  • nucacidhunter
    replied
    1000 ng total RNA as input is OK but as you can see the size of larger peak is proportional to the PCR cycle number indicating over cycling. I would think that 7 cycles would have been enough.

    Using Good quality RNA results in high PCR yields and I would suggest using BA DNA1000 kit or diluting the libraries to below 1 ng/ul for running on HS Chip.

    Keeping PCR cycles the same for the samples that are going to be compared for differential expression is also a good practice to avoid PCR induced biases. This is less important if gene expression analysis is not the aim.

    Leave a comment:


  • Andersen
    replied
    Originally posted by nucacidhunter View Post
    Possible reasons:
    1-Too many PCR cycles
    2- Too much input

    I wonder what was the difference between samples 4 and 5 with others and also if samples have been run on BA without dilution.
    Not a whole lot was done differently, between 4-5 and the others. I believe that the only difference was that 4-5 had 8 cycles and 1-3 had 10 cycles and 6 had 9 cycles. But this was due to the lower amplification test score before amplifying the libraries.

    We use 1000 ng as starting material, is this too high, or is it a too much input for the bioanalyzer you think of?



    Originally posted by pmiguel View Post
    Bubble products.
    The good news is that they usually sequence fine.
    TruSeq kits ask for 15 cycles of amplification, which is usually crazy high. Later cycles begin to kinetically favor products re-annealing to other over primers annealing/new strand polymerization. Since library molecules are a complex mixture only the adapters anneal leaving a floppy single stranded section in the middle of the heteroduplex amplicons that causes them to migrate slowly during electrophoresis. Or so the lore goes.

    --
    Phillip
    I like your phrasing - Or so the lore goes. Is there any way to control if this is the case, or can I avoid this by any means?


    THANKS TO BOTH OF YOU FOR YOUR HELP!

    Leave a comment:


  • pmiguel
    replied
    Bubble products.
    The good news is that they usually sequence fine.
    TruSeq kits ask for 15 cycles of amplification, which is usually crazy high. Later cycles begin to kinetically favor products re-annealing to other over primers annealing/new strand polymerization. Since library molecules are a complex mixture only the adapters anneal leaving a floppy single stranded section in the middle of the heteroduplex amplicons that causes them to migrate slowly during electrophoresis. Or so the lore goes.

    --
    Phillip

    Leave a comment:


  • nucacidhunter
    replied
    Possible reasons:
    1-Too many PCR cycles
    2- Too much input

    I wonder what was the difference between samples 4 and 5 with others and also if samples have been run on BA without dilution.

    Leave a comment:


  • Andersen
    started a topic Double peak on Truseq RNA-seq library

    Double peak on Truseq RNA-seq library

    Hi All,

    I have prepared some Truseq RNA-seq libraries and i have 2 peaks.

    I have been reading some previous threads that it could be bead carry-over, but I do not really see any beads in my library and I think some of the libraries look a bit different. Do you have any suggestions to what the problem could be?

    I used another protocol than usual to remove rRNA but I do not think it should be the problem (Epicentre - Ribo Zero : http://www.illumina.com/products/rib...-kit-faqs.html)
    Attached Files

Latest Articles

Collapse

  • SEQadmin2
    How Immunogenomics Decodes Immunity’s Genetic Blueprint
    by SEQadmin2




    The immune system’s power comes from its genetic diversity, allowing myriad threats to be neutralized through first recognizing foreign antigens. That diversity is also what makes the immune system so difficult to study. Recent advances in sequencing technology and computational biology, however, are giving researchers new tools to understand immune responses and immune-related diseases in greater detail.

    This convergence of genetics, immunology, and computation...
    09-01-2026, 05:41 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, 09-09-2026, 12:14 PM
0 responses
30 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 09-09-2026, 11:33 AM
0 responses
25 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 09-03-2026, 10:22 AM
0 responses
33 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 09-02-2026, 12:32 PM
0 responses
48 views
0 reactions
Last Post SEQadmin2  
Working...