    Rules for making your own index

    We're planning to add a second index to the TruSeq v2 kit because we need more multiplexing than just 24. Are there any rules on making your own index?

    I asked Illumina and they said to make sure that for each cycle A,T,C,G are all represented because the MiSeq has to "focus" or else the cycle is lost. So this means I can't make a universal tag since there will be cycles where all my base reads would only consist of a single nucleotide! Is this accurate?
  • ETHANol
    Senior Member
    • Feb 2010
    • 308

    #2
    Not really sure if I am getting what you are saying but it's best to keep all the hamming distance >1 for all the barcodes.
    --------------
    Ethan

    Comment

    • kentk
      Member
      • Dec 2011
      • 17

      #3
      Yes I should maximize hamming distance.
      But for example I have 4 indexes...

      5' ATGCAT
      5' TGAACG
      5' GCTGTC
      5' AGCTGC

      An Illumina representative mentioned that I can't use that index set because the first, second and last bases will not have all of the four bases. So one the flowcell for base 1, I'll have signals for A, T, G clusters but not for C so our machine (MiSeq) will trash that cycle. Well this is what I understood from our conversation.

      Any thoughts?

      Comment

      • ETHANol
        Senior Member
        • Feb 2010
        • 308

        #4
        On the HiSeq, you need balance nucleotide composition at the beginning of the sequencing read but not the barcode read. Otherwise you could only multiplex in multiples of four. Which was one of the drawbacks of putting the barcode at the beginning of the sequencing read. Maybe the MiSeq is more picky about the barcode read, I don't know.
        --------------
        Ethan

        Comment

        • HESmith
          Senior Member
          • Oct 2009
          • 512

          #5
          Actually, ETHANol's statement is not 100% accurate. On the HiSeq, high cluster densities (900-1000K) have a more deleterious effect on index reads than inserts. We've had several flow cells with good cluster calling (80-90% PF) and high quality scores (mean ~38), yet fewer than 50% of the indices were called accurately. In some cases, pseudotiles at the inflow side (which contain higher cluster densities) have completely dropped out (i.e., no basecalling) during the index read after producing high-quality insert reads. The problem can be mitigated by balancing the ratio of index bases that are excited by the same laser (A/C or G/T).

          If your second index is at the start of read one, then you absolutely have to use all four bases in roughly equal proportions for the first four cycles (which is when cluster calling occurs).

          Comment

          • ETHANol
            Senior Member
            • Feb 2010
            • 308

            #6
            HESmith, Thanks for the correction. I'm curious here. How do you determine that the indices are called incorrectly? How do I go about performing QC on the index read?
            --------------
            Ethan

            Comment

            • ETHANol
              Senior Member
              • Feb 2010
              • 308

              #7
              It's funny, I found this on the internet some time ago and follow it but Illumina hasn't followed up on it so I just assumed it wasn't a problem. Apparently, it can be. Which leads one to ask, why is this not mentioned in any of the library preparation manuals.

              I think pmiguel has said that base balanced base composition for the index read is important on the HiScan.


              1. Some sequencing experiments require the use of fewer than 12 index sequences in a lane with a high cluster density. In such cases, select indexes carefully to ensure optimum base calling and demultiplexing by having different bases at each cycle of the index read. Illumina recommends the following sets of indexes for low-level pooling experiments.
              Pool of 2 samples:
              • Index #6 GCCAAT • Index #12 CTTGTA

              Pool of 3 samples:
              • Index #4 TGACCA • Index #6 GCCAAT • Index #12 CTTGTA

              Pool of 6 samples: • Index #2 CGATGT • Index #4 TGACCA • Index #5 ACAGTG • Index #6 GCCAAT • Index #7 CAGATC • Index #12 CTTGTA
              --------------
              Ethan

              Comment

              • kentk
                Member
                • Dec 2011
                • 17

                #8
                Thanks guys. Yes I was planning to introduce a 5' index. Being able to multiplex only at multiples of 4 isn't a problem. Just need to multiplex into the hundreds.

                I think I've read the same post by pmiguel mentioning index reads should always contain a A/C and G/T at each position that is why I was curious why all bases should be in equal proportion.

                Comment

                • HESmith
                  Senior Member
                  • Oct 2009
                  • 512

                  #9
                  Originally posted by ETHANol View Post
                  HESmith, Thanks for the correction. I'm curious here. How do you determine that the indices are called incorrectly? How do I go about performing QC on the index read?
                  I examined the frequencies of different indices in the Undetermined directory. The most common were one-base mismatches with the correct indices (we required perfect matches for demultiplexing), but there were nearly as many with two or more mismatches. Also, there were some pseudotiles with all Ns in the index despite high quality insert reads. Note that we observed this problem only at very high cluster densities.

