    cDNA synthesis kits with titanium

    I want to sequence ESTs from a RNA pool. I heard good things about SMART cDNA synthesis kits from Clotech and Evrogen. Also, I read different papers that have used this technology. My concern it is that I also read from "cDNA Frequently Asked Questions" from Roche, that SMART kits are not recommended with titanium kits because the size of the cDNA fragments obtained, ussually small in size and fail to nebulize properly. Any suggestions?
  • kmcarr
    Senior Member
    • May 2008
    • 1181

    #2
    Could you post a link to that FAQ. I searched the web site but could not find it. I'm interested in what they mean by the cDNA fragments are usually small in size.

    We have run many, many cDNA projects on the 454 using GS 20, GS-FLX and now Titanium. The vast majority of these used SMART cDNA preparation, some with the Evrogen normalization, some without. We have had generally good results. Compared to genomic DNA sequencing cDNA sequencing generally produces fewer filter passed reads (more failures due to mixed reads) and the mean length is shorter. There is also more variability from library to library which probably is a DNA quality issue. I'd be happy to provide more information if you like.

    Comment

    • BeFuPe
      Junior Member
      • Mar 2009
      • 3

      #3
      Thank you very much for your quick response. I would appreciate as much information as you can provide me since we are new in this business. The FAQ document it was provided from Roche Support Department by email, it is not public in the website but I can forward it to you if you wish.

      Comment

      • jOn
        Member
        • Oct 2008
        • 19

        #4
        For the reasons already pointed out by kmcarr

        "Compared to genomic DNA sequencing cDNA sequencing generally produces fewer filter passed reads (more failures due to mixed reads) and the mean length is shorter. There is also more variability from library to library which probably is a DNA quality issue."

        I would advise doing a comprehensive titration for each cDNA sample processed. Also if you're going ahead with normalization you'll notice a deffinate shift in fragment sizes when run on a gel.

        Comment

        • gerald2545
          Member
          • Nov 2008
          • 22

          #5
          We have run many, many cDNA projects on the 454 using GS 20, GS-FLX and now Titanium. The vast majority of these used SMART cDNA preparation, some with the Evrogen normalization, some without. We have had generally good results. Compared to genomic DNA sequencing cDNA sequencing generally produces fewer filter passed reads (more failures due to mixed reads) and the mean length is shorter. There is also more variability from library to library which probably is a DNA quality issue. I'd be happy to provide more information if you like.
          Hi,
          we run only few project on our 454 (titanium only) and we were surprised about the amount of filtered reads and the length of the high quality reads. could you please tell me (if possible) the mean length and the percentage of passed filter?

          For example, for our last run (4 regions, cDNA from chicken and duck), for 2 regions (chicken) we only have ~15% of passed filters (whereas Roche expects more than 50% for a titanium run - information from GSFLX Titanium metrics, but it may be for a genomic DNA project). Problems came from dots (cf attached spreadsheet).

          Gerald
          Attached Files

          Comment

          • jOn
            Member
            • Oct 2008
            • 19

            #6
            Hi Gerald

            Are you doing a size exclusion with the cDNA and a titration (concentrate the titration at the lower end)? Also, because fragment size is usually shorter for cDNA it may be worth loading the beads on the PTP at a lower density. As I've said before don't expect cDNA runs to be as successful as runs using gDNA.

            Comment

            • kmcarr
              Senior Member
              • May 2008
              • 1181

              #7
              Originally posted by gerald2545 View Post
              Hi,
              we run only few project on our 454 (titanium only) and we were surprised about the amount of filtered reads and the length of the high quality reads. could you please tell me (if possible) the mean length and the percentage of passed filter?

              For example, for our last run (4 regions, cDNA from chicken and duck), for 2 regions (chicken) we only have ~15% of passed filters (whereas Roche expects more than 50% for a titanium run - information from GSFLX Titanium metrics, but it may be for a genomic DNA project). Problems came from dots (cf attached spreadsheet).

              Gerald
              Gerald,

              Your filtering output does not surprise me at all. We have seen very similar results with many of our cDNA runs on the 454. In the past, with the GS FLX Standard, most failed reads were rejected by the Mixed filter; now with the Titanium it seems that more are rejected by either the Dot or Short Quality filter. I don't know if this a difference in the sequence itself or in the filtering algorithms.

