  • kobeho24
    Member
    • Apr 2015
    • 32

    #166
    Originally posted by Simone78 View Post
    Hi Gary,
    I just tried with 0.2% SDS, in the same way as I am doing after tagmention to strip the Tn5 off the DNA before PCR and it didn´t work. I also tried SDS + NaOH/HCl (skipping columns or precipitation, as they recommend in the protocol) but, as I said, there was probably too much salt at that point and the PCR didn´t work.
    /Simone
    Hi Simone,
    I don't know if you noticed that the newly published TGIRT paper in the journal RNA. They use the TGIRT to do the RT of human plasma RNA. And I realized that there is a fragmentation step prior to RT when dealing with whole cell total RNA as they suggested. Thus, on my perspective, defeat its better processivity compared to retroviral RT. You mentioned that the average size of cDNA generated by TGIRT is 4-5k. I just wonder if you have an in-house protocol to do so. It would be much appreciated if you let me know.
    BTW, I noticed that the TGIRT can template switch not only RNA but also DNA, which means it might not be compatible with single-cell RNA-seq, unless we extract the total RNA from a single cell and do rRNA-depletion before RT, am I right?

    Best!
    Gary

    Comment

    • amcg
      Junior Member
      • Nov 2015
      • 1

      #167
      Hi all,
      I'm hoping to use the Smart-seq2 protocol to make rna-seq libraries from small populations (200-500) of cells collected by FACS. In this case would you advise sorting them directly into the lysis buffer (triton-x and RNase inhibitor) described in the Nature Protocols paper, or would it be better to extract the RNA first and then proceed with the reverse transcription on purified RNA?

      Many thanks!

      Comment

      • chaichontat
        Junior Member
        • Sep 2015
        • 3

        #168
        Hi Simone,

        I used the Smart-seq2 protocol using the Superscript IV. For some reason, the reaction failed completely (no cDNA products were observed). When I did not add 1M betaine (Sigma) and 6mM MgCl2, the RT reaction went beautifully, but there's no template switching. The template was 10 ng of purified PBMCs RNA and the TSO was synthesized by Eurogentec. Could you provide me with the reaction conditions you used with SSRT IV?

        Thanks!
        Chaichontat

        Comment

        • Simone78
          Senior Member
          • Oct 2010
          • 208

          #169
          Originally posted by amcg View Post
          Hi all,
          I'm hoping to use the Smart-seq2 protocol to make rna-seq libraries from small populations (200-500) of cells collected by FACS. In this case would you advise sorting them directly into the lysis buffer (triton-x and RNase inhibitor) described in the Nature Protocols paper, or would it be better to extract the RNA first and then proceed with the reverse transcription on purified RNA?

          Many thanks!
          Hi,
          even if your cells have a lot of RNA I wouldn´t do it. This is only my opinion, others might think differently. Every time you transfer RNA before any amplification step you lose molecules. I would sort directly into the lysis buffer, but be aware that 0.2% Triton might not be sufficient for lysing so many cells.
          Best,
          Simone

          Comment

          • Simone78
            Senior Member
            • Oct 2010
            • 208

            #170
            Originally posted by chaichontat View Post
            Hi Simone,

            I used the Smart-seq2 protocol using the Superscript IV. For some reason, the reaction failed completely (no cDNA products were observed). When I did not add 1M betaine (Sigma) and 6mM MgCl2, the RT reaction went beautifully, but there's no template switching. The template was 10 ng of purified PBMCs RNA and the TSO was synthesized by Eurogentec. Could you provide me with the reaction conditions you used with SSRT IV?

            Thanks!
            Chaichontat
            Hi,
            I did exactly as described in the Nat Prot paper, just replacing SSRTII with SSRTIV and doing the RT for 15 min @ 50 degrees and inactivated the enzyme for 10 min @ 80 degrees.. I tested it only on tot RNA (10 pg) and I actually noticed that when I didn´t add betaine and MgCl2 I got worse results than when using them...the opposite of what you found. in my case the template switching went ok, at least based on the cDNA yield and size.
            Best,
            Simone

            Comment

            • Simone78
              Senior Member
              • Oct 2010
              • 208

              #171
              Originally posted by kobeho24 View Post
              Hi Simone,
              I don't know if you noticed that the newly published TGIRT paper in the journal RNA. They use the TGIRT to do the RT of human plasma RNA. And I realized that there is a fragmentation step prior to RT when dealing with whole cell total RNA as they suggested. Thus, on my perspective, defeat its better processivity compared to retroviral RT. You mentioned that the average size of cDNA generated by TGIRT is 4-5k. I just wonder if you have an in-house protocol to do so. It would be much appreciated if you let me know.
              BTW, I noticed that the TGIRT can template switch not only RNA but also DNA, which means it might not be compatible with single-cell RNA-seq, unless we extract the total RNA from a single cell and do rRNA-depletion before RT, am I right?

