Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • BTS
    Member
    • Sep 2010
    • 19

    #1

    NEB Fragmentase

    We are trying to fragment genomic DNA using NEB's Fragmentase enzyme with a 20-23 minute incubation. Our goal is to reach fragments near 300bp for sequencing on GAIIx. Has anyone used this enzyme and if so what are your thoughts on it? When cleaning with a Qiagen Qiaquick kit, I'm getting recoveries of 35-40% initial starting material (starting with 1-5 ug). I'm thinking that a lof of DNA may be chopped small and then not caught by the columns, they don't capture fragments under 70bp, but I am not able to confirm this. Thanks!
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    We have not tried Fragmentase yet.

    Your estimation of the amount of starting DNA is based upon what assay? Unless you use fluorimetry or visualization (on a gel) against a known mass standard, then it is likely your will over-estimate the amount of starting DNA you have. There are lots of confounding contaminants in a typical DNA prep (RNA -- whether you RNAsed or not, and trace phenol being the two I most commonly see.) UV spectrophotometry will almost always give you an over-estimate for this reason. (After fragmenting your purification step will largely remove these contaminants. Hence the low apparent yields.)

    --
    Phillip

    Comment

    • BTS
      Member
      • Sep 2010
      • 19

      #3
      Oops, I forgot to include that. We quantified our DNA via Qubit Flourometer prior to fragmenting AND after.

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        Then you are likely correct. Although it is possible the Qiaquick columns have lower capacity than you expect.

        You could run aliquots from a sample before and after Qiaquick on a gel or bioanalyzer chip to verify.

        Mechanical fragmentation methods become more inefficient at shearing as they approach their lower size limit. As a result it is possible to tune them to put a large percentage of fragments above a size limit. With an enzyme there would be no real physical limitation on cleaving smaller and smaller fragments. So losing a higher percentage of your DNA on the low end may part of the price you pay for using an enzymatic method.

        On the plus side, the fragment ends may be more ligate-able after enzymatic fragmentation. Possibly.

        --
        Phillip

        Comment

        • ChickSeq
          Junior Member
          • Jun 2010
          • 5

          #5
          How much elution buffer are you using to elute your sample off of the Qiaquicks? There is a pretty direct correlation between the amount used to elute and the elution recovery. See the attached file.

          Hope that helps,

          Jen
          Attached Files

          Comment

          • BTS
            Member
            • Sep 2010
            • 19

            #6
            I was eluting with two washes of 30 uL in a QiaQuick column and diluting to a final volume of 100 uL. I read similar information about the columns as what you posted and have changed my protocol to one wash with 100 uL. The protocol you posted is much more detailed and useful than the one I found. Thanks a bunch!

            Comment

            • ChickSeq
              Junior Member
              • Jun 2010
              • 5

              #7
              No problem, glad to help.

              Comment

              • HGENETIC
                still a novice
                • Jul 2010
                • 34

                #8
                We tried using this enzyme, we had it automated on a Biomek Fx and it seemed to work OK the first time we did the expt. Then we repeated everything using some 48 samples and everything failed, I contacted NEB and they first tried to tell me I hadn't mixed the enzyme properly so I told them there was a mix step incorporated into the robot program. In the end they gave us our money back and we now use a Covaris, it seems the enzyme is very unstable and doesn't give consistent results. My advice would be to try something different if you can.

                Comment

                • vehuardo
                  Member
                  • Jun 2012
                  • 11

                  #9
                  I have used NEBnext fragmentase for approx 300 bacterial genomes. Works well!

                  Comment

                  Latest Articles

                  Collapse

                  • SEQadmin2
                    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                    by SEQadmin2



                    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                    ...
                    10-14-2026, 09:55 AM
                  • SEQadmin2
                    New Genomics Technologies Take Aim at Long-Standing Limits
                    by SEQadmin2


                    Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                    We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                    ...
                    09-28-2026, 10:25 AM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by SEQadmin2, Yesterday, 11:13 AM
                  0 responses
                  17 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, Yesterday, 10:48 AM
                  0 responses
                  18 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-15-2026, 10:52 AM
                  0 responses
                  36 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-14-2026, 05:39 AM
                  0 responses
                  42 views
                  0 reactions
                  Last Post SEQadmin2  
                  Working...