I have not once been able to fully dissolve the precipitate in the emPCR additive, even after heating to 55C. It does not seem to affect the sequencing results but I have noticed especially with the LV reaction that this precipitate seems to reappear during bead recovery and makes the bead pellet very difficult to break up after centrifugation.
Unconfigured Ad
Collapse
X
-
Tags: None
-
We have also seen a crystallized precipitate after heating GC additive. I suggest spinning down the tube after heating and only removing GC additive that is found above the pellet. We have found GC additive to aggregate around the bead pellet during LVE breaking, which disrupts washing and enrichment.
-
Hi Cambridge454,
Allow me to lend a hand. Are you familiar with Technical Bulletin 011-2009? If not, it is available on the my454 customer site
(http://www.454.com/my454) under the "Technical Bulletins" section.
Section 3.1 of that document, Step 1c. reads as follows:
"If a precipitate is observed after thawing the tube of emPCR Additive, vortex the tube well to dissolve it.
If precipitation remains in the emPCR Additive after vortexing, heat the reagent
to 55°C in a heat block for up to five minutes to aid in dissolving. If residual
precipitate remains, briefly centrifuge the tube and use the supernatant in the
preparation of the Live Amplification Mix."
In other words, the recommended course is to avoid carrying over residual precipitate, as suggested by guz.
Best regards,
Jason
Technical Product Manager
454 Life Sciences, A Roche CompanyTechnical Product Manager
454 Life Sciences, A Roche Company
Comment
-
Hi Jason,
Yes, I have read and followed the instructions regarding the additive in the Technical Bulletin 011-2009. There appears to be no precipitate in the Live Amp Mix, it only reappears again after bead recovery in the LVE, which makes the bead pellet very difficult to resuspend and wash.
Comment
-
I think I have a fix for this issue you mention (from TCB-09011). I believe the issue derives largely from this section of the protocol (section 3.5 "Bead Recovery):Originally posted by Cambridge454 View PostHi Jason,
Yes, I have read and followed the instructions regarding the additive in the Technical Bulletin 011-2009. There appears to be no precipitate in the Live Amp Mix, it only reappears again after bead recovery in the LVE, which makes the bead pellet very difficult to resuspend and wash.
The problem is that 1X Enhancing Fluid TW appears to cause the waxy deposit that makes resuspending the beads so difficult. So the longer you spend trying to resuspend the beads, the worse the clumping situation becomes.4. Add 10 ml of 1X Enhancing Fluid TW and vortex well to resuspend. If vortexing is not
sufficient to resuspend completely, use a glass rod or a spatula to break bead aggregates.
Note: It is important to fully suspend the bead suspension before adding
isopropanol in the next step to ensure complete mixing and to prevent clumping.
If, instead, you add the 1X Enhancing Fluid TW, vortex briefly and immediately add isopropanol, the formation of precipitate is much less. Clumping then does not seem to be an issue.
--
Phillip
Comment
-
Thanks for the advice. I think this might be my problem, as I usually pour off the isopropanol supernatant, add the 1x Enhancing Fluid TW, vortex, vortex, vortex, break clumps with glass rod, vortex, vortex, vortex, then add isopropanol.
Will give your fix a try.
Comment
-
We've found that trying to use a spatula to break up the aggregates makes the clumping worse. Vortex well after additing the Enhancing fluid and then go right to the isopropanol as mentioned above. Better to spend more time vortxing after the isopropanol addition that before. I've managed to get rid of all of the clumps this way.
Comment
-
I have heard it referred to as a polymerase "stabilizer" by one of the "old ones"[1]. That doesn't tell you much -- even if it is correct, however.Originally posted by Old guy View PostAnybody know what the emPCR additive is?
--
Phillip
1. By "old one", I mean someone who ran the instrument back when it was in its original incarnation as the GS-20.
Comment
Latest Articles
Collapse
-
by SEQadmin2
Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?
Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating...-
Channel: Articles
10-14-2026, 09:55 AM -
-
by SEQadmin2
Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.
We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing...-
Channel: Articles
-
ad_right_rmr
Collapse
News
Collapse
| Topics | Statistics | Last Post | ||
|---|---|---|---|---|
|
Started by SEQadmin2, Yesterday, 11:13 AM
|
0 responses
21 views
0 reactions
|
Last Post
by SEQadmin2
Yesterday, 11:13 AM
|
||
|
Started by SEQadmin2, Yesterday, 10:48 AM
|
0 responses
18 views
0 reactions
|
Last Post
by SEQadmin2
Yesterday, 10:48 AM
|
||
|
Started by SEQadmin2, 10-15-2026, 10:52 AM
|
0 responses
36 views
0 reactions
|
Last Post
by SEQadmin2
10-15-2026, 10:52 AM
|
||
|
Started by SEQadmin2, 10-14-2026, 05:39 AM
|
0 responses
42 views
0 reactions
|
Last Post
by SEQadmin2
10-14-2026, 05:39 AM
|
Comment