Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • aspiring
    Junior Member
    • Apr 2017
    • 4

    #1

    How to get rid of Shoulder Peak in Illumina Library

    Hello Everyone!
    I am preparing Library using Illumina TruSeq (Starting Total RNA 1.5 microgram with RIN: 10). I have been trying to do this a couple of times but every time I do this a shoulder peak of ~248 bp slowly creeps in. I would be very much thankful if you could please help me explain this and get rid of it.
    Thank you very much!
    Attached Files
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    I am not sure the instrument that you have run the libraries. Those peaks may not be related to library prep. I have seen those when the instrument is not in a stable position for instance the table is moved or it is near a centrifuge and the vibrations affects the instrument during loading.

    Comment

    • luc
      Senior Member
      • Dec 2010
      • 469

      #3
      What type of sample are you studying? This might be caused by a predominant short transcript?

      Comment

      • aspiring
        Junior Member
        • Apr 2017
        • 4

        #4
        Originally posted by luc View Post
        What type of sample are you studying? This might be caused by a predominant short transcript?
        Hi Thank you for your response. I am studying HeLa RNA samples. The OD 260/280 and OD 260/230 ratios were 2.03 & 2.01 whereas the RIN value was 10, which shows that the starting RNA sample is not degraded and very pure.

        Comment

        • aspiring
          Junior Member
          • Apr 2017
          • 4

          #5
          Originally posted by nucacidhunter View Post
          I am not sure the instrument that you have run the libraries. Those peaks may not be related to library prep. I have seen those when the instrument is not in a stable position for instance the table is moved or it is near a centrifuge and the vibrations affects the instrument during loading.
          Thank you for your response. The instrument has always been stable as we have dedicated table for the bio analyzer. There is a small table top centrifuge to spin down the samples but we use it before turning on the bio analyzer.

          Comment

          • nucacidhunter
            Jafar Jabbari
            • Jan 2013
            • 1250

            #6
            It this case luc explanation is more likely the cause. Aligning reads to reference or BLAST search should help to identify overrepresented transcripts/library fragments. Sometimes they could be result of insufficient polyA selection or depletion that can be corrected but sometimes result of biological state of cells which may not require any action.

            Comment

            • aspiring
              Junior Member
              • Apr 2017
              • 4

              #7
              Originally posted by nucacidhunter View Post
              I am not sure the instrument that you have run the libraries. Those peaks may not be related to library prep. I have seen those when the instrument is not in a stable position for instance the table is moved or it is near a centrifuge and the vibrations affects the instrument during loading.
              It seems as if the stability of tape station affects the generation of peak while running the sample in the tape.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                The Factors Reshaping Genome Assembly
                by SEQadmin2



                When the Human Genome Project declared the human genome finished in 2003, portions of the sequence were still missing, including many of its most repetitive and structurally complex regions.1 Still, for the next two decades, researchers analyzed their sequencing data against the single linear reference that came out of this effort. Released in 2013 as GRCh38, that reference has powered countless discoveries, but it has gaps and cannot capture the genetic diversity of the world’s...
                11-24-2026, 07:04 AM
              • SEQadmin2
                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                by SEQadmin2





                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification,...
                10-30-2026, 06:46 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 11-23-2026, 11:11 AM
              0 responses
              39 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 11-17-2026, 06:27 AM
              0 responses
              47 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 11-12-2026, 07:59 AM
              0 responses
              68 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 11-05-2026, 09:25 AM
              0 responses
              68 views
              0 reactions
              Last Post SEQadmin2  
              Working...