    shearing question using Covaris

    Would like to understand something when I use Covaris for shearing.

    I have DNA that has been diluted to 5ug in 120ul for shearing or 41ng/ul. I shear and measure concentration by Picogreen. But it is about 30ng/ul after shearing. Why does the concentration drop? It is only sheared DNA, not cleaned. Is it due to the DNA being less compact after shearing? Does it have anything to do that DNA is fragmented now and this causes the concentration to drop? I have tried other DNAs and their concentrations drop just like this.
  • ECO
    --Site Admin--
    • Oct 2007
    • 1360

    #2
    Single stranded regions. (?)

    Comment

    • LMcSeq
      Member
      • Feb 2009
      • 33

      #3
      Agreed-PicoGreen is intended to measure double stranded DNA. If the shearing resulted in any single stranded material, that would not be detected.

      Comment

      • Hamid
        Senior Member
        • Sep 2009
        • 108

        #4
        Hi seqgirl123,

        What you are noticing is how pico green method is used for concentration determination. Two points of interest:

        1. Pico green binding to DNA and its intensity is dependent on the size of DNA. It is recommended that you generate a standard curve using DNA in in the same size range as the distribution of your fragments after shearing. If you use a standard curve that has been generated using large DNA such as lambda, the concentration from that standard curve will not apply accurately to the smaller sheared material.
        2. Quite a few service labs use a combination of pico green and A260/A280 to check the quality and concentration of their samples. If you test the samples you mentioned on a nano drop or equivalent system, you will notice that there is very very minimal loss of your samples.

        Thank you

        Hamid

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
          by SEQadmin2





          NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification,...
          10-30-2026, 06:46 AM
        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, Yesterday, 07:59 AM
        0 responses
        13 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 11-05-2026, 09:25 AM
        0 responses
        32 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 11-03-2026, 11:54 AM
        0 responses
        45 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-28-2026, 09:57 AM
        0 responses
        170 views
        0 reactions
        Last Post SEQadmin2  
        Working...