Seqanswers Leaderboard Ad

Collapse
X
 
  • Filter
  • Time
  • Show
Clear All
new posts
  • isalarubi
    Member
    • Feb 2013
    • 12

    Problems with 16S sequencing 454Jr

    Hi everybody!

    We're new with the 454 Jr, and in our first run of 16S to determine the stool microbiota we found some problems.

    We've had rawWells,304998 and then only keyPassWells,12489. When we see the color system, all are in pink as no key.

    Anybody knows what is happening?

    Thank you
  • ajthomas
    Senior Member
    • Mar 2010
    • 165

    #2
    Are the key pass wells all control beads? If so, I would guess that you grabbed the wrong tube when you annealed the sequencing primer and used the enrichment primer instead.

    Comment

    • isalarubi
      Member
      • Feb 2013
      • 12

      #3
      Thank you ajthomas, but we are quite sure that we did not use the primers in the wrong way.

      We do not understand the program very well yet. Probably our answer would be in the files, but we do not know whre we have to see

      Comment

      • bunce
        Member
        • Sep 2012
        • 55

        #4
        Hi isalarubi,
        Are you using LibA or LibL? - In our experience with amplicon sequencing on the Gs-jnr you want to aim for only 180,000-ish raw wells. The lack of complexity in the amplicon seems to cause the plate to flare. My advice is to be very careful about your bead:template ratio and load conservatively (i.e about 4% which is just below the line on your bead counter). If your control bead are not sequencing well it is often because you are overloaded. Cheers Mike

        Comment

        • isalarubi
          Member
          • Feb 2013
          • 12

          #5
          Thank you Mike,

          we are using Lib L. Definitely we need to be more conservative with the bead:template ratio, so that we were working with a higher one as in the training told us, but in our case it doesn't work.

          Thank you so much!!

          Isabel

          Comment

          • bunce
            Member
            • Sep 2012
            • 55

            #6
            For Lib-L with amplicons drop the emPCR primer down to a quarter and aim to load 180,000-ish beads. The platform is very sensitive to bead:template ratios - my advice if you are getting going is to use a qPCR assay to dial-in your ratio…. just because the bead enrichment is "in the window" is no guarantee that the ratio was 'correct'. good luck!

            Comment

            • newmoon
              Member
              • Oct 2010
              • 11

              #7
              would you post the filtermetrics summary so we can look at the whole run including the read length and control DNA metrics?

              You were overloading the JR, but I don't think that's the reason the key pass well was so low.

              Comment

              • isalarubi
                Member
                • Feb 2013
                • 12

                #8
                Dear Newmoon,

                here you have the files. I hope you can give us some information from them.

                Thank you!
                Attached Files

                Comment

                • newmoon
                  Member
                  • Oct 2010
                  • 11

                  #9
                  hi isalarubi,

                  this is not exactly what I want to see.
                  In the GS Run Browser, you can find different tags for wells, readlength, control DNA...
                  I have an example of filtermetrics from a Flx run / 2 region PTP here. If you can find these numbers for your run, I might have a clue about the problem.
                  (I have hard time to upload the file, too big. I past it here.)


                  Filter Metrics, GACT
                  GACT (Library) Region
                  1 2 Total
                  Raw Wells 945,367 955,708 1,901,075
                  Key Pass Wells 897,232 900,375 1,797,607

                  Failed Dot 23,309 10,154 33,463
                  Mixed 239,743 146,537 386,280
                  Short Quality 161,548 147,548 309,096
                  Short Primer 9,296 2,925 12,221

                  Passed Filter Wells 457,269 587,106 1,044,375

                  % Dot + Mixed 29.32 17.4 23.35
                  % Short 19.04 16.71 17.87
                  % Passed Filter 50.96 65.21 58.1



                  Read Length and Quality Metrics, Read Length, GACT
                  GACT (Library) Region
                  1 2 Total
                  Raw Wells 945,367 955,708 1,901,075
                  Key Pass Wells 897,232 900,375 1,797,607
                  Passed Filter Wells 457,269 587,106 1,044,375
                  Total Bases 149,917,353 208,858,977 358,776,330

                  Length Average 327.85 355.74 343.53
                  Length Std Deviation 105.42 101.17
                  Longest Reads Length 1,193 1,196 1,196
                  Shortest Reads Length 40 40 40
                  Median Reads Length 349 377 365



                  Control DNA Metrics
                  Region
                  1 2 Total
                  Raw Wells 945,367 955,708 1,901,075
                  All ControlDNA Wells 17,340 20,625 37,965
                  Average
                  % Match 200 bp, 100% 80.4 77.64 78.9
                  ≥98% 92.68 91.13 91.84
                  ≥95% 94.94 93.64 94.23

                  Comment

                  • isalarubi
                    Member
                    • Feb 2013
                    • 12

                    #10
                    Dear newmoon,

                    sorry, for the wrong file, I think this new one is the good one.

