  • nucacidhunter
    replied
    1000 ng total RNA as input is OK but as you can see the size of larger peak is proportional to the PCR cycle number indicating over cycling. I would think that 7 cycles would have been enough.

    Using Good quality RNA results in high PCR yields and I would suggest using BA DNA1000 kit or diluting the libraries to below 1 ng/ul for running on HS Chip.

    Keeping PCR cycles the same for the samples that are going to be compared for differential expression is also a good practice to avoid PCR induced biases. This is less important if gene expression analysis is not the aim.

    Leave a comment:


  • Andersen
    replied
    Originally posted by nucacidhunter View Post
    Possible reasons:
    1-Too many PCR cycles
    2- Too much input

    I wonder what was the difference between samples 4 and 5 with others and also if samples have been run on BA without dilution.
    Not a whole lot was done differently, between 4-5 and the others. I believe that the only difference was that 4-5 had 8 cycles and 1-3 had 10 cycles and 6 had 9 cycles. But this was due to the lower amplification test score before amplifying the libraries.

    We use 1000 ng as starting material, is this too high, or is it a too much input for the bioanalyzer you think of?



    Originally posted by pmiguel View Post
    Bubble products.
    The good news is that they usually sequence fine.
    TruSeq kits ask for 15 cycles of amplification, which is usually crazy high. Later cycles begin to kinetically favor products re-annealing to other over primers annealing/new strand polymerization. Since library molecules are a complex mixture only the adapters anneal leaving a floppy single stranded section in the middle of the heteroduplex amplicons that causes them to migrate slowly during electrophoresis. Or so the lore goes.

    --
    Phillip
    I like your phrasing - Or so the lore goes. Is there any way to control if this is the case, or can I avoid this by any means?


    THANKS TO BOTH OF YOU FOR YOUR HELP!

    Leave a comment:


  • pmiguel
    replied
    Bubble products.
    The good news is that they usually sequence fine.
    TruSeq kits ask for 15 cycles of amplification, which is usually crazy high. Later cycles begin to kinetically favor products re-annealing to other over primers annealing/new strand polymerization. Since library molecules are a complex mixture only the adapters anneal leaving a floppy single stranded section in the middle of the heteroduplex amplicons that causes them to migrate slowly during electrophoresis. Or so the lore goes.

    --
    Phillip

    Leave a comment:


  • nucacidhunter
    replied
    Possible reasons:
    1-Too many PCR cycles
    2- Too much input

    I wonder what was the difference between samples 4 and 5 with others and also if samples have been run on BA without dilution.

    Leave a comment:


  • Andersen
    started a topic Double peak on Truseq RNA-seq library

    Double peak on Truseq RNA-seq library

    Hi All,

    I have prepared some Truseq RNA-seq libraries and i have 2 peaks.

    I have been reading some previous threads that it could be bead carry-over, but I do not really see any beads in my library and I think some of the libraries look a bit different. Do you have any suggestions to what the problem could be?

    I used another protocol than usual to remove rRNA but I do not think it should be the problem (Epicentre - Ribo Zero : http://www.illumina.com/products/rib...-kit-faqs.html)
    Attached Files

Latest Articles

Collapse

  • SEQadmin2
    Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
    by SEQadmin2



    CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

    Despite this, “CRISPR helped turn genome editing from a specialized technique into
    ...
    07-31-2026, 11:01 AM
  • SEQadmin2
    Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
    by SEQadmin2


    Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

    The systematic characterization of the human proteome has
    ...
    07-20-2026, 11:48 AM
  • SEQadmin2
    Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
    by SEQadmin2



    Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
    ...
    07-09-2026, 11:10 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, Today, 10:13 AM
0 responses
10 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 07-31-2026, 02:55 AM
0 responses
21 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 07-24-2026, 12:17 PM
0 responses
19 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 07-23-2026, 11:41 AM
0 responses
17 views
0 reactions
Last Post SEQadmin2  
Working...