1000 ng total RNA as input is OK but as you can see the size of larger peak is proportional to the PCR cycle number indicating over cycling. I would think that 7 cycles would have been enough.
Using Good quality RNA results in high PCR yields and I would suggest using BA DNA1000 kit or diluting the libraries to below 1 ng/ul for running on HS Chip.
Keeping PCR cycles the same for the samples that are going to be compared for differential expression is also a good practice to avoid PCR induced biases. This is less important if gene expression analysis is not the aim.
Unconfigured Ad
Collapse
X
-
Not a whole lot was done differently, between 4-5 and the others. I believe that the only difference was that 4-5 had 8 cycles and 1-3 had 10 cycles and 6 had 9 cycles. But this was due to the lower amplification test score before amplifying the libraries.Originally posted by nucacidhunter View PostPossible reasons:
1-Too many PCR cycles
2- Too much input
I wonder what was the difference between samples 4 and 5 with others and also if samples have been run on BA without dilution.
We use 1000 ng as starting material, is this too high, or is it a too much input for the bioanalyzer you think of?
I like your phrasing - Or so the lore goes. Is there any way to control if this is the case, or can I avoid this by any means?Originally posted by pmiguel View PostBubble products.
The good news is that they usually sequence fine.
TruSeq kits ask for 15 cycles of amplification, which is usually crazy high. Later cycles begin to kinetically favor products re-annealing to other over primers annealing/new strand polymerization. Since library molecules are a complex mixture only the adapters anneal leaving a floppy single stranded section in the middle of the heteroduplex amplicons that causes them to migrate slowly during electrophoresis. Or so the lore goes.
--
Phillip
THANKS TO BOTH OF YOU FOR YOUR HELP!
Leave a comment:
-
Bubble products.
The good news is that they usually sequence fine.
TruSeq kits ask for 15 cycles of amplification, which is usually crazy high. Later cycles begin to kinetically favor products re-annealing to other over primers annealing/new strand polymerization. Since library molecules are a complex mixture only the adapters anneal leaving a floppy single stranded section in the middle of the heteroduplex amplicons that causes them to migrate slowly during electrophoresis. Or so the lore goes.
--
Phillip
Leave a comment:
-
Possible reasons:
1-Too many PCR cycles
2- Too much input
I wonder what was the difference between samples 4 and 5 with others and also if samples have been run on BA without dilution.
Leave a comment:
-
Double peak on Truseq RNA-seq library
Hi All,
I have prepared some Truseq RNA-seq libraries and i have 2 peaks.
I have been reading some previous threads that it could be bead carry-over, but I do not really see any beads in my library and I think some of the libraries look a bit different. Do you have any suggestions to what the problem could be?
I used another protocol than usual to remove rRNA but I do not think it should be the problem (Epicentre - Ribo Zero : http://www.illumina.com/products/rib...-kit-faqs.html)Attached FilesTags: None
Latest Articles
Collapse
-
by SEQadmin2
The immune system’s power comes from its genetic diversity, allowing myriad threats to be neutralized through first recognizing foreign antigens. That diversity is also what makes the immune system so difficult to study. Recent advances in sequencing technology and computational biology, however, are giving researchers new tools to understand immune responses and immune-related diseases in greater detail.
This convergence of genetics, immunology, and computation...-
Channel: Articles
09-01-2026, 05:41 AM -
ad_right_rmr
Collapse
News
Collapse
| Topics | Statistics | Last Post | ||
|---|---|---|---|---|
|
Started by SEQadmin2, 09-09-2026, 12:14 PM
|
0 responses
30 views
0 reactions
|
Last Post
by SEQadmin2
09-09-2026, 12:14 PM
|
||
|
Started by SEQadmin2, 09-09-2026, 11:33 AM
|
0 responses
25 views
0 reactions
|
Last Post
by SEQadmin2
09-09-2026, 11:33 AM
|
||
|
Started by SEQadmin2, 09-03-2026, 10:22 AM
|
0 responses
33 views
0 reactions
|
Last Post
by SEQadmin2
09-03-2026, 10:22 AM
|
||
|
Started by SEQadmin2, 09-02-2026, 12:32 PM
|
0 responses
48 views
0 reactions
|
Last Post
by SEQadmin2
09-02-2026, 12:32 PM
|
Leave a comment: