    ChIP-Seq library - Microplex - bad profiles

    We have trouble with ChIP-seq library...
    Does anybody have experience with diagenode MicroPlex Library Preparation kit v2 C05010014 (48 rxns, 48 indices)? We have bad bioanalyser profiles after library preparation with enrichment in very large fragments which seem to be concateners... ??? after denaturation and renaturation the profile come back to something more normal... very strange... we use low DNA quantity

    so your expertise would be really appreciated!
    Thanks a lot in advance!
    Last edited by Valerie_Gaudin; 04-05-2016, 08:37 AM. Reason: Unprecise title
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    The kit seems to be rebranded ThruPLEX DNA-seq kit. Your description indicates PCR overcycling. For a definitive answer posting library (before PCR purification) and input profiles would be helpful.

    Comment

    • Valerie_Gaudin
      Junior Member
      • Mar 2015
      • 5

      #3
      Thanks I attached the aberrant profiles
      Attached Files

      Comment

      • Valerie_Gaudin
        Junior Member
        • Mar 2015
        • 5

        #4
        In attached file, a more complete description of the different steps..
        Attached Files

        Comment

        • Diagenode
          Junior Member
          • Apr 2016
          • 3

          #5
          ChIP-Seq library - Microplex

          Hi Valerie,

          Our tech support team have tried to call you today to discuss further your problem.
          But in summary we believe that your library his overamplified. We recommand to amplify less cycles.
          Let's discuss further on the phone...

          Best regards,
          Diagende's support team

          Comment

          • nucacidhunter
            Jafar Jabbari
            • Jan 2013
            • 1250

            #6
            With equivalent of this kit (ThruPLEX DNA-Seq) following PCR cycle number based on input gives good results. To avoid overcycling you can do 1 less cycle based on recommendation then run a sample of PCR on ScreenTape or other device. If you get <2ng/ul you can do extra cycles to increases yield by assuming doubling yield with each cycle. You should also see smaller peak at 60-70 bp in profile for primers. Lack of small peak indicates that reaction has run out of primers and all sorts of artefacts will show up. Glad that tech support is looking after the issue.

            Comment

            • Valerie_Gaudin
              Junior Member
              • Mar 2015
              • 5

              #7
              Thanks for advices! checking the 60-70 bp peak is a good point!! We will do!
              We also purified the sample before doing library with Ampure. Indeed for input DNA we got a peak on bioanalyser around 30-50 bp (steps1-2 in atached file) do you know whether that can affect library preparation or if there are better alternative to remore this small DNA fraction?
              Thanks in advance!
              Attached Files

              Comment

              • nucacidhunter
                Jafar Jabbari
                • Jan 2013
                • 1250

                #8
                Looking at #2 it seems that you have successfully removed the small peak (~50 bp) present in #1. That should not affect library prep and some protocols even call for size-selection in a narrow range. Purified input should be eluted in TE0.1 or Tris buffer because high EDTA adversely affects library prep with this kit.

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                  by SEQadmin2


                  I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

                  Here are nine questions we think about, in roughly the order they matter, before...
                  06-18-2026, 07:11 AM
                • SEQadmin2
                  From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                  by SEQadmin2


                  Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                  The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                  ...
                  06-02-2026, 10:05 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, 06-26-2026, 11:10 AM
                0 responses
                15 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 06-17-2026, 06:09 AM
                0 responses
                49 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 06-09-2026, 11:58 AM
                0 responses
                107 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 06-05-2026, 10:09 AM
                0 responses
                125 views
                0 reactions
                Last Post SEQadmin2  
                Working...