    #181
    Hi Simone,

    With reference to your trials of 2ul tagmentation, did you have any luck reducing the size of your libraries? Even with 250pg input, I find that my final libraries are average 600bp on the BA, which makes our sequencing core apprehensive about loading them onto the HiSeq. I guess the limiting factor is loaded enzyme - did you ever try "re-loading" the commercial product with adapters?

    Thanks in advance.

    Comment

    • adamreid
      Junior Member
      • Nov 2009
      • 6

      #182
      Problems with low input single-cell RNA-seq

      We are trying to sequence RNA from individual small cells.

      We are sorting by FACS and using SmartSeq2 and Nextera XT to make libraries.

      We have had several problems.

      We get rRNA contamination from both E. coli (perhaps due to contaminated reverse transcriptase) and from our target organism. This is strange because SmartSeq2 should only amplify polyA mRNAs.

      Secondly when we sequence the libraries we get adaptor contamination, sometimes from the strand-switching oligo and also from the Nextera transposase.

      Has anyone noticed similar problems?

      Cheers,

      Adam

      Comment

      • Simone78
        Senior Member
        • Oct 2010
        • 208

        #183
        Originally posted by daniel007 View Post
        Hi Simone,

        With reference to your trials of 2ul tagmentation, did you have any luck reducing the size of your libraries? Even with 250pg input, I find that my final libraries are average 600bp on the BA, which makes our sequencing core apprehensive about loading them onto the HiSeq. I guess the limiting factor is loaded enzyme - did you ever try "re-loading" the commercial product with adapters?

        Thanks in advance.
        Hi,
        we routinely perform tagmentation in 2 ul volume and my input is generally 100 pg (and used 12 cycles PCR). Even then, sometimes I get libraries that are a bit too long, so I am thinking to reduce it to 50 pg (and 14 cycles PCR post-tagm).
        I never tried to re-load the commercial Tn5. I am not sure it´s possible or, at least, I never thought about it!
        Best,
        Simone

        Comment

        • Simone78
          Senior Member
          • Oct 2010
          • 208

          #184
          Originally posted by kobeho24 View Post
          Hi Simone,
          Between these two, which one's better? It seems that SMARTscribe is similar to SSII and Maxima H- is similar to SSIV. Besides, as SMARTscribe is 100U/ul while Maxima H- is 200U/ul, do we still make it 100U in the 10ul RT reaction as it is in the Smart-seq2 protocol? Am I right?

          Best!
          Gary
          sorry for the very late reply. If you are still interested, I can say that Maxima H- works well and now I would use that. I recently tried again with SMARTscribe but I didn´t get great results and I don´t really know the reason since it used to work quite well.
          Best,
          Simone

          Comment

          • immpdaf
            Junior Member
            • Sep 2015
            • 4

            #185
            Hi everyone,
            I would be very grateful if anyone could give me some suggestions in our single-cell RNA seq data analysis part.

            we have 2 groups of single cells (one normal single cells and one disease single cells), we performed single-cell RNA sequencing. Our library is made using SMART-SEQ2 protocol and it is single-end. We have around 4 million reads / single cell.

            Now, using Differential gene Expression analysis, we are going to find significant genes which are upregulated or downregulated in disease cells group with regards to normal group.
            So, which normalization technique could you recommend? Our bioinformatician uses TMM to normalize raw counts and he applies R package Monocle to perform DE.
            He believes that if we use RPKM, we will get many false positive genes, since we are not comparing genes in one sample, but we are comparing different samples. Do you think it is right?

            Many thanks in advance.

            Comment

            • skwek1
              Member
              • Feb 2016
              • 11

              #186
              Biotin-TSO-LNA?

              [QUOTE=jwfoley:

              Also, you mention using iso nucleotides as your 5' blocker. Have you done the test using biotin instead? That seems more common in the literature, and is a lot cheaper (plus it's difficult to get LNAs and iso bases from the same company). In my experience it works just fine to eliminate the "hedgehog".[/QUOTE]

              Hi jwfoley,
              Did you use the same TSO primer sequence published by Simone with LNA but biotinylate the nucleotides in the front? Can you please send me the sequence and indicate which nucleotides are biotinylated? Did you order this from Exiqon?
              Has anyone compared iso-TSO vs biotinylated-TSO vs iso-TSO-LNA vs biotinylated-TSO-LNA?
              Thanks!
              Serena

              Comment

              • skwek1
                Member
                • Feb 2016
                • 11

                #187
                Hi Simone,
                I will like to sequence single immune cells and there are concatamers in the H2O control. Can you please give me the sequence for ISO-TSO that you use to make your library? Does it still have LNA or just all rG? Also where do you order the iso-primers from?
                Thanks,
                Serena

                Comment

                • seqal
                  Junior Member
                  • Mar 2016
                  • 1

                  #188
                  Hi Simone (or anyone else),

                  Can you explain to me the benefit of having the same ISPCR sequence on the polyT and TSO? Does it help with the amplification to have one primer instead of two? Would a single strand with the ISPCR sequence at the 5' end and the reverse complement of ISPCR at the 3' end not loop around and bind to each other, blocking both ends?

                  Thanks!

                  Comment

                  • kobeho24
                    Member
                    • Apr 2015
                    • 32

                    #189
                    Originally posted by seqal View Post
                    Hi Simone (or anyone else),

                    Can you explain to me the benefit of having the same ISPCR sequence on the polyT and TSO? Does it help with the amplification to have one primer instead of two? Would a single strand with the ISPCR sequence at the 5' end and the reverse complement of ISPCR at the 3' end not loop around and bind to each other, blocking both ends?

