Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • Hamid
    Senior Member
    • Sep 2009
    • 108

    #46
    Hi Shanshan,

    With regards to you seeing multiple bands on your westerns, it is quite normal. That is occurring because most antibodies for western, with the exception of a few histone ones, are not monoclonal, so there were be background bands. Even quite a few monoclonal antibodies will recognize more than one epitope.
    Also when working with tissues, you will have higher background due to the presence of antibodies against the animal used to generate the antibodies.
    When we run westerns for epitope integrity, we cut the membrane such that it incorporates the area representing the size of the protein of interest. By doing that you reduce the amount of antibody you are using, but also render the binding to the region of interest more efficient.

    Can you post your post IP bioanalyzer results as well as your western if possible?
    Did you run the post IP DNA on a high sensitivity chip? If so, then the high sensitivity chip is very sensitive to salts, and will always give a higher size range and distribution as compared to the regular sensitivity chip. We are almost done with a technical note that clearly shows that and should have that available for posting by next week.


    Thank you

    hamid
    Last edited by Hamid; 04-27-2012, 12:26 PM.

    Comment

    • shanshanzhou
      Junior Member
      • Mar 2012
      • 9

      #47
      Oh, I also tried to filter out the "crap" using miracloth and shear after cleaning. Got the same result with crude lysate other than reduced amount of DNA recovered...

      Shanshan

      Comment

      • shanshanzhou
        Junior Member
        • Mar 2012
        • 9

        #48
        Good thoughts, wasted so much antibody on westerns

        Comment

        • shanshanzhou
          Junior Member
          • Mar 2012
          • 9

          #49
          Originally posted by Hamid View Post
          Hi Shanshan,

          With regards to you seeing multiple bands on your westerns, it is quite normal. That is occurring because most antibodies for western, with the exception of a few histone ones, are not monoclonal, so there were be background bands. Even quite a few monoclonal antibodies will recognize more than one epitope.
          Also when working with tissues, you will have higher background due to the presence of antibodies against the animal used to generate the antibodies.
          When we run westerns for epitope integrity, we cut the membrane such that it incorporates the area representing the size of the protein of interest. By doing that you reduce the amount of antibody you are using, but also render the binding to the region of interest more efficient.

          Can you post your post IP bioanalyzer results as well as your western if possible?
          Did you run the post IP DNA on a high sensitivity chip? If so, then the high sensitivity chip is very sensitive to salts, and will always give a higher size range and distribution as compared to the regular sensitivity chip. We are almost done with a technical note that clearly shows that and should have that available for posting by next week.


          Thank you

          hamid
          Yep, I used the high sensitivity chip, and I used EB buffer come with the column from Qiagen. Please see attachment for the post IP results, I'll try to take a pic of my western result later.

          I attached another older pre-CHIP bioanalyzer results from several samples with time courses (sorry it's not labeled), even for the 'good' ones, the peak doesn't drop very fast dragging a long tail towards bigger fragment. Is this worrisome to you? Or its just different properties from different cell types.
          Thanks so much!

          Best,
          Shanshan
          Attached Files

          Comment

          • tamara.tjitrowirjo
            Junior Member
            • Apr 2012
            • 8

            #50
            Hi...

            I have ran Western blot with my ChIP and Input (only sheared chromatin without any IP) samples. I did not run reverse cross-linking. It depends on the condition of your Western blot. We use sample buffer containing beta-mercaptoethanol and boil them for 5 minutes. This will denature the protein and release all the DNA that are bound, essentially the same as reverse cross-linking procedure. I then ran them on a denaturing gel and probe with antibody against my protein of interest. It works well.

            I sheared my samples using Branson 250D probe sonicator for 15 cycles (total ON sonication is 7 minutes 30 seconds).
            Attached Files

            Comment

            • pmiguel
              Senior Member
              • Aug 2008
              • 2328

              #51
              Boiling DNA that you want to sequence seems crazy to me. But maybe the BME prevents some of the damage that would otherwise occur?

              --
              Phillip

              Comment

              • tamara.tjitrowirjo
                Junior Member
                • Apr 2012
                • 8

                #52
                Noooooo Phillip... This is for running the Western. I would never boil my DNA that will be sequenced... Those DNA for sequencing undergoes reverse cross-linking at 65C overnight followed by RNase A and Proteinase K treatment and then phenol chloroform isolation and then sample prep.

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #53
                  Ah. Thanks for the clarification.

                  --
                  Phillip

                  Comment

                  • shanshanzhou
                    Junior Member
                    • Mar 2012
                    • 9

                    #54
                    Originally posted by tamara.tjitrowirjo View Post
                    Hi...

                    I have ran Western blot with my ChIP and Input (only sheared chromatin without any IP) samples. I did not run reverse cross-linking. It depends on the condition of your Western blot. We use sample buffer containing beta-mercaptoethanol and boil them for 5 minutes. This will denature the protein and release all the DNA that are bound, essentially the same as reverse cross-linking procedure. I then ran them on a denaturing gel and probe with antibody against my protein of interest. It works well.

                    I sheared my samples using Branson 250D probe sonicator for 15 cycles (total ON sonication is 7 minutes 30 seconds).
                    That's a nice band, I did the same by boiling 5 min and got really messy stain, start thinking maybe its the antibody I use. Trying again with some reverse crosslinked sample today, see if there will be a difference. What power setting do you use on the Branson? We have a digital fisher one that allow you to change Joule and %Amplitude, I can understand %Amplitude, don't know what should I set for Joule...

