Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • g861@hotmail.com
    Junior Member
    • Jun 2011
    • 6

    #1

    Same library to get different percentage of enriched beads

    We used the same dilution from the same library to do emPCR amplification. All the condition is the same, the same lot kit and the same person to do this experiment. These interval experimental time is 3 days. We did twice but got different volume beads, one is 10% and another is almost 30%. How come the result is so different?? Does anyone has the same experience?? I'm so confused that the result is convincable?
    Mandy:rolleyes:
  • suefo
    Member
    • Jul 2011
    • 36

    #2
    Are you using LVs, MVs, SVs or JR kits? Our JR kits tend to be consistent. But we see variation on the others - especially the MVs.

    Comment

    • g861@hotmail.com
      Junior Member
      • Jun 2011
      • 6

      #3
      JR means Junior? I used junior kit.
      Mandy:rolleyes:

      Comment

      • suefo
        Member
        • Jul 2011
        • 36

        #4
        Yes, Junior kits. I'm stumped as these tend to be the most consistent in our hands.

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          Yes, we see these sorts of enrichment yield variations frequently. (Actually we use MV emPCR kits most of the time.)

          How about room temp during emulsion formation? Seems like one major factor could be microreactor size, or microreactor size distribution. It seems like the viscosity of the emulsion oils would play a role here. Viscosity is often impacted by temperature. So if your lab is like mine you can see 4-5 oC temperature swings day to day. Maybe that is the cause? Just speculation, though.

          Does anyone routinely check the size of microreactors with a microscope?

          --
          Phillip

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            Oh, another possible factor: how many times the emPCR kit gets freeze-thawed. I was surprised to see that Roche has the DNA capture beads stored frozen. A typical MV emPCR kit might be freeze-thawed 4 or 5 times. Each time all the beads are re-washed. Seems fairly traumatic to me.

            --
            Phillip

            Comment

            • g861@hotmail.com
              Junior Member
              • Jun 2011
              • 6

              #7
              Is MV emPCR kits for FLX system? I can't change our protocol. My manager said how Roche trainer taught us then we did it. We wanna set up this system.
              We always keep room temp at 22oC, we didn't change this condition, so there is no possible. And in my opinion emPCR kit is okay, emPCR kit wasn't freeze-thawed. Per kit is for one experiment. So we just thawed it for once.
              Mandy:rolleyes:

              Comment

              • MissDNA
                Senior Member
                • Nov 2010
                • 146

                #8
                Yes, MV kits are used for FLX, usually when we divide the PTP in 4 or 8 regions.
                In our lab we mostly use 1 kit per experiment too and we observe enrichment differences from time to time. We always titrate our libraries but scaling up from SV to MV is still a mystery to us. Luckily our Guess-O-meter works well.
                For Sv kits when we thaw the kit for the first time we aliquot all the beads, and just wash the ones we are going to use for a given experiment.

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #9
                  Originally posted by [email protected] View Post
                  [...]
                  We always keep room temp at 22oC, we didn't change this condition, so there is no possible. And in my opinion emPCR kit is okay, emPCR kit wasn't freeze-thawed. Per kit is for one experiment. So we just thawed it for once.
                  As implied earlier in the thread, we also see large variations of this sort from emPCR reactions. My guess is that after doing 100's of emPCRs you learn what the critical parameters are. Sadly the people doing 100's of emPCRs either don't read this forum at all or don't feel the need to post in response to questions of this sort.

                  I can only suggest some of the factors we try to lock down:

                  (1) Use low bind tips and tubes for all work with diluted libraries. Even low levels of binding by plastics will alter the concentration of your diluted libraries.

                  (2) Check the calibration on pippettes periodically by weighing water on an analytical scale.

                  (3) It would be nice to add some additional quality control points to the procedure. I think looking at the microreactors under a microscope might be useful. Also maybe running a little of the aqueous phase of the emPCR reaction on a bioanalyzer chip might help.

                  Anyone else have recommendations for making the emPCR protocol more "robust" (repeatable with similar results)?