              Our read lengths for cDNA samples often show a bimodal distribution, with one peak at the expected 450-500nt range and second peak in the 100-150nt range. This results in a lower mean and median read length but those statistics don't give the full picture in these cases.

              Remember that all of the performance numbers which Roche publishes (e.g. 50% filter passed reads, 400-450nt mean read length) are based on sequencing very well behaved genomic DNA.

              Our protocol for preparing cDNA for 454 (SMART kit with modified first strand primer) was originally developed when our instrument was a GS-20, then carried through to the GS FLX. It may be a time to develop a new strategy for the Titanium.

              Comment

              • Emanuel Heitlinger
                Member
                • Mar 2009
                • 11

                #8
                Hi all,
                the Evrogen MINT kit worked better than the SMART kit for me. The SMART kit produced strange concatenated primer/adaptor artefacts. Currently it is used without any polyT modifications producing ~75.000 good quality sequences of ~250nt from 1/4 plate GS FLX run.

                Only problem: High amount of rRNA contamination ~50% in the best preparations.
                Does anybody else have this problem using SMART or MINT kits?

                I'm looking forward to get my first titanium runs...

                Comment

                • Emanuel Heitlinger
                  Member
                  • Mar 2009
                  • 11

                  #9
                  Wowwww
                  I just found all this:

                  Comment

                  • Melissa
                    Senior Member
                    • Aug 2008
                    • 124

                    #10
                    Thanks Emanuel for the website!

                    The sequencing center told me they have previously experienced the same problem. Sequencing mainly SMART concatemers. They suggested I use the minimum number of PCR cycles. I'm also expecting my titanium run.

                    Comment

                    • jmbon
                      Junior Member
                      • Nov 2009
                      • 2

                      #11
                      Does a broken T primer help ?

                      Hello everybody
                      We had on our first cDNA sequencing try, using a MINT kit used on a 1/4 Titanium plate. No concatenates, but a bimodal length distribution with a first peak around 70 nt and a second around 500 nt; about 50000 reads are below 100 nt, and 100000 above. Reads with a 5' polyT make up 50 % of the short, bad class. We used a first strand primer with 30 T, which I suspect to bring difficulties with pyrosequencing. Does anybody has experience with broken T primers ?

                      Thanks ahead!

                      Comment

                      • pmiguel
                        Senior Member
                        • Aug 2008
                        • 2328

                        #12
                        My impression of Roche's policy towards cDNA library construction is that they were not satisfied with any of the methods available and therefore they refused to recommend any of them. In the fullness of time Roche would release a method or kit that would produce acceptable results. I find this maddening as both Applied Biosystems and Illumina have had expression analysis solutions available for many months. Roche/454, who have been in the "next gen" game a year longer, are still dragging their feet.

                        But I also get the impression there were a couple of methods that at least approached the Platonic ideal of a perfect 454 cDNA library. These were (1)ribosomal depletion (eg, polyA+ isolation) of total RNA, random primed cDNA synthesis. Nebulization, etc as for genomic libraries or (2) oligo dT 1st strand synthesis using an offset restriction enzyme site (eg Mme or EcoP) adaptor. The adaptor and most of the polyA/T tail then is removed from the cDNA via restriction digest. The latter would likely involve some tedious tricks (eg methylation) to avoid cutting Mme or EcoP sites in the cDNA proper.

                        Anyway, we are dancing around something here and Roche is actually somewhat less mysterious than I think would normally be their wont. That is, they explicitly warn of the negative effects of polyA or polyT tracts. These tracts will blaze bright, obscuring nearby beads and possibly consuming all the dATP (or dTTP) in the flow before reaching the end of the tract. Hence the creation of the "broken oligo dT primer". This method apparently does not satisfy Roche, but their disapproval is somewhat muted.

                        But I think there is a second, related, issue in play. About this, Roche is more typically inscrutable, but one hears whispers. That is, one is led to believe that the image analysis software is sensitive to well-to-well bleed-over. Two beads sitting in adjacent wells of a PTP, if they emit too similar a sequence, are presumed to be "overlapping" and one or both are discarded. I'm not privy to what the threshold is for such a filter, if it does even exist, to kick in. It is of some significance, however; because SMRT cap adaptors, for instance, if they are sufficient to trigger high levels of bleed-over filtration, would be problematic. Especially if, as I believe is the case, the fragmentation broken ends are less often ligated than the extant capped ends.