                    Shanshan

                    Comment

                    • tamara.tjitrowirjo
                      Junior Member
                      • Apr 2012
                      • 8

                      #55
                      hi Shanshan,

                      Thanks for the comment! =) I am very happy with the Western...

                      You need to optimize it based on your cells. But from our experience since they are all nuclei extracted, they differ only slightly. But then again, all our cells are from human breast.

                      Amplitude setting for 50%, we do not have Joule setting so I don't know anything about it. Good luck!

                      Comment

                      • shanshanzhou
                        Junior Member
                        • Mar 2012
                        • 9

                        #56
                        Originally posted by tamara.tjitrowirjo View Post
                        hi Shanshan,

                        Thanks for the comment! =) I am very happy with the Western...

                        You need to optimize it based on your cells. But from our experience since they are all nuclei extracted, they differ only slightly. But then again, all our cells are from human breast.

                        Amplitude setting for 50%, we do not have Joule setting so I don't know anything about it. Good luck!
                        Thanks tamara

                        Comment

                        • ALive
                          Junior Member
                          • Jan 2012
                          • 8

                          #57
                          Hi Everyone,

                          I've done some Western blotting, and I didn't see a decrease in the amount of protein as the sonication time was increased... The conditions I tried were:

                          2 minutes cross-linking -> sonication for 5 minutes; 15 minutes; 20 minutes
                          5 minutes cross-linking -> sonication for 5 minutes; 15 minutes; 20 minutes
                          8 minutes cross-linking -> sonication for 5 minutes; 15 minutes; 20 minutes

                          To prepare the samples, I used sample buffer containing beta-mercaptoethanol and put them >90C for 10 minutes. But I was seeing two bands - the smaller was the correct size, and the larger looked almost double the size so could be a dimer (? this is my first time Western blotting, I don't know a lot about them ). So, I reverse cross-linked the samples and then did the same sample buffer step as above. This seemed to almost get rid of the larger band.

                          I've attached a picture of a western I did today.

                          1: Input - sample buffer
                          2. Input - reverse cross-linked + sample buffer
                          3. 2 minute cross-link, 5 minute sonication - sample buffer
                          4. 2 minute cross-link, 20 minute sonication - sample buffer
                          5. 2 minute cross-link, 20 minute sonication - reverse cross-linked + sample buffer
                          6. 8 minute cross-link, 5 minute sonication - sample buffer
                          7. 8 minute cross-link, 20 minute sonication - sample buffer
                          8. 8 minute cross-link, 20 minute sonication - reverse cross-linked + sample buffer

                          So it looks to me like I'm not destroying the epitope by over-sonicating... Maybe I’m not using enough protease inhibitors in the antibody incubation step or wash step?

                          Thanks in advance for any comments

                          Ali
                          Attached Files

                          Comment

                          • Hamid
                            Senior Member
                            • Sep 2009
                            • 108

                            #58
                            Hi Alive,

                            The western does look good, and your protein is quiet small in size which should render it more resistant to damage during shearing. It does seem like it is since I don't see a major difference between the 5 minute, and 20 minute processing. Your protein seems quiet abundant . One thing to remember is that just because you are seeing the band on the western, it does not not mean that it is still associated or linked to the DNA fragment. The western will still pickup proteins not associated with the chromatin. The only way to find that out will be do do an IP.
                            My suggestion is that you pick the samples that generated the best shearing profile based on the bioanalyzer results, and do an IP on a 5 minute and 20 minutes processed samples. comparing the post-IP bioanalyzer will then give you a good indication of your epitope(protein+chromatin) integrity.

                            Thank you

                            Hamid

                            Comment

                            • ALive
                              Junior Member
                              • Jan 2012
                              • 8

                              #59
                              Thanks Hamid. I'll do that and post my results.

                              Btw, I'm looking at histone modifications and used anti H4K20me1 for the Western.

                              Cheers, Ali

                              Comment

                              • tamara.tjitrowirjo
                                Junior Member
                                • Apr 2012
                                • 8

                                #60
                                Hello people...!!

                                Forgot to report back. I had that problem where I extracted my ChIP sample prep DNA on gel about 200-300bp (with adapters) and run bioanalyzer and caliper and both of them reported that my fragments are about 4-5kb! I went on to sequence them anyways and they are about 150bp (without adapters). Got sequencing results okay, no problem with extra large fragments.

                                Hard evidence that I cannot trust the bioanalyzer and caliper, at least for my case. Good old gel is trustworthy! =D

                                Comment

                                Latest Articles

                                Collapse

                                • SEQadmin2
                                  16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                  by SEQadmin2




                                  NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                  Today, 06:46 AM
                                • SEQadmin2
                                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                  by SEQadmin2



                                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                  ...
                                  10-14-2026, 09:55 AM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by SEQadmin2, 10-28-2026, 09:57 AM
                                0 responses
                                25 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-28-2026, 09:51 AM
                                0 responses
                                15 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-22-2026, 11:13 AM
                                0 responses
                                71 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-22-2026, 10:48 AM
                                0 responses
                                48 views
                                0 reactions
                                Last Post SEQadmin2  
                                Working...