                    Thanks!
                    Hi seqal,
                    For you first question, I read some literature (Chenchik et al. 1998)about template switching, which is the main mechanism of SMART-seq2. It suggests that reverse transcriptase can search for potential acceptor template sites when cDNA synthesis is interrupted. To efficiently catalyze template switching, the acceptor template should have a homologous region with the 3' end of cDNA.
                    I think it is due to some intrinsic properties of RTase in the course of retroviral replication. That's why we make part of both sides' sequence identical.
                    I might be wrong, please correct my misunderstanding if so

                    Best!
                    Gary

                    Comment

                    • jwfoley
                      Senior Member
                      • Jun 2009
                      • 183

                      #190
                      Originally posted by skwek1 View Post
                      Hi jwfoley,
                      Did you use the same TSO primer sequence published by Simone with LNA but biotinylate the nucleotides in the front? Can you please send me the sequence and indicate which nucleotides are biotinylated? Did you order this from Exiqon?
                      No, my design was somewhat different. I remain skeptical of LNAs as discussed previously in this thread. As also discussed previously, Exiqon is the only vendor of LNAs in North America.

                      But if you biotinylate the oligo, just stick it on the 5' end. IDT's code for this is /5Biosg/.

                      Comment

                      • SunPenguin
                        Member
                        • Aug 2015
                        • 38

                        #191
                        Hey guys,

                        In doing the smart-seq 2 protocol for single cell RNA seq, my colleagues and I find that sometimes we see very uneven preamplification. There are often these "twin peaks" at around 1200bp and 1700bp. I don't see this problem if we repeat the same protocol, but with regular TSO (without LNA). I know in the beginning of this thread, someone mentioned something similar, but it seems like people don't know what it is. Can anyone comment further on the phenomenon?

                        For bulk samples, LNA works well for us, and is our default. For single cells, it seems rather different.

                        PS these are immune cells, if that's relevant.
                        Last edited by SunPenguin; 03-23-2016, 09:12 AM.

                        Comment

                        • skwek1
                          Member
                          • Feb 2016
                          • 11

                          #192
                          Originally posted by jwfoley View Post
                          No, my design was somewhat different. I remain skeptical of LNAs as discussed previously in this thread. As also discussed previously, Exiqon is the only vendor of LNAs in North America.

                          But if you biotinylate the oligo, just stick it on the 5' end. IDT's code for this is /5Biosg/.
                          Thank you for your response jwfoley! Can you lease share the sequences of your TSO, oligodT and IS primers? Should I biotinylate all?

                          Please email me so I can ask you more questions. I have problems amplifying single cell using the Smartseq2 protocol, though 10 cells and 100 cells worked fine. I am using biotinylated TSO(LNA), biotinylated oligodt and biotinylated IS primers. I'm getting quite desperate and would appreciate your help!

                          Thanks,
                          Serena
                          [email protected]
                          Last edited by skwek1; 03-23-2016, 03:33 PM.

                          Comment

                          • SunPenguin
                            Member
                            • Aug 2015
                            • 38

                            #193
                            Hey Simone,

                            I recently saw a paper, where they put at 3-carbon spacer on the 3' end of the TSO for template switching RT. Supposedly it's to reduce "primer founded amplification." I have never seen this before. Do you have any opinion on that?

                            Comment

                            • skwek1
                              Member
                              • Feb 2016
                              • 11

                              #194
                              Originally posted by SunPenguin View Post
                              Hey guys,

                              In doing the smart-seq 2 protocol for single cell RNA seq, my colleagues and I find that sometimes we see very uneven preamplification. There are often these "twin peaks" at around 1200bp and 1700bp. I don't see this problem if we repeat the same protocol, but with regular TSO (without LNA). I know in the beginning of this thread, someone mentioned something similar, but it seems like people don't know what it is. Can anyone comment further on the phenomenon?

                              For bulk samples, LNA works well for us, and is our default. For single cells, it seems rather different.

                              PS these are immune cells, if that's relevant.
                              Hi Sunpenguin,
                              Can you please give me the sequence for your TSO (without LNA) for SMartseq2? Where do you order it from? I am doing single cell RNAseq for immune cells and am having more primers dimers than cDNA for single cell even when I have biotinlylated all the primers. For bulk immune cells, 10 and above, it is fine.
                              Thanks!
                              Serena

                              Comment

                              • skwek1
                                Member
                                • Feb 2016
                                • 11

                                #195
                                Originally posted by SunPenguin View Post
                                Hey Simone,

                                I recently saw a paper, where they put at 3-carbon spacer on the 3' end of the TSO for template switching RT. Supposedly it's to reduce "primer founded amplification." I have never seen this before. Do you have any opinion on that?
                                Hi,
                                Can you please give the reference?
                                Thanks!

                                Comment

                                Latest Articles

                                Collapse

                                • SEQadmin2
                                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                  by SEQadmin2



                                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                  ...
                                  10-14-2026, 09:55 AM
                                • SEQadmin2
                                  New Genomics Technologies Take Aim at Long-Standing Limits
                                  by SEQadmin2


                                  Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                  We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                  ...
                                  09-28-2026, 10:25 AM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by SEQadmin2, 10-22-2026, 11:13 AM
                                0 responses
                                27 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-22-2026, 10:48 AM
                                0 responses
                                20 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-15-2026, 10:52 AM
                                0 responses
                                36 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-14-2026, 05:39 AM
                                0 responses
                                43 views
                                0 reactions
                                Last Post SEQadmin2  
                                Working...