                  --
                  Phillip

                  Comment

                  • g861@hotmail.com
                    Junior Member
                    • Jun 2011
                    • 6

                    #10
                    Originally posted by MissDNA View Post
                    Yes, MV kits are used for FLX, usually when we divide the PTP in 4 or 8 regions.
                    Sorry, I don't understand you mean you separate PTP into 4 or 8 regions? I don't know about FLX- Titanium system, we only have Junior- Titanium machine. In junior, we can't divide PTP.
                    Mandy:rolleyes:

                    Comment

                    • g861@hotmail.com
                      Junior Member
                      • Jun 2011
                      • 6

                      #11
                      Originally posted by pmiguel View Post
                      (3) It would be nice to add some additional quality control points to the procedure. I think looking at the microreactors under a microscope might be useful. Also maybe running a little of the aqueous phase of the emPCR reaction on a bioanalyzer chip might help.
                      How to look at the microreactors under a microscope?
                      Another, emPCR is for DNA amplification in beads. DNA is already captured by beads, is that any effect in bioanalyzer running?
                      Mandy:rolleyes:

                      Comment

                      • MissDNA
                        Senior Member
                        • Nov 2010
                        • 146

                        #12
                        Originally posted by [email protected] View Post
                        Sorry, I don't understand you mean you separate PTP into 4 or 8 regions? I don't know about FLX- Titanium system, we only have Junior- Titanium machine. In junior, we can't divide PTP.
                        Oh and I am not familiar with Junior at all. With FLX we can divide the plate in 2, 4, 8 and 16 regions using a rubber gasket. So depending on how many regions we divide the plate, we can choose in between different emulsion oil presentations: LV (large volume), MV (medium volume) and SV (small volume). In our lab we usually use LV for 2 regions, MV for 4 and SV for 8 or 16.

                        Comment

                        • ajthomas
                          Senior Member
                          • Mar 2010
                          • 167

                          #13
                          I've seen this from time to time as well. Just recently I had a couple of libraries I titrated using an SV kit then scaled up to an MV kit. With the amount of DNA that gave me ~10% enrichment at the SV scale, I got only ~2-3% with the MV kit. I had to redo it to get enough beads. This time, I remade my dilution from the library stock and used the same amount as before. The result was the same ~10% enrichment I should have gotten the first time. It seems that an extra freeze-thaw cycle of 10^7 concentration library reduced its titer. I can't say for sure that's what happened, but that's they only way I can explain it at the moment.

                          Comment

                          • pmiguel
                            Senior Member
                            • Aug 2008
                            • 2328

                            #14
                            Originally posted by [email protected] View Post
                            How to look at the microreactors under a microscope?
                            Just pippette a little on to a microscope slide, put a cover slip over it (if necessary), and look under the microscope. Take a picture, if possible. Then, over time, you might see differences that correlate with emPCR results.
                            Originally posted by [email protected] View Post
                            Another, emPCR is for DNA amplification in beads. DNA is already captured by beads, is that any effect in bioanalyzer running?
                            You don't want to run the beads, just a ul of the aqueous solution from which you recover the beads. I guess only a tiny percentage of the total PCR reaction products are captured by the DNA capture beads.

                            --
                            Phillip

                            Comment

                            • ajthomas
                              Senior Member
                              • Mar 2010
                              • 167

                              #15
                              Originally posted by pmiguel View Post
                              You don't want to run the beads, just a ul of the aqueous solution from which you recover the beads. I guess only a tiny percentage of the total PCR reaction products are captured by the DNA capture beads.

                              --
                              Phillip
                              One of the PCR primers for emPCR is attached to the bead, so one strand of the PCR products is always attached to the bead. Some of the other strand will be in the aqueous phase, but recovering that would be difficult, since it will be in a large volume of isopropanol mixed with the emulsion oil. If you want to run a Bioanalyzer trace of the emPCR products, your best bet would be to use the supernatant from the melt solution step after breaking the emulsions. It would probably be a good idea to EtOH precipitate it first to get rid of the NaOH.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                Today, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              22 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              14 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              68 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              48 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...