    Help with denaturing low concentration Illumina PE Library

    Hi all
    I have an Illumina PE library that is of very low concentration (~1.5nM). If I speedvac I should be able to get 2nM to do the denaturation, but I'll have to do it in 10µL rather than the stated 20µL as I don't have the volume. Has anyone done this before?
    Thanks
  • GW_OK
    Senior Member
    • Sep 2009
    • 411

    #2
    Add 13.3 ul of your 1.5nM library to 4.6 ul of EB buffer (10mM Tris)
    Add 1 ul 2N NaOH and incubate 5 minutes
    Add 1 ul 2N HCL to neutralize your NaOH
    You've now got 20 ul of a 1000pM ssDNA library.
    Dilute to your favorite loading concentration with chilled hyb buffer.

    Using this method I have successfully loaded libraries as low as to 193pM.

    Comment

    • ScottC
      Senior Member
      • Jan 2008
      • 244

      #3
      I can vouch for that method too... I've successfully sequenced libraries the same way. I think there's another thread on the topic here somwhere.

      GW_OK: Out of interest, how do you assess that your low-concentration libraries have been successfully prepared (other than sequencing)? Do you use real-time PCR?

      Scott.

      Comment

      • GW_OK
        Senior Member
        • Sep 2009
        • 411

        #4
        qPCR with the Kapa kit. It's the only way to go in my opinion. (Well, at least the qPCR part. You can pick your own reagents.)

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
          by SEQadmin2




          NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
          Today, 06:46 AM
        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 10-28-2026, 09:57 AM
        0 responses
        22 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-28-2026, 09:51 AM
        0 responses
        14 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 11:13 AM
        0 responses
        67 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 10:48 AM
        0 responses
        47 views
        0 reactions
        Last Post SEQadmin2  
        Working...