    16S Library Plant Tissue Troubles

    Hi all I thought I would seek advice here as I have tried a number of tweaks and not getting the results I would expect. Here is the background

    DNA was extracted from plant stems to allow for a survey of bacterial endophytes using 16S amplicons in the v4 region

    Samples were extracted using a Qiagen DNeasy Plant extraction kit, samples were quantified and run on a gel to check for quality. Yes it will be mostly plant DNA but still, this was done.
    Library preparation was done following the Schloss Wet Lab SOP as I have done numerous times before for soil, roots, and yes some plant tissue (mostly combo runs when plant tissue as involved).
    Samples were pools and placed on MiSeq and the following stats were obtained on my latest v2 500 cycle (2 x 250bp) run.
    @ cycle 244 I got
    Density 722K/mm2
    Passing Filter 72.2%
    PhiX spiked at 9.2%
    Qscore >=30 32.9% and this only went down to about 25% by the end of the run

    This was not the first time to run this library as I did a nano flow cell run with it and had spike in PhiX and a much higher concentration to increase diversity and yet still ended up with a low pass filter and poor qscore.

    I have, in the past run a library with these stem samples (a subset of 4 samples and two different extraction types Qiagen and MoBio on each of the 4 samples, to determine the best extraction method/bacterial yield) on a nano flow cell (2 x 250) and got
    These stats:

    Density 520k/mm2
    Passing filter 88.5%
    Qscore >= 93.6%
    PhiX was at 13.6
    Yes there was a fair bit of chloroplast/host DNA contamination but was getting about 12-20 % out after processing through Mothur.

    I have also run other soil samples using the reagents and all are fine. I have also a large number of various library preps under my belt but am at a bit of loss on this one…

    Any ideas?
  • ATϟGC
    Member
    • Jun 2013
    • 56

    #2
    Hi Sean,

    Aside from PhiX, are you taking any measures to increase diversity in your amplicon libraries such as staggering/offsetting and/or adding other loci?


    At a glance, your cluster densities do not appear to be too high but if you have access to SAV (Sequence Analysis Viewer) for you run it might help if you posted the figures of %base/per cycle, density charts of the flow cell, and the density box plots. If not, the "per tile sequence quality" graph from FastQC might help show if there is differential quality over the length of the flow cell. Your amplicon run might not have been overclustered per se, but it might have been overclustered for a low-diversity run and 10% or more PhiX might not have been enough.

    Here is a document on Miseq Overclustering in case you find it helpful:

    The Nuffield Department of Medicine (NDM) at the University of Oxford has a global reach and significant breadth in terms of capabilities and capacity.


    My guess is that your nano Mobio/Qiagen run was of whole-genome libraries rather than amplicons and may have worked better due to their high diversity.

    I have seen people get lower PF% (~70%) with single amplicon sequencing despite moderate PhiX spike-in (~15-20%) if they do not stagger/offset their read 1 locus-specific primers.

    Comment

    • seanw
      Junior Member
      • Nov 2014
      • 3

      #3
      Follow up info.

      Hi,

      here is some clarification on things.

      There is no stagger within the primer pairs to increase the diversity and I have not added anything else, but this had not been an issue before (I guess it may be now). I also had a nano run of this library with a greater concentration(30-40%) of PhiX but that still had the issue.

      The MoBio/Qaigen subsample library run was a 16S amplicon using the same primer sets that are giving me problem. It was a trial run I used to make sure this was doable with these types of samples.


      Here are some screen shots and yes it certianly looks like a diversity issues...
      Attached Files

      Comment

      • ATϟGC
        Member
        • Jun 2013
        • 56

        #4
        Hi again Sean,

        Sorry for making that guess/assumption. I forgot that eukaryotes have 16S RNA as well.

        So it does appear that you have very low base diversity, you have rapidly declining q scores that appear to coincide with increasing %G at base 120 of read 1, and most of your clusters passing filter come from lower densities (an indicator of overclustering). I am still leaning toward your troubles being from low-diversity and moderate overclustering but it could also be from bad reagents or something else.

        This video goes over a lot of the same things as the document I added earlier:


        I have lost one amplicon run to an expired kit which gave us a bunch of junk G nucleotides and rapidly declining q-scores. A re-run with the same library with new MiSeq reagents gave us excellent data. That being said, I have seen some colleagues who do not stagger/offset and/or mix in other loci get worse %PF and Qscores than I do with the same sequencing provider.

        I have included 3 different offset/staggered amplicons in the 5 MiSeq amplicon runs I have done and they have all worked well with ~10-15% PhiX(aside from a 6th run that had the bad reagents). Perhaps someone with more experience interpreting SAV results will be able to tell you which problem your results are indicative of.

        Some screen shots from FastQC would likely help with diagnosis as well.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Nine Things a Sample Prep Scientist Thinks About Before Sequencing
          by SEQadmin2


          I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

          Here are nine questions we think about, in roughly the order they matter, before...
          06-18-2026, 07:11 AM
        • SEQadmin2
          From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
          by SEQadmin2


          Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


          The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
          ...
          06-02-2026, 10:05 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, Today, 05:37 AM
        0 responses
        5 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-26-2026, 11:10 AM
        0 responses
        16 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-17-2026, 06:09 AM
        0 responses
        50 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-09-2026, 11:58 AM
        0 responses
        109 views
        0 reactions
        Last Post SEQadmin2  
        Working...