  • MissDNA
    Senior Member
    • Nov 2010
    • 146

    #1

    Issues with new XLR70 lot

    Only recently we have received a lot of XLR70 containing the new control beads. Since them I have observed a decrease of the read length for our library and control type 1 as well. We have had a few runs where the control length was bellow 300 bp. I contact Roche support and explained the situation. The lot we have been using is #93801720. I was told there was already other complains on that lot.
    Has anyone experienced any problem with that lot or any of the new lots containing the new beads?
    Thanks
  • genes123
    Junior Member
    • Jul 2010
    • 7

    #2
    We had the exact same experience with that lot#. Finally Roche agreed to replace a few kits but never acknowledged it was the entire lot that was faulty (we didn't get anything above 350, but they only replaced runs we got at >300bp. We went back to >400bp as soon as we switched lot#)

    Comment

    • genes123
      Junior Member
      • Jul 2010
      • 7

      #3
      I re-checked and our lot# was actually 93801620. There is already another thread on problems with 93801720

      Comment

      • MissDNA
        Senior Member
        • Nov 2010
        • 146

        #4
        Thank you! I am gonna look for the other thread.
        We haven´t had any official word from Roche yet and I doubt they will acknowledge there is a problem with the lot and make a recall. We are stuck with 16 boxes of that lot. The last couple of runs were not so terrible. Since we´ve started using that lot none of our runs had reads longer than 350. I am currently waiting for answers on the 3 runs I forwarded to support.
        I don´t understand why they change products that are working well. It happened with the gaskets, then emPCR kits and now with XRL70.

        Comment

        • genes123
          Junior Member
          • Jul 2010
          • 7

          #5
          We made a pretty convincing case by sequencing the same library on 2 different lots and by plotting the stats of all our runs over the past 6 months (outsiders all had the same lot# in common) , but it still took 1 month and a half to get feedback and some replacement kits from Roche.

          Comment

          • MissDNA
            Senior Member
            • Nov 2010
            • 146

            #6
            Oh yeah, I did the same. I sent them an spreadsheet with the data of all the runs we made from November to now, with lot numbers and such. They keep finding all sorts of excuses and I keep repeating myself. I got an email last Friday, with a lot of nonsense to which I replied and haven´t heard back from them since. I imagine it is hard for them to admit there is problem with the lot, when they know I have so many boxes in stock.

            Comment

            • Soulbee
              Member
              • Jun 2010
              • 25

              #7
              Agreed. I had to also do many test runs before I got any replacement oils. Do you know if they are coming up with any new developments as far as the software or reagents go? I feel that compared to Illumina or Solid, Roche has been very quiet.. just wondering what is going on.

              Comment

              • MissDNA
                Senior Member
                • Nov 2010
                • 146

                #8
                Our problem persists. We got several replacement kits but so far no concrete answer on our problem. We have been waiting for control beads to perform a Tf run since November. It should arrive today. A Roche representative is coming to our lab today and I will see what he has to say. Since we started using the new lots of XLR70 with the new control beads we haven´t had a run that reached 400+bp median size of reads. Several runs have been bellow 300. Finally Roche has asked us to not perform any runs till the Tf run is made.
                We had a problem with emPCR around October but my suspicious were on the emPCR reagent. I think I was right because since that lot ended we haven´t had any problems.
                I think they are working on larger reads. Something around 800-1000. At least they told me that was a reason for the new control beads with 2 sizes of fragements.

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                  by SEQadmin2



                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                  ...
                  10-14-2026, 09:55 AM
                • SEQadmin2
                  New Genomics Technologies Take Aim at Long-Standing Limits
                  by SEQadmin2


                  Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                  We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                  ...
                  09-28-2026, 10:25 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, Yesterday, 11:13 AM
                0 responses
                21 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, Yesterday, 10:48 AM
                0 responses
                18 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-15-2026, 10:52 AM
                0 responses
                36 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-14-2026, 05:39 AM
                0 responses
                42 views
                0 reactions
                Last Post SEQadmin2  
                Working...