    How to pool ATAC-seq libraries in equimolar concentrations?

    Hi! I have 24 ATAC libraries prepared for sequencing. For this, I want to pool them all together. I ran them on the TapeStation and I have the distinctive nucleosomal banding pattern (see attached files). I'm hesitating a bit on how to properly estimate the molarity of every library when I have multiple bands, since the Tape is not always detecting the same bands. Any suggestion about this? Should I just use Qubit quantification for pooling?
    Attached Files
  • kmcarr
    Senior Member
    • May 2008
    • 1181

    #2
    Originally posted by ECorrales View Post
    Hi! I have 24 ATAC libraries prepared for sequencing. For this, I want to pool them all together. I ran them on the TapeStation and I have the distinctive nucleosomal banding pattern (see attached files). I'm hesitating a bit on how to properly estimate the molarity of every library when I have multiple bands, since the Tape is not always detecting the same bands. Any suggestion about this? Should I just use Qubit quantification for pooling?
    Use Region Analysis on the TapeStation software instead of Peak Analysis to determine an average size for each sample. Set up a Region from ~150-1000bp on all samples and use that calculate the average size. Use this average size and Qubit measurements for each to then calculate a molar concentration for each library.

    Before you do this though, correct the Upper Marker in wells E1, A2, C2, F2 and G2 in the 10-24 file. The Upper Marker should be manually set tot he taller of the two peaks (Upper Marker peak height should alway be almost 2X the Lower Marker peak height.)

    Comment

    • ECorrales
      Junior Member
      • Nov 2019
      • 3

      #3
      Originally posted by kmcarr View Post
      Before you do this though, correct the Upper Marker in wells E1, A2, C2, F2 and G2 in the 10-24 file. The Upper Marker should be manually set tot he taller of the two peaks (Upper Marker peak height should alway be almost 2X the Lower Marker peak height.)
      Thank you for your reply. Would you say these peaks corresponds to the upper marker? I was assuming they were larger un-tagmented regions, so I was planning to do a second size selection round with Ampure beads to remove these peaks. But if these are the markers then a second size selection might be unnecessary.

      Comment

      • kmcarr
        Senior Member
        • May 2008
        • 1181

        #4
        Originally posted by ECorrales View Post
        Thank you for your reply. Would you say these peaks corresponds to the upper marker? I was assuming they were larger un-tagmented regions, so I was planning to do a second size selection round with Ampure beads to remove these peaks. But if these are the markers then a second size selection might be unnecessary.
        What I am saying is that for those 5 lanes I mentioned there are 2 very closely spaced peaks, one labeled as a peak of ~1,260-1,270bp and the peak just after that labeled as the Upper marker. The software has incorrectly identified the later migrating peak as the Upper marker. In the TapeStation analysis software you should select the peak which is labeled ~1,270 and manually set that as the upper marker. If you do not correct the upper marker then the size calculation for these lanes will be incorrect.

        Comment

        • ECorrales
          Junior Member
          • Nov 2019
          • 3

          #5
          Originally posted by kmcarr View Post
          What I am saying is that for those 5 lanes I mentioned there are 2 very closely spaced peaks, one labeled as a peak of ~1,260-1,270bp and the peak just after that labeled as the Upper marker. The software has incorrectly identified the later migrating peak as the Upper marker. In the TapeStation analysis software you should select the peak which is labeled ~1,270 and manually set that as the upper marker. If you do not correct the upper marker then the size calculation for these lanes will be incorrect.
          Oh, I see. Thank you so very much for clarifying it!

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
            by SEQadmin2



            CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

            Despite this, “CRISPR helped turn genome editing from a specialized technique into
            ...
            07-31-2026, 11:01 AM
          • SEQadmin2
            Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
            by SEQadmin2


            Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

            The systematic characterization of the human proteome has
            ...
            07-20-2026, 11:48 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, Today, 10:05 AM
          0 responses
          5 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 08-13-2026, 12:22 PM
          0 responses
          31 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 08-11-2026, 10:35 AM
          0 responses
          24 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 08-06-2026, 07:41 AM
          0 responses
          38 views
          0 reactions
          Last Post SEQadmin2  
          Working...