fewer spots than expected from qPCR, but this was when using a custom read 1 primer.

                              Comment

                              Latest Articles

                              Collapse

                              • seqadmin
                                Current Approaches to Protein Sequencing
                                by seqadmin


                                Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
                                04-04-2024, 04:25 PM
                              • seqadmin
                                Strategies for Sequencing Challenging Samples
                                by seqadmin


                                Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
                                03-22-2024, 06:39 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by seqadmin, 04-11-2024, 12:08 PM
                              0 responses
                              26 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 04-10-2024, 10:19 PM
                              0 responses
                              29 views
                              0 likes
                              Last Post seqadmin  
                              Started by seqadmin, 04-10-2024, 09:21 AM
                              0 responses
                              25 views