Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • litc
    Member
    • Oct 2010
    • 24

    #1

    Why RAD reads depth vary extremely?

    I have got a RAD data set and assessed it. It seems the depth of RAD fragment vary differently, extremely different.

    the reads(extract same reads) with depth 1 occupied 40%
    but there are also others different reads have the depth from 2~100000.

    I cannot image that a single reads have 100000 fold, is it a repeat in the genome? but are there so many repeat in genome?
  • DZhang
    Senior Member
    • Jun 2010
    • 177

    #2
    Hi Litc,

    Yes, RAD fragments have extremely high coverage. It is kind of expected. Please check RAD -related publications to understand the methodology.

    Best regards,
    Douglas

    Comment

    • johnomics
      Junior Member
      • Sep 2009
      • 5

      #3
      Yes, RAD data varies greatly in read depth, above and beyond what is seen in genomic libraries. There are two main causes.

      Firstly, the high coverage is usually because duplicates and repeats do create very large stacks. Because RAD loci are anchored to particular positions, and the loci cover many fewer positions than is typical for a genomic DNA library, these depths can be much higher than for genomic libraries, where coverage is smeared over many consecutive positions.

      Secondly, the low coverage is because there is a strong correlation between restriction fragment length and read depth, for restriction fragments below 10 Kb in size. The shorter the restriction fragment, the less coverage. We believe this is due to incomplete shearing. Short fragments are not sheared as efficiently as long ones, and so many fragments in the range 1 Kb - 10 Kb are not cut and are filtered out when you size select Illumina-length fragments (300-700bp or so). This makes it difficult to separate real RAD loci from sequencing error at the low end.

      We have a paper in revision with Molecular Ecology right now discussing this issue; I hope it will be out within the next few months.

      John Davey

      Comment

      • DZhang
        Senior Member
        • Jun 2010
        • 177

        #4
        Hi John,

        Thank you for elaborating on this. Please do inform us when your paper is out.

        Best regards,
        Douglas

        Comment

        • litc
          Member
          • Oct 2010
          • 24

          #5
          Thank DZhang, and also thank johnomics's explanation. johnomics's explanations are good, but I also doubt that repeat fragment in genome is the only reason for the very high depth of some RAD. some RAD reads are very over-represented, they have a depth of 100000, is it means that RAD_A(suppose RAD_A have a depth of 100000) are 20000 times more repeat than RAD_B(have a depth of 5)? Are there something wrong in amply process or processes in RAD library preparation for some fragment are likely to be amplied while other fragments not.

          Comment

          • johnomics
            Junior Member
            • Sep 2009
            • 5

            #6
            There could be amplification bias; RAD suffers from the same PCR bias that normal genomic libraries suffer from (see Mike Quail & Sanger Institute papers for more on this), so will tend to amplify GC rich regions more often than AT rich. But which genome are you working with? It's quite possible for a repeat to occur tens of thousands of times; there are hundreds of thousands of LINEs and SINEs in the human genome, for example.

            Comment

            • litc
              Member
              • Oct 2010
              • 24

              #7
              Thank johnomics, I'm working with peanut, It may be the bias and repeat the influence the depth of RAD.

              Comment

              • johnomics
                Junior Member
                • Sep 2009
                • 5

                #8
                Molecular Ecology paper mentioned above now out in Early View, open access:

                Davey JW, Cezard T, Fuentes-Utrilla P, Eland C, Gharbi K, Blaxter ML (2012). Special features of RAD Sequencing data: implications for genotyping. Molecular Ecology, Genotyping by Sequencing special issue, Early View, doi:10.1111/mec.12084

                http://onlinelibrary.wiley.com/doi/1...mec.12084/full

                Comment

                • JackieBadger
                  Senior Member
                  • Mar 2009
                  • 385

                  #9
                  does double digest RAD not remedy the uneven coverage issue?

                  Comment

                  • johnomics
                    Junior Member
                    • Sep 2009
                    • 5

                    #10
                    It should do, but we haven't done any direct tests. Would be very happy to hear from others with ddRAD data.

