    SureSelect Hyb#4 - SDS

    Hi all,

    Could anyone tell me the role of SDS (Hyb#4 buffer I'm assuming) in the hybridisation reaction of SureSelectXT?

    It is causing us some problems and we were wondering how disastrous it would be to titrate it back, or do away with it altogether.

    Thanks!
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    SDS destabilises bonds between probes and library fragments (one of hybridization stringency factors) thereby reduces non-specific hybridization. Omitting it from hyb reaction would result in lower on target capture.

    Comment

    • Squall
      Member
      • Feb 2015
      • 15

      #3
      Thanks nucacidhunter.

      Do you think that it is a linear response? So if we reduced the SDS by 20, 40, 80% etc, that the reduction in on-target would increase accordingly? Or is the SDS present in excess in the reaction?

      Comment

      • nucacidhunter
        Jafar Jabbari
        • Jan 2013
        • 1250

        #4
        I have not seen any formula to calculate the effects. I do not think that it is linear. One way to determine the effect empirically would be to reduce %SDS to the highest that downstream application can tolerate. If off-target levels are high then one can increase hybridisation temperature by 2C to compensate for reduced SDS content. Google searching DNA hybridization kinetics or dynamics should retrieve some related papers.

        Comment

        • melinn21
          Junior Member
          • Jan 2015
          • 1

          #5
          I am also experiencing issues with this. While our data seems to be coming out ok, I'm finding that if I heat and then cool the hyb buffer mix (as the protocol says to do if precipitation forms) I somehow lose volume and never have enough to add to the probes (even with 5% overage). It's very strange. If I measure the volume while its still warm I have the correct volume if I measure while RT it is not enough and the precipitate has essentially returned.

          Comment

          • Squall
            Member
            • Feb 2015
            • 15

            #6
            Thanks hunter for suggestions, will cycle them into the testing plans!

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              How Immunogenomics Decodes Immunity’s Genetic Blueprint
              by SEQadmin2




              The immune system’s power comes from its genetic diversity, allowing myriad threats to be neutralized through first recognizing foreign antigens. That diversity is also what makes the immune system so difficult to study. Recent advances in sequencing technology and computational biology, however, are giving researchers new tools to understand immune responses and immune-related diseases in greater detail.

              This convergence of genetics, immunology, and computation...
              Yesterday, 05:41 AM
            • SEQadmin2
              Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
              by SEQadmin2



              CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

              Despite this, “CRISPR helped turn genome editing from a specialized technique into
              ...
              07-31-2026, 11:01 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 08-24-2026, 10:32 AM
            0 responses
            42 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 08-20-2026, 11:17 AM
            0 responses
            48 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 08-18-2026, 10:05 AM
            0 responses
            55 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 08-13-2026, 12:22 PM
            0 responses
            50 views
            0 reactions
            Last Post SEQadmin2  
            Working...