    PCR amplicon library - DNA smearing

    I've been making two-step amplicon libraries for years now with very few problems. The original amplicons are tailed with partial Nextera xt adapters and then I use Phusion HF to add the rest of the adapter with 6 cycles of PCR and KAPA Pure beads for cleanup. Recently, the yield has been much lower than expected and, rather than a pretty clean peak on the Bioanalyzer, there's been a lot of high molecular weight smearing. The samples labeled SPS01-07 are after the second round of amplication while the samples labeled Pre are before. Suggestions as to what might be going on?
    Click image for larger version

Name:	Screen Shot 2019-11-07 at 1.34.30 PM.png
Views:	1
Size:	58.4 KB
ID:	309612
  • luc
    Senior Member
    • Dec 2010
    • 469

    #2
    Not a good explanation, nevertheless: perhaps your stock of index primers is degrading or got contaminated? You might want to order a few fresh oligos and compare?

    Comment

    • aesterle
      Junior Member
      • Jan 2020
      • 2

      #3
      Hey there, microgirl:

      I was wondering if you tried what luc suggested and if this worked.

      Also, do you have a low annealing temperature on your PCR cycles? I figure it's stayed the same, but since other things have been changed, maybe the thermocycler programming needs to be altered as well.

      I've only been doing library prep for around 1.5 yr, but it surprised me how low the annealing temps in a lot of protocols are set to. We anneal at 50C!

      Comment

      Latest Articles

      Collapse

      • SEQadmin2
        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
        by SEQadmin2



        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
        ...
        10-14-2026, 09:55 AM
      • SEQadmin2
        New Genomics Technologies Take Aim at Long-Standing Limits
        by SEQadmin2


        Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

        We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
        ...
        09-28-2026, 10:25 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, 10-22-2026, 11:13 AM
      0 responses
      23 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-22-2026, 10:48 AM
      0 responses
      20 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-15-2026, 10:52 AM
      0 responses
      36 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-14-2026, 05:39 AM
      0 responses
      42 views
      0 reactions
      Last Post SEQadmin2  
      Working...