    discrepancy between qubit and bioanalyzer results

    When I do the QC for my bead cleaned-up pre-amplified cDNA (Smartseq2) using AATI fragment analyzer and bioanalyzer, the concentration was very low (0.01ng/uL) and almost no peaks. But when I measured the same samples using Qubit, I got about 0.1ng/uL. Why are these methods giving different concentrations? Thanks!
  • cmbetts
    Senior Member
    • Jun 2012
    • 120

    #2
    The FA/BA is probably a more accurate measurement (although I trust qubit more for concentrations >1ng/ul). It's easy to get background readings of 0.1-0.3ng/ul by qubit. Did you run a negative control for comparison?

    Comment

    • lijing2019
      Junior Member
      • May 2019
      • 5

      #3
      Originally posted by cmbetts View Post
      The FA/BA is probably a more accurate measurement (although I trust qubit more for concentrations >1ng/ul). It's easy to get background readings of 0.1-0.3ng/ul by qubit. Did you run a negative control for comparison?
      OK, I see. I will include a negative control to see how it goes. Thank you for your reply.

      Comment

      Latest Articles

      Collapse

      • SEQadmin2
        Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
        by SEQadmin2



        CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

        Despite this, “CRISPR helped turn genome editing from a specialized technique into
        ...
        07-31-2026, 11:01 AM
      • SEQadmin2
        Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
        by SEQadmin2


        Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

        The systematic characterization of the human proteome has
        ...
        07-20-2026, 11:48 AM
      • SEQadmin2
        Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
        by SEQadmin2



        Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
        ...
        07-09-2026, 11:10 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, Today, 10:13 AM
      0 responses
      13 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 07-31-2026, 02:55 AM
      0 responses
      23 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 07-24-2026, 12:17 PM
      0 responses
      19 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 07-23-2026, 11:41 AM
      0 responses
      18 views
      0 reactions
      Last Post SEQadmin2  
      Working...