  • pmiguel
    replied
    Originally posted by kmcarr View Post
    Yes, it certainly would be nice. To check for B-adapter I output the untrimmed FASTA from the SFF file (sffinfo -s -n) and then search the reads against the B-adapter sequence with cross_match or fuzznuc (or your favorite sequence search tool).
    Yes.
    As it turns out cross_match and fuzznuc are my favorite sequence search tools.
    Have you tried to see long (read-killing) homopolymer runs this way? Or do they not show up in even the untrimmed sequence?

    --
    Phillip

    Leave a comment:


  • kmcarr
    replied
    Originally posted by pmiguel View Post
    Wow, thanks for that information. Seems absolutely crazy for Roche not to provide a metric or graph that shows the length of library molecules for which a B-adapter sequence was located. With that metric you can immediately see many library issues. Without it you can waste your time looking for instrument problems.
    Yes, it certainly would be nice. To check for B-adapter I output the untrimmed FASTA from the SFF file (sffinfo -s -n) and then search the reads against the B-adapter sequence with cross_match or fuzznuc (or your favorite sequence search tool).

    Leave a comment:


  • pmiguel
    replied
    Originally posted by kmcarr View Post
    Phillip, reads will only be reported as short primer if they are rejected because they are too short (≤ 84 flows, ~50nt) after trimming off the B-adapter sequence. If an insert is 250nt long the raw read will likely traverse into the adapter which will be trimmed from the final read output. This will not be included in the shortPrimer filter count, that only reports reads which are rejected, not simply trimmed.
    Wow, thanks for that information. Seems absolutely crazy for Roche not to provide a metric or graph that shows the length of library molecules for which a B-adapter sequence was located. With that metric you can immediately see many library issues. Without it you can waste your time looking for instrument problems.

    I have been trying to convince Applied Biosystems of the same thing.
    --
    Phillip

    Leave a comment:


  • seqAll
    replied
    Originally posted by LMcSeq View Post
    Hi all,
    Quick question: I'm finding the my average frag length on the high sensitivity chip at the end of the rapid protocol is on the larger side. My sequencer results have had low average fragment lengths. I'm wondering if the larger frags aren't being amplified because they're too big for the microreactors, causing preferential amplification of smaller frags. I'm using the Covaris settings recommended for 500bp average. Anyone else having any similar issue?
    Hi LMcSeq,

    Do you have leftover enrichment beads? If so, you can run normal PCR with a few of these beads as template, using the amplification primers, for a few cycles (with the same cylcling condition as emPCR, and maybe one or two cycles, so not to skew the size distribution). Then run on a gel to check the size distribution.

    Next time, you can save the supernatant after melting (the supernatant are single-stranded templates released off the beads), and run them on a gel as a control step before you proceed to sequencing.

    It would be interesting to see if it is the sequencing part (or maybe the algorithm) that can not take such long fragments.
    Last edited by seqAll; 03-05-2010, 01:47 AM.

    Leave a comment:


  • kmcarr
    replied
    Originally posted by pmiguel View Post
    Okay, but if short primer is in spec, that means you do not have a large number of reads that traverse the insert completely. So preferential amplification of your small templates is not the issue.
    Phillip, reads will only be reported as short primer if they are rejected because they are too short (≤ 84 flows, ~50nt) after trimming off the B-adapter sequence. If an insert is 250nt long the raw read will likely traverse into the adapter which will be trimmed from the final read output. This will not be included in the shortPrimer filter count, that only reports reads which are rejected, not simply trimmed.

    Leave a comment:


  • pmiguel
    replied
    Originally posted by LMcSeq View Post
    That's what I suspect is the problem, but I'm not sure how to effectively troubleshoot it with the new rapid protocol since the chip isn't run until the end of all of the steps are complete. I don't want to have to make libraries over and over...
    I'm not sure if I should do a calibration with my XP beads and check the avg. frag length with different Covaris shearing times (don't know if it translates to the new procedure with the sizing solution and standard XP volume).
    Any suggestions?
    We would just size fractionate with an agarose gel if we got a library too much outside the size range. I realize that is not optimum for large numbers of samples, though.

    Also, why not use nebulization instead of sonication? I believe nebulized DNA would have a higher fraction of ligatable ends after end-repair than sonicated DNA.

    --
    Phillip

    Leave a comment:


  • LMcSeq
    replied
    That's what I suspect is the problem, but I'm not sure how to effectively troubleshoot it with the new rapid protocol since the chip isn't run until the end of all of the steps are complete. I don't want to have to make libraries over and over...
    I'm not sure if I should do a calibration with my XP beads and check the avg. frag length with different Covaris shearing times (don't know if it translates to the new procedure with the sizing solution and standard XP volume).
    Any suggestions?

    Leave a comment:


  • pmiguel
    replied
    Originally posted by LMcSeq View Post
    Hi Phillip,
    Thanks for your reply. We've had our 454 FAS out and he said that its likely a problem in library prep or emPCR based on the output from the support tool (which customers can't view themselves). He said the short primer is in spec.