                  You can use SAV or HCS to visualize the Q-scores for the index cycles. They are usually a bit lower than read one; if they're a lot lower, be concerned. A high fraction (>3-4%) of reads in the Undetermined directory is another indication of poor index reads.

                  Harold

                  Comment

                  • ETHANol
                    Senior Member
                    • Feb 2010
                    • 308

                    #10


                    With a 5' index you have the invariant T required for adapter ligation in all libraries. I guess it doesn't cause too much of a problem because people use this strategy, but it is something to think about nonetheless. Has this caused problems for anyone?
                    --------------
                    Ethan

                    Comment

                    • ETHANol
                      Senior Member
                      • Feb 2010
                      • 308

                      #11
                      Thanks Harold!
                      --------------
                      Ethan

                      Comment

                      • HESmith
                        Senior Member
                        • Oct 2009
                        • 512

                        #12
                        Originally posted by kentk View Post
                        Thanks guys. Yes I was planning to introduce a 5' index. Being able to multiplex only at multiples of 4 isn't a problem. Just need to multiplex into the hundreds.

                        I think I've read the same post by pmiguel mentioning index reads should always contain a A/C and G/T at each position that is why I was curious why all bases should be in equal proportion.
                        There's a distinction b/t the Illumina index read (which is separate) vs. barcodes that are incorporated at the start of your insert. In addition to cluster calling, I believe that the measured signal intensities for the first four cycles are used to calibrate values that are utilized for the remainder of the run (e.g., signal-to-noise), which would obviously affect the data if the bases are not equally represented in those cycles.

                        For the index read, A/C vs. G/T is usually sufficient to discriminate between a small number of barcodes.

                        Harold

                        Harold

                        Comment

                        • kentk
                          Member
                          • Dec 2011
                          • 17

                          #13
                          Originally posted by ETHANol View Post
                          http://www.plosone.org/article/info%...l.pone.0016607
                          With a 5' index you have the invariant T required for adapter ligation in all libraries. I guess it doesn't cause too much of a problem because people use this strategy, but it is something to think about nonetheless. Has this caused problems for anyone?
                          Article looks interesting. I'll have to read it through first. Thanks again ETHANol

                          You mean the T for the T-A ligation right? No I don't think it's a problem because that T (or actually its complement A) anneals to the last base of the sequencing primer so essentially it's not part of the read

                          Comment

                          • HESmith
                            Senior Member
                            • Oct 2009
                            • 512

                            #14
                            Originally posted by ETHANol View Post
                            http://www.plosone.org/article/info%...l.pone.0016607

                            With a 5' index you have the invariant T required for adapter ligation in all libraries. I guess it doesn't cause too much of a problem because people use this strategy, but it is something to think about nonetheless. Has this caused problems for anyone?
                            I assume you mean barcodes that are part of the adapter. Unless your index is only three bases long, you should be okay (but I haven't done the experiment). You could also resolve the problem by using indices of different length so the T is phase-shifted, and balance the other nucleotides for that cycle.

                            Comment

                            • TonyBrooks
                              Senior Member
                              • Jun 2009
                              • 303

                              #15
                              On a related note; I have a bunch of indexes (The Sanger 96-plex ones) that I'd like to use for low plexing too (4plex). These indexes are 8 bases long. Is there anything stopping just reading the first 6 bases (as per standard illumina indexing) on the GAIIx/HiSeq as long as there is AC/GT balance at all 6 positions? I only want to do this on one lane, so don't need to read 8 cycles on the other 7 lanes.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                New Genomics Technologies Take Aim at Long-Standing Limits
                                by SEQadmin2


                                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                ...
                                Yesterday, 10:25 AM
                              • SEQadmin2
                                How Immunogenomics Decodes Immunity’s Genetic Blueprint
                                by SEQadmin2




                                The immune system’s power comes from its genetic diversity, allowing myriad threats to be neutralized through first recognizing foreign antigens. That diversity is also what makes the immune system so difficult to study. Recent advances in sequencing technology and computational biology, however, are giving researchers new tools to understand immune responses and immune-related diseases in greater detail.

                                This convergence of genetics, immunology, and computation...
                                09-01-2026, 05:41 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, Today, 09:51 AM
                              0 responses
                              9 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 09-25-2026, 09:06 AM
                              0 responses
                              32 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 09-23-2026, 11:05 AM
                              0 responses
                              27 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 09-18-2026, 11:37 AM
                              1 response
                              47 views
                              0 reactions
                              Last Post pekgio
                              by pekgio
                               
                              Working...