    Bioanalyzing long ATAC-Seq libraries

    Hi All,

    I am part of a sequencing core, trying to QC a client's ATAC-Seq libraries via the bioanalyzer on a DNA high sensitivity chip. Prior to running the samples I checked the concentrations with a qubit fluorometer and diluted them all to ~4-6 ng/uL. Perhaps a little high for the high sensitivity chip, but not enough to overload it.

    After talking with the client I believed the problems was long fragments carrying over between samples, so I tried to space them out, only loading samples in position 1, 4, 7, and 10. Unfortunately, I am still getting garbage traces after the first sample. Does anyone have any recommendations for how to handle these samples?

    I've attached the traces as an example - you can see the characteristic ATAC peaks in sample 1, and then some higher molecular weight material near the marker, continuing into the next two traces.
    Attached Files
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    We had some Nextera libraries that looked like this recently -- turned out they were 10X overloaded.

    Also, they are a client's ATAC-seq library? We recently had a set of these where the client had done and ampure upper cut and given us the supernatant without further clean-up! So they were somewhere north of 1M NaCl plus whatever PEG was in the ampure. We pulled the library down by adding more ampure, followed by the normal ampure washing/eluting the beads-- then everything looked fine.

    --
    Phillip

    Comment

    • JoeKutch
      Member
      • Oct 2016
      • 17

      #3
      Thanks for the reply Philip:

      I don't think they are overloaded, but I can dilute further and try them again with some other samples. Qubit and then diluting should have prevented that.

      They are a client's libraries - the ATAC protocol he used does not call for any clean-up, so there aren't any PEG/salts in them. After finding that out, we recommended a bead uppercut with 0.4x beads followed by 2.5x to recollect the libraries. He gave us a very limited amount for QC or I'd try it myself. Once he brings those newly cleaned libraries I'll update.

      Thanks again for the help!

      Comment

      Latest Articles

      Collapse

      • SEQadmin2
        Nine Things a Sample Prep Scientist Thinks About Before Sequencing
        by SEQadmin2


        I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

        Here are nine questions we think about, in roughly the order they matter, before...
        06-18-2026, 07:11 AM
      • SEQadmin2
        From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
        by SEQadmin2


        Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


        The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
        ...
        06-02-2026, 10:05 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, Yesterday, 11:10 AM
      0 responses
      7 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 06-17-2026, 06:09 AM
      0 responses
      42 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 06-09-2026, 11:58 AM
      0 responses
      104 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 06-05-2026, 10:09 AM
      0 responses
      125 views
      0 reactions
      Last Post SEQadmin2  
      Working...