                    Thank you!!

                    Isabel
                    Attached Files

                    Comment

                    • newmoon
                      Member
                      • Oct 2010
                      • 11

                      #11
                      I only see the filter metrics. It was bad with the key pass wells being so low.

                      what about the read length, is median read length around 350bp?

                      and what about control DNA metrics, is %match 95% above 85%? If the control DNA was good, that means the run was good, but your beads had problem.If the control DNA has 95% match below 75%, I would think the run itself had problem, may not the beads problem.

                      The only time we had low key pass wells was when our machine had fluidic issue which leaked during the flow. If you know how to look at the images, you may find a clue.

                      Have you sent a run report to 454 tech support for a diagnosis? They should be able to tell you if it's machine problem or beads problem.

                      Comment

                      • newmoon
                        Member
                        • Oct 2010
                        • 11

                        #12
                        I don't know why the reply I posted yesterday didn't show up here.

                        anyway, I would suggest to look at the control DNA metrics which you didn't post in the last post. If the control %match 95% was above 80, the run itself was good, which means the beads had problem. If the control %match 95% is below 75, the run had problem, which could be the reagents or the machine had problem.

                        have you sent a run report to 454 tech support? They should be able to tell you if the machine or the beads had problem.

                        Comment

                        • isalarubi
                          Member
                          • Feb 2013
                          • 12

                          #13
                          Thanks a lot NEwmoon,

                          we are in touch with the technical support from Roche, and well, they have said that the machine is ok, and that it could be only an isolated event, and I hope so.

                          We are in the middle of a new run, and I hope it will be ok.

                          Fingers crossed!!

                          Comment

                          Latest Articles

                          Collapse

                          • seqadmin
                            Pathogen Surveillance with Advanced Genomic Tools
                            by seqadmin




                            The COVID-19 pandemic highlighted the need for proactive pathogen surveillance systems. As ongoing threats like avian influenza and newly emerging infections continue to pose risks, researchers are working to improve how quickly and accurately pathogens can be identified and tracked. In a recent SEQanswers webinar, two experts discussed how next-generation sequencing (NGS) and machine learning are shaping efforts to monitor viral variation and trace the origins of infectious...
                            Today, 11:48 AM
                          • seqadmin
                            New Genomics Tools and Methods Shared at AGBT 2025
                            by seqadmin


                            This year’s Advances in Genome Biology and Technology (AGBT) General Meeting commemorated the 25th anniversary of the event at its original venue on Marco Island, Florida. While this year’s event didn’t include high-profile musical performances, the industry announcements and cutting-edge research still drew the attention of leading scientists.

                            The Headliner
                            The biggest announcement was Roche stepping back into the sequencing platform market. In the years since...
                            03-03-2025, 01:39 PM
                          • seqadmin
                            Investigating the Gut Microbiome Through Diet and Spatial Biology
                            by seqadmin




                            The human gut contains trillions of microorganisms that impact digestion, immune functions, and overall health1. Despite major breakthroughs, we’re only beginning to understand the full extent of the microbiome’s influence on health and disease. Advances in next-generation sequencing and spatial biology have opened new windows into this complex environment, yet many questions remain. This article highlights two recent studies exploring how diet influences microbial...
                            02-24-2025, 06:31 AM

                          ad_right_rmr

                          Collapse

                          News

                          Collapse

                          Topics Statistics Last Post
                          Started by seqadmin, 03-20-2025, 05:03 AM
                          0 responses
                          26 views
                          0 reactions
                          Last Post seqadmin  
                          Started by seqadmin, 03-19-2025, 07:27 AM
                          0 responses
                          33 views
                          0 reactions
                          Last Post seqadmin  
                          Started by seqadmin, 03-18-2025, 12:50 PM
                          0 responses
                          25 views
                          0 reactions
                          Last Post seqadmin  
                          Started by seqadmin, 03-03-2025, 01:15 PM
                          0 responses
                          190 views
                          0 reactions
                          Last Post seqadmin  
                          Working...