    heat denature or not?

    Hei,

    I have fusion amplicons for one way sequencing (designed to be used with emPCR lib-L kit). Does one need to heat denature DNA before adding to Capture beads as it is described in emPCR lib-L protocol?
  • behavin
    Member
    • Oct 2011
    • 11

    #2
    I would say it would be a safe bet to heat denature, as you'll likely get lower than anticipated enrichment otherwise.

    Comment

    • ajthomas
      Senior Member
      • Mar 2010
      • 167

      #3
      You don't need to denature ampicons of that sort. You only need to denature the library if both strands can function in emPCR, which is the case with libraries prepared by ligating on rapid library adapters. Those adapters are Y-shaped, with the A adapter on one strand and the B adapter on the other strand. Therefore, both strands have an A and B adapter and both can serve as a template for emPCR. Amplicon libraries made with fusion primers have the A and B sequence on one each, so only one strand can function in emPCR.

      You can denature if you want, but it won't make any difference.

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        Originally posted by ajthomas View Post
        You don't need to denature ampicons of that sort. You only need to denature the library if both strands can function in emPCR, which is the case with libraries prepared by ligating on rapid library adapters. Those adapters are Y-shaped, with the A adapter on one strand and the B adapter on the other strand. Therefore, both strands have an A and B adapter and both can serve as a template for emPCR. Amplicon libraries made with fusion primers have the A and B sequence on one each, so only one strand can function in emPCR.

        You can denature if you want, but it won't make any difference.
        I think you are missing a subtlety here. The salient issue is whether both strands of the ds template are full "strand" and "strand complement". In essence, are both strands going to generate the same template to be captured by the bead, or will they each generate a different template.

        Y adapters create 2 different library molecules -- one in one orientation (with respect to the A adapter), the other in the other orientation. They are annealed together in the center, but upon emPCR will generate two different templates. When sequenced these should produce discordant sequences -- hence should produce a mixed bead 100% of the time.

        The non-Y adapter templates do not have this issue. So you do not need to denature them. However if you do denature them, your CPB should effectively double. Right? Two strands, either can be amplified in an emulsion reactor.

        We just denature everything now. Because if you have amplicons you will probably have some primer dimer annealed to your full length amplicons. I would rather peel this off into its own reactor where it will ruin that bead, but will allow the full length amplicon to produce a good sequence.

        Of course the denaturation can be problematic. If you do it in a small volume, you can end up throwing you concentration way off because of evaporation.

        --
        Phillip

        Comment

        • ajthomas
          Senior Member
          • Mar 2010
          • 167

          #5
          Originally posted by pmiguel View Post
          I think you are missing a subtlety here. The salient issue is whether both strands of the ds template are full "strand" and "strand complement". In essence, are both strands going to generate the same template to be captured by the bead, or will they each generate a different template.

          Y adapters create 2 different library molecules -- one in one orientation (with respect to the A adapter), the other in the other orientation. They are annealed together in the center, but upon emPCR will generate two different templates. When sequenced these should produce discordant sequences -- hence should produce a mixed bead 100% of the time.

          The non-Y adapter templates do not have this issue. So you do not need to denature them. However if you do denature them, your CPB should effectively double. Right? Two strands, either can be amplified in an emulsion reactor.
          Well, yes, that's basically what I was saying; you just stated it more explicitly. In the case of the amplicon, only one strand will function in emPCR because only one strand has the B adapter while the other has the reverse compliment of B. That second strand will, however, produce a copy that can then bind to the bead in the second round. Molecules with the Y-adapter will produce two strands (of different sequence) that can bind the bead in the first round.

          And yes, the Y-adapter will effectively double the cpb, although such a comparison requires 100% efficiency in the library prep. In practice, amplicon libraries produced with fusion primers will be used at a lower cpb ratio and will be more consistent because there is no ligation step, and every molecule will be functional (assuming the primers are good quality and all full length).
          We just denature everything now. Because if you have amplicons you will probably have some primer dimer annealed to your full length amplicons. I would rather peel this off into its own reactor where it will ruin that bead, but will allow the full length amplicon to produce a good sequence.

          Of course the denaturation can be problematic. If you do it in a small volume, you can end up throwing you concentration way off because of evaporation.

          --
          Phillip
          I hadn't really considered the idea of primer dimers annealed to the molecules. In my own sequencing, I haven't any evidence that that happens frequently, but I suppose it's possible. It would be interesting to an experiment to see if one really does get fewer mixed/dot reads if the amplicons are denatured first.

          I've also thought about the evaporation issue, and for that reason I have generally avoided denaturing amplicons. I've always figured that if it isn't needed, don't do it because that's just one more opportunity to introduce variability or for something to go wrong.

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM
          • SEQadmin2
            New Genomics Technologies Take Aim at Long-Standing Limits
            by SEQadmin2


            Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

            We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
            ...
            09-28-2026, 10:25 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 10-22-2026, 11:13 AM
          0 responses
          61 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 10:48 AM
          0 responses
          42 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-15-2026, 10:52 AM
          0 responses
          75 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-14-2026, 05:39 AM
          0 responses
          74 views
          0 reactions
          Last Post SEQadmin2  
          Working...