    RNA integrity number Drosophila

    Hello @all
    as we want to start an RNA-Seq analysis we try to determine the quality of our RNA (Drosophila heads) first. When we did this via Bioanalyzer an error occurred as expected. The measurement in a TapeStation results in a RIN though. I can`t explain why because still no 28S peak is visible. The RIN is the ration of 28S to 18S as I understood this.
    Therefore I don`t know if I can rely on that. Does someone have experience with that?
    Thank you!
    Jule

    ps: attached you see the trace of one sample with both machines
    Attached Files
  • ericchow
    Junior Member
    • Dec 2012
    • 4

    #2
    It's a result of the 28S from insects being split. See this article: https://www.ncbi.nlm.nih.gov/pmc/articles/PMC3016993/

    Comment

    • luc
      Senior Member
      • Dec 2010
      • 469

      #3
      As consequence from what Eric wrote, one often has to eyeball the RNA integrity for insect samples on the Bioanalyzer (I am looking ratio of the rRNA peak size to the smaller debris fragments).
      Your example trace looks great.

      Comment

      • Elugop
        Junior Member
        • Nov 2017
        • 2

        #4
        Thanks - I try it with these samples now. Fingers crossed

        Comment

        • HarbsSM
          Junior Member
          • Nov 2017
          • 1

          #5
          Hi,
          I'd suggest using the Bioanalyzer over the Tapestation in the future. It gives a more granular electropherogram, whereas the Tapestation directly images the gel image from the run. This results in a less detailed electropherogram. Your trace looks good, but the RIN/RINe values are effectively redundant for insect material. Eyes are the best measure.

          The link above is a great explanation!

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM
          • SEQadmin2
            New Genomics Technologies Take Aim at Long-Standing Limits
            by SEQadmin2


            Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

            We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
            ...
            09-28-2026, 10:25 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, Yesterday, 11:13 AM
          0 responses
          17 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, Yesterday, 10:48 AM
          0 responses
          18 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-15-2026, 10:52 AM
          0 responses
          36 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-14-2026, 05:39 AM
          0 responses
          42 views
          0 reactions
          Last Post SEQadmin2  
          Working...