    We are shearing whole genomic microbial DNA, not cDNAs. I think that was another contributor to this thread.
    Okay, but if short primer is in spec, that means you do not have a large number of reads that traverse the insert completely. So preferential amplification of your small templates is not the issue.

    It could be that you have too many templates longer than the maximum optimum amplification length of the emPCR kit (~1 kb, I think). Those would tend to produce lightly templated beads that might give you short reads.

    --
    Phillip

    Leave a comment:


  • LMcSeq
    replied
    Hi Phillip,
    Thanks for your reply. We've had our 454 FAS out and he said that its likely a problem in library prep or emPCR based on the output from the support tool (which customers can't view themselves). He said the short primer is in spec.

    We are shearing whole genomic microbial DNA, not cDNAs. I think that was another contributor to this thread.

    Leave a comment:


  • pmiguel
    replied
    Originally posted by LMcSeq View Post
    Hi all,
    Quick question: I'm finding the my average frag length on the high sensitivity chip at the end of the rapid protocol is on the larger side. My sequencer results have had low average fragment lengths. I'm wondering if the larger frags aren't being amplified because they're too big for the microreactors, causing preferential amplification of smaller frags. I'm using the Covaris settings recommended for 500bp average. Anyone else having any similar issue?
    It could be. Typical PCR reactions yield vastly more product for short templates than for longer ones.

    But it could also be that your read lengths are short for some other reason. If your templates are actually short you should have a high number in your "Short Primer" metric. If not, then something else is likely the source of your short reads.

    I will mention one: polyA, of course, is the bane of 454 runs...

    Even with random primed cDNA libraries, it is possible to run into polyA problems. It should be less, but polyA+ RNA isolation will enrich for polyA. If you start with heavily degraded RNA and pull out polyA from that, you may get a high percentage of your cDNA being polyA or polyT--even if you used random primed reverse transcription.

    --
    Phillip

    Leave a comment:


  • LMcSeq
    replied
    Hi all,
    Quick question: I'm finding the my average frag length on the high sensitivity chip at the end of the rapid protocol is on the larger side. My sequencer results have had low average fragment lengths. I'm wondering if the larger frags aren't being amplified because they're too big for the microreactors, causing preferential amplification of smaller frags. I'm using the Covaris settings recommended for 500bp average. Anyone else having any similar issue?

    Leave a comment:


  • pmiguel
    replied
    Originally posted by Old guy View Post
    Do you use different standards to quantitate based on the GC content of the genome or based on the size of the genome?
    No. Most of the time these figures are "ballpark" anyway. 50% high or low will make little difference.

    Phenol or RNA, though, can easily contribute >90% of the UV absorbance of a genomic DNA sample. 10-fold low on the amount of input DNA does make a difference.

    --
    Phillip

    Leave a comment:


  • Old guy
    replied
    Flourimeter question

    Originally posted by pmiguel View Post
    Before. But use a fluorimeter to quantitate. DNA preps, especially genomic DNA preps are often >90% RNA (sometimes degraded to short pieces with RNase). Or phenol! Phenol absorbs strongly at 270 nm. So if you use UV spectrophotometry, make sure your peak is at 260, not 270.

    --
    Phillip
    Do you use different standards to quantitate based on the GC content of the genome or based on the size of the genome?

    Leave a comment:


  • RCJK
    replied
    Hi JohnG,
    I used the Rapid Library MID adaptors provided by Roche. They provide a kit with 12 MIDs. The RL adaptors and MID adaptors are different than those of the general library prep. They are now Y adaptors labeled with FAM (I believe) for easier quantification at the end of the protocol.

    Cheers

    Leave a comment:


  • JohnG
    replied
    Hello,

    RCJK which type of MID tags have you used for rapid library preparation? Have you tried to use MID tags from standard library protocol?

    I've asked Roche support about using MIDs from standard protocol for preparing Rapid library, but they told me that it is not recommended.
    I think that they probably changed primers concentration used in adaptor ligation step. Both standard and rapid MIDs can be used later in emPCR so primers sequence is probably the same.

    Cheers

    John

    Leave a comment:

Latest Articles

Collapse

  • SEQadmin2
    Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
    by SEQadmin2


    Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

    The systematic characterization of the human proteome has
    ...
    07-20-2026, 11:48 AM
  • SEQadmin2
    Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
    by SEQadmin2



    Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
    ...
    07-09-2026, 11:10 AM
  • SEQadmin2
    Cancer Drug Resistance: The Lingering Barrier to Rising Survival
    by SEQadmin2



    Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

    There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
    07-08-2026, 05:17 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, 07-24-2026, 12:17 PM
0 responses
31 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 07-23-2026, 11:41 AM
0 responses
23 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 07-20-2026, 11:10 AM
0 responses
214 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 07-13-2026, 10:26 AM
0 responses
79 views
0 reactions
Last Post SEQadmin2  
Working...