              Best!
              Gary
              Hi Gary,
              yes, I saw the paper, thanks! I didn´t know that the TGIRT can do strand switch on DNA as well (this was not mentioned when I contacted the Lambowitz lab). When I first did my tests some time ago I used tot RNA. Unfortunately, it didn´t work at single cell level (pg) and I just dropped it. therefore, I never sequenced anything and never tried on a real cell. I do have a protocol that worked with 100 ng (!) RNA but after the publication of this new article, I want to try again using some modifications and see if I can get it to work with less RNA. So I can´t really share it right now, sorry!
              Best,
              Simone

              Comment

              • chaichontat
                Junior Member
                • Sep 2015
                • 3

                #172
                Originally posted by Simone78 View Post
                Hi,
                I did exactly as described in the Nat Prot paper, just replacing SSRTII with SSRTIV and doing the RT for 15 min @ 50 degrees and inactivated the enzyme for 10 min @ 80 degrees.. I tested it only on tot RNA (10 pg) and I actually noticed that when I didn´t add betaine and MgCl2 I got worse results than when using them...the opposite of what you found. in my case the template switching went ok, at least based on the cDNA yield and size.
                Best,
                Simone
                Hi Simone,

                It turned out that my PCR primers were the problem. I was too fixated on the template switching process that I completely ignored the preamplification. Upon replacing the primers, everything works beautifully. I really thank you for this brilliant protocol!

                Chaichontat

                Comment

                • KroSeq
                  Junior Member
                  • Dec 2014
                  • 4

                  #173
                  SSRT IV vs SSRT II

                  Originally posted by Simone78 View Post
                  Hi,
                  I did exactly as described in the Nat Prot paper, just replacing SSRTII with SSRTIV and doing the RT for 15 min @ 50 degrees and inactivated the enzyme for 10 min @ 80 degrees.. I tested it only on tot RNA (10 pg) and I actually noticed that when I didn´t add betaine and MgCl2 I got worse results than when using them...the opposite of what you found. in my case the template switching went ok, at least based on the cDNA yield and size.
                  Best,
                  Simone
                  Hej Simone,
                  what is the benefit of using SSRT IV instead of SSRT II?

                  Comment

                  • Simone78
                    Senior Member
                    • Oct 2010
                    • 208

                    #174
                    Originally posted by KroSeq View Post
                    Hej Simone,
                    what is the benefit of using SSRT IV instead of SSRT II?
                    the RT reaction is way faster than with the SSRTII but, apart from that, I don´t see any improvement in the reaction (at least with my settings!). The prices are almost the same, the SSRTIV is a bit more expensive.

                    Comment

                    • longwood
                      Junior Member
                      • May 2014
                      • 5

                      #175
                      Superscript II contamination update

                      E. Coli contamination of superscript II has been confirmed by the manufacturer. I know a bunch of people reported this on this forum a few months back, so thought I'd share this letter (attached) I received from Thermo Fisher. The letter basically says they have started testing for E. Coli DNA as part of their routine test now and those who received the tainted lots will be sent free replacement upon request.
                      Attached Files

                      Comment

                      • kobeho24
                        Member
                        • Apr 2015
                        • 32

                        #176
                        Originally posted by Simone78 View Post
                        Hi Gary,
                        yes, I saw the paper, thanks! I didn´t know that the TGIRT can do strand switch on DNA as well (this was not mentioned when I contacted the Lambowitz lab). When I first did my tests some time ago I used tot RNA. Unfortunately, it didn´t work at single cell level (pg) and I just dropped it. therefore, I never sequenced anything and never tried on a real cell. I do have a protocol that worked with 100 ng (!) RNA but after the publication of this new article, I want to try again using some modifications and see if I can get it to work with less RNA. So I can´t really share it right now, sorry!
                        Best,
                        Simone
                        Hi Simone,
                        Sounds really great that you do have a protocol. What I wonder is RT condition for non-fragmented RNA. Just incubate the RNA with TGIRT at proper temperature for 15 min? Looking forward to your successful modified protocol.

