    Truseq RNA Prep Kit Questions

    Hi, hoping to get a few answers/see what others are getting from the Truseq RNA sample kit.

    I have just completed several trial runs, from my results using Bioanalyser to qualify success the libraries seem to be ~1/4 of what is shown in the handbook, thankfully they are ~260bp. So I am confident the protocol has worked, but am now wondering what to do to increase yield.

    Does anyone use any modifications to the standard procedure? I have been thinking:
    (i) increase in PCR enrichment cycles; anecdotally the PCR enrichment step does not suffer from PCR duplicates: is this really the case?
    (ii) use low-bind plates; I have used what we had in the lab but should probably find some low-bind plates: has anyone found this to make a difference in yield? Any recommendations for plates?
    (iii) is it because my RNA is not good quality? RINs ~8 which seems to be acceptable. Was going to try with some very high quality RNA but wanted to know others experience. Surely if it is fragmented during lib. prep. it will not matter so much about the RNA integrity if it isn't badly degraded(?)

    I will sequence and run qPCR but would like advice about issues of yield. First time doing libraries and the lab I am in has a DIY which has worked very well so not sure whether to go along with Truseq or just do the longer (but working) lab method.

    Thanks in advance,

    Bruce.
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    Hi Bruce,
    (i) If you need more library to fill out your run -- getting that extra library by increasing enrichment cycles probably will result in additional PCR duplicates. Stands to reason...
    (ii) Depends on the binding capacity of normal plates. I don't know what that is. Maybe 1 ng?
    (iii) Yes. The TruSeq RNA kit entails a polyA RNA binding. If your RNA is degraded, then only that RNA adjacent to the polyA tail will be retained. The 5' ends of the RNA strands will be lost. That, among other things, will lower yield.

    --
    Phillip

    Comment

    • bruce01
      Senior Member
      • Mar 2011
      • 160

      #3
      Thanks for your reply Phillip,

      yes, it does stand to reason that PCR dupes are created with additional cycles, for our DIY method we have reduced to 7 or 8 cycles as this is a big problem. I have heard from technicians that dupes are reduced in the Illumina kit but have not seen any proof of this (yet!)

      I am going to try a sample with very high RIN and 28/16 ratio to determine if degradation plays a major part in my lower yield. I can see how degradation at 5' would affect yield, just hoping the method isn't very sensitive to this.

      Illumina techsupport have suggested a vacuum concentrator or preparing more library and pooling, but I am not too keen as we are cutting it fine with number of preps in the kit as it is.

      Bruce.

      Comment

      • bruce01
        Senior Member
        • Mar 2011
        • 160

        #4
        Hi Phillip,

        thanks for the reply. Was waiting to test a few more samples prior to replying myself.

        (i) Illumina techsupport suggest: making another library to add to the previous one to boost yield; use a vacuum concentrator (without heat). Agree with increasing PCR cycles, do not want to bias more than we already do.

        (ii+iii) constructed a very nice library (akin to the image in the handbook) using a very good quality RNA sample (RIN 9.5, 28/18 2.1). Also left XP steps for longer to increase binding which may have helped.

        There is a 'bump' after the ~260bp library, ~550-2000bp which I assume is insert/adapter concatenators(?)

        Cheers,

        Bruce.

        Comment

        Latest Articles

        Collapse

        • GATTACAT
          Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
          by GATTACAT
          Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
          07-01-2026, 11:43 AM
        • SEQadmin2
          Nine Things a Sample Prep Scientist Thinks About Before Sequencing
          by SEQadmin2


          I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

          Here are nine questions we think about, in roughly the order they matter, before...
          06-18-2026, 07:11 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 07-02-2026, 11:08 AM
        0 responses
        22 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-30-2026, 05:37 AM
        0 responses
        23 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-26-2026, 11:10 AM
        0 responses
        22 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-17-2026, 06:09 AM
        0 responses
        55 views
        0 reactions
        Last Post SEQadmin2  
        Working...