                    Comment

                    • JackieBadger
                      Senior Member
                      • Mar 2009
                      • 385

                      #11
                      We should have a data set within a couple of weeks

                      Comment

                      • Hanne
                        Junior Member
                        • Dec 2009
                        • 6

                        #12
                        Hi all,

                        We are inspired by Peterson et al. paper regarding ddRAD and are going to try out their protocol. Does anyone have experience with this? I guess you have JackieBadger? Do you have any data yet? How many samples did you pool? And did you use the HiSeq or MiSeq?

                        Best,
                        Hanne

                        Comment

                        • JackieBadger
                          Senior Member
                          • Mar 2009
                          • 385

                          #13
                          Originally posted by Hanne View Post
                          Hi all,

                          We are inspired by Peterson et al. paper regarding ddRAD and are going to try out their protocol. Does anyone have experience with this? I guess you have JackieBadger? Do you have any data yet? How many samples did you pool? And did you use the HiSeq or MiSeq?

                          Best,
                          Hanne
                          We have done a few runs of ddRAD, but still fine-tuning.
                          Number of samples depends on your genome size/number of expected RE cuts/required depth. This is something you have to really play with/estimate on a case by case basis. We are using the MiSeq, in the last run we pooled ~ 40 individuals with human sized genomes in one run ...equivalent to one illumina lane.... I think this is too many samples.
                          When I have more concrete info I will update

                          Comment

                          • Hanne
                            Junior Member
                            • Dec 2009
                            • 6

                            #14
                            Originally posted by JackieBadger View Post
                            We have done a few runs of ddRAD, but still fine-tuning.
                            Number of samples depends on your genome size/number of expected RE cuts/required depth. This is something you have to really play with/estimate on a case by case basis. We are using the MiSeq, in the last run we pooled ~ 40 individuals with human sized genomes in one run ...equivalent to one illumina lane.... I think this is too many samples.
                            When I have more concrete info I will update
                            Thanks

                            I have a lot of questions regarding the protocol;
                            1. What software did you use to perform in silico digestions?
                            2. In the ligation molarity calculator, row 3. It looks like they only use 50 ng (0,05 ug) DNA in the reaction, is this correct?
                            3. I’m confused by the procedure described in Dynabeads® M-270 Streptadivin and talked to the producer (DYNAL). They informed me to use at least 100 ul of the beads, how much did you use?
                            4. They recommend the step 2.1.4 “Immobilization of Nucleic Acids”, with 3 washes and resuspension of beads in elution buffer. 100 ul (?) in TE buffer? Or water?
                            5. I have learned that the Streptavidin-Coupled Dynabeads® can be used directly in a PCR reaction, is this what you did?
                            6. DYNAL also told me that it is possible to Release Immobilized Biotinylated Molecules; Separation of two DNA strands can be done by either alkali or temperature treatment. Using alkali, you elute off the non-biotinylated strand with 0.1M NaOH.” Have you tried this?
                            7. Can you please share the PCR program you used? The PCR primers have annealing temperature at 72 °C…
                            8. In your protocol it says; For each Pippin Prep elution or gel extraction, set up 4-8 PCR reactions in 20ul total volume: For each PCR, combine ~20ng of size-selected sample. Do you measure this, or just split the sample, with or without beads, into 4-8 reactions?
                            9. The Norwegian Sequencing Center (http://www.sequencing.uio.no/) informed us that monochromatic motives (ie AATTC entire flow cell) can be difficult to handle for the MiSeq., so if this is the case, there will have to be adjusted slightly in the run parameters, so the failure to detect this motif (AATTC). However, get round this by blending the library with PhIX to 50%, and lowering the clustering density to 50% of normal. The result of course is that we only get 25% of the possible output of the MiSeq… What did you do?

                            Sorry for all my questions...

                            Comment

                            • JackieBadger
                              Senior Member
                              • Mar 2009
                              • 385

                              #15
                              Sorry I can't answer any of your questions. Our lab tech does all of the library prep..I sit at a computer all day
                              In term of MiSeq having issues with low complexity libraries. This shouldn't be a problem for RAD work because low complexity isnt an issue. Just make sure you have a wise choice of barcodes so that cluster generation is not hindered.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM
                              • SEQadmin2
                                New Genomics Technologies Take Aim at Long-Standing Limits
                                by SEQadmin2


                                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                ...
                                09-28-2026, 10:25 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              24 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              20 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-15-2026, 10:52 AM
                              0 responses
                              36 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-14-2026, 05:39 AM
                              0 responses
                              42 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...