                        Gary

                        Comment

                        • chaichontat
                          Junior Member
                          • Sep 2015
                          • 3

                          #177
                          Hi Simone,

                          I'm trying to amplify one gene of interest; however, I couldn't get a specific band when I added the whole 10 ul of the RT reaction (final PCR volume was 25 ul). The band only appeared when I added 1 ul of the RT mix, suggesting that there's some inhibition of PCR by the RT mix.

                          However, there was one deviation from your protocol: I used the component version (not ReadyMix) of the KAPA HiFi Hot Start. Do you think that this could have been a factor?

                          Thanks!
                          Chaichontat

                          Comment

                          • kobeho24
                            Member
                            • Apr 2015
                            • 32

                            #178
                            SSII contamination

                            Hi all,
                            Recently I’ve done 3 runs of single-cell RNA-seq (Smart seq2) trial. I just followed the original protocol. A highly reproducible background curve can be observed even in negative controls (only nuclease free water) when using our new SuperScript II RTase, which cannot be seen when using SSIII, or without RTase. Our bad SSII lot no. is 1688365.
                            Click image for larger version

Name:	QQ??20160121185909.png
Views:	1
Size:	25.3 KB
ID:	305041

                            Click image for larger version

Name:	QQ??20160121190101.png
Views:	1
Size:	53.2 KB
ID:	305042

                            Click image for larger version

Name:	QQ??20160121190137.png
Views:	1
Size:	24.6 KB
ID:	305043

                            Does any of you guys have any recommendation of substitutive MMLV RTase? I saw ProtoScripts II & SS IV so far, any candidate else?

                            Best wishes!
                            Gary
                            Last edited by kobeho24; 01-21-2016, 06:47 PM.

                            Comment

                            • Simone78
                              Senior Member
                              • Oct 2010
                              • 208

                              #179
                              Originally posted by kobeho24 View Post
                              Hi all,
                              Recently I’ve done 3 runs of single-cell RNA-seq (Smart seq2) trial. I just followed the original protocol. A highly reproducible background curve can be observed even in negative controls (only nuclease free water) when using our new SuperScript II RTase, which cannot be seen when using SSIII, or without RTase. Our bad SSII lot no. is 1688365.
                              [ATTACH]4169[/ATTACH]

                              [ATTACH]4170[/ATTACH]

                              [ATTACH]4171[/ATTACH]

                              Does any of you guys have any recommendation of substitutive MMLV RTase? I saw ProtoScripts II & SS IV so far, any candidate else?

                              Best wishes!
                              Gary
                              I would use either SMARTscribe (Clontech) or Maxima H- (Thermo Scientific).

                              Comment

                              • kobeho24
                                Member
                                • Apr 2015
                                • 32

                                #180
                                Originally posted by Simone78 View Post
                                I would use either SMARTscribe (Clontech) or Maxima H- (Thermo Scientific).
                                Hi Simone,
                                Between these two, which one's better? It seems that SMARTscribe is similar to SSII and Maxima H- is similar to SSIV. Besides, as SMARTscribe is 100U/ul while Maxima H- is 200U/ul, do we still make it 100U in the 10ul RT reaction as it is in the Smart-seq2 protocol? Am I right?

                                Best!
                                Gary

                                Comment

                                Latest Articles

                                Collapse

                                • SEQadmin2
                                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                  by SEQadmin2



                                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                  ...
                                  10-14-2026, 09:55 AM
                                • SEQadmin2
                                  New Genomics Technologies Take Aim at Long-Standing Limits
                                  by SEQadmin2


                                  Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                  We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                  ...
                                  09-28-2026, 10:25 AM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by SEQadmin2, 10-15-2026, 10:52 AM
                                0 responses
                                26 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-14-2026, 05:39 AM
                                0 responses
                                32 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-13-2026, 09:56 AM
                                0 responses
                                31 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-09-2026, 11:14 AM
                                0 responses
                                28 views
                                0 reactions
                                Last Post SEQadmin2  
                                Working...