                        --
                        Phillip

                        Comment

                        • pmiguel
                          Senior Member
                          • Aug 2008
                          • 2328

                          #13
                          Originally posted by Emanuel Heitlinger View Post
                          Hi all,
                          The SMART kit produced strange concatenated primer/adaptor artefacts.
                          I finally figured out what must be the source of adaptor concatamers that plague some SMART libraries.

                          If you are using the reverse transcriptase to add the 5' adaptor via MMLVs natural TdT activity (one to five C's are said to be added), then 5' concatamers would be the expected result. This is because MMLV will reverse transcribe RNA until it hits the end of the template, then it adds some C's and waits around for something to anneal at those C's. The 5' SMART oligo can anneal there. If it does, MMLV continues until it reaches the end of the 5' SMART oligo. (In essence MMLV has "switched" templates -- sometimes this is called "strand switching" or "template switching".)

                          After reaching the end of the new template (the SMART oligo) MMLV would again add non-templated C's.

                          This allows another 5' SMART oligo to anneal to the nascent strand, and the process can repeat over and over again, creating a string of concatamers.

                          The way to avoid this is to not add the 5' SMART adaptor to the reverse transcription. Instead just let the 5' SMART adaptor get added during 2nd strand synthesis/ amplification PCR. Basically a step-out from the non-templated C's that would terminate all the 1st strand cDNAs produced by MMLV. Taq polymerase doesn't add non-templated C's to the end of nascent strands, so no concatamers will result.

                          --
                          Phillip

                          Comment

                          • jmbon
                            Junior Member
                            • Nov 2009
                            • 2

                            #14
                            Originally posted by pmiguel View Post
                            But I think there is a second, related, issue in play. About this, Roche is more typically inscrutable, but one hears whispers. That is, one is led to believe that the image analysis software is sensitive to well-to-well bleed-over. Two beads sitting in adjacent wells of a PTP, if they emit too similar a sequence, are presumed to be "overlapping" and one or both are discarded. I'm not privy to what the threshold is for such a filter, if it does even exist, to kick in. It is of some significance, however; because SMRT cap adaptors, for instance, if they are sufficient to trigger high levels of bleed-over filtration, would be problematic. Especially if, as I believe is the case, the fragmentation broken ends are less often ligated than the extant capped ends.

                            --
                            Phillip
                            Thank you Phillip, this was a quick and informative answer. I do not fully follow your point about primers and bleed-over, though. If a primer sequence cause bleed over from well-to-well, many PCR products should be difficult to sequence in depth, and presumably not only SMART RT-PCR products. In our run with Evrogen RT_PCR kit, reads bearing in 5' the M1 cDNA primer are indeed in large excess (73%, for =<50% expected), so at least the M1 sequence seems safe for that matter. Could it be that what is bleeding is rather the too bright light of the polyT strech that follow every other 5'M1, making the neighbor beads difficult to read ?

                            Comment

                            • pmiguel
                              Senior Member
                              • Aug 2008
                              • 2328

                              #15
                              Originally posted by jmbon View Post
                              Thank you Phillip, this was a quick and informative answer. I do not fully follow your point about primers and bleed-over, though. If a primer sequence cause bleed over from well-to-well, many PCR products should be difficult to sequence in depth, and presumably not only SMART RT-PCR products. In our run with Evrogen RT_PCR kit, reads bearing in 5' the M1 cDNA primer are indeed in large excess (73%, for =<50% expected), so at least the M1 sequence seems safe for that matter. Could it be that what is bleeding is rather the too bright light of the polyT strech that follow every other 5'M1, making the neighbor beads difficult to read ?
                              I did not mean that the adaptor sequence caused bleed-over, only that it might look sufficiently like bleed-over to trigger the bleed-over filter.

                              Again, this is nearly pure speculation on my part, but the the reads that trigger the bleed-over filter will not be seen among your filter pass reads. They would have been filtered out.

                              I presume, the bleed-over filter would only be triggered if two beads in close physical proximity gave a similar sequence for some number of flows.

                              --
                              Phillip

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                10-30-2026, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              34 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              22 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              72 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              48 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...