    Junior and Flx compatibility

    Has anyone used Junior to sequence library made using GS FLX Titanium emPCR L kit? We tried doing so as our FLX sequencer is down but found only very few reads passed filter (even for control beads). So I am guessing there is some compatibility issue?

    any thoughts please?

    thanks!
  • RCJK
    Senior Member
    • May 2009
    • 156

    #2
    AFAIK the chemistries should be completely compatible. What type of library was it and what filters did the reads not pass?

    Comment

    • ngseq
      Member
      • Mar 2010
      • 19

      #3
      Thanks RCJK.

      It is a shotgun ligated library.

      here is what I got in the QualityFilterMetrics for the bad run (note control beads failed just as badly):

      key CATG GACT
      NumKeyPassed 261 248226
      NumDotFailed 6 30740
      NumMixedFailed 68 111699
      NumTrimmedTooShortQuality 109 102992
      NumTrimmedTooShortPrimer 0 437
      TotalPassedFilter 78 2357

      I re-ran emPCR using the Junior emPCR Titanium kit and got much improved result. Here is what the QualityFilterMetrics file looks:

      key ATGC CATG GACT
      NumKeyPassed 4054 2732 134569
      NumDotFailed 1 4 2157
      NumMixedFailed 59 26 10656
      NumTrimmedTooShortQuality 27 265 18689
      NumTrimmedTooShortPrimer 0 0 301
      TotalPassedFilter 3867 2437 102746

      Any clue what went wrong? Thanks!
      Last edited by ngseq; 08-11-2011, 12:24 PM.

      Comment

      • RCJK
        Senior Member
        • May 2009
        • 156

        #4
        Hmm...I'm not too sure. For the bad run there also looks to be a high level of mixed reads that were rejected. This can be due to a few things: the library not being denatured prior to addition to the emPCR, too high a cpb leading to too high an enrichment, broken emulsions. Perhaps any or all of those occurred and maybe this also would affect the number filtered for short quality? That one I'm not very sure of but maybe someone else has some thoughts?

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          Originally posted by RCJK View Post
          [...] too high a cpb leading to too high an enrichment, [...]
          Is it well known that if your cpb is really high (like >10x) you can end up getting enrichment percentages that look acceptable?

          In such cases, I think the numbers of positive beads overwhelm the system. Positive beads get lost because there are not enough enrichment beads to immobilize them all.

          There are clues during enrichment that something is badly wrong. But if you ignore the odd behavior and look only at the final %enrichment you can be led astray.

          I ask because we have run into this on both the SOLiD and the GS-FLX. But, best I can remember, I have never heard any warnings from Applied Biosystems or Roche about this possibility.

          --
          Phillip

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM
          • SEQadmin2
            New Genomics Technologies Take Aim at Long-Standing Limits
            by SEQadmin2


            Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

            We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
            ...
            09-28-2026, 10:25 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 10-15-2026, 10:52 AM
          0 responses
          28 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-14-2026, 05:39 AM
          0 responses
          36 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-13-2026, 09:56 AM
          0 responses
          31 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-09-2026, 11:14 AM
          0 responses
          29 views
          0 reactions
          Last Post SEQadmin2  
          Working...