  • induivain
    Junior Member
    • Sep 2011
    • 9

    #1

    Bad quality at the end of the read 150pb Miseq

    Hi everyone!!!

    I am working with Sure select capture , and the miseq in a DNAseq experiment.

    This is my first time that I use the paired end read at 150bp , and I have a problem at the end of the reads, I have a bias in the last bases. I remove all reads with Q<30 , remove adaptors with cutadapt and trimomatic, but the bias is not resolved.

    The mean of the insert size is 198bp , and the adaptors are the tipical for illumina.

    The size mean of the library is 350bp before sequecing.

    this happens someone???

    Thanks in advance.
    Attached Files
    Last edited by induivain; 04-26-2013, 03:22 PM.
  • MLog
    Member
    • Jan 2010
    • 36

    #2
    Hi induivain,

    I observe exactly the same effect (even stronger, example attached) on my 250-bp runs. My example is from a recent run of an exome sample, but it seems that this bias is not related to the exome capture, because I see the same on total DNA samples also: http://makarich.fbb.msu.ru/temporary...ce_content.png
    Attached Files

    Comment

    • induivain
      Junior Member
      • Sep 2011
      • 9

      #3
      Hi MLog,
      Thank you for you reply
      It is good to know that I am not the only one with this effect.
      I am very curious, is posible to know what is the library size before sequencing?

      Comment

      • MLog
        Member
        • Jan 2010
        • 36

        #4
        Originally posted by induivain View Post
        I am very curious, is posible to know what is the library size before sequencing?
        Certainly yes - agilent or agarose electrophoresis. (Though actually Agilent sometimes generates weird results). I think this bias could a be result of short inserts, those that are shorter than the read length. In this case you will be reading the adapter sequence. You mention that the insert size is about 200 bp - but this is only the average value. Depending on the distribution of fragment lengths you still can have a number of fragments that are shorter than 150 bp.

        Comment

        • induivain
          Junior Member
          • Sep 2011
          • 9

          #5
          You mention that the insert size is about 200 bp - but this is only the average value. Depending on the distribution of fragment lengths you still can have a number of fragments that are shorter than 150 bp.
          I was thinking the same , but only 10 % has a minor size that 150bp, and I only use reads between 151bp to 300bp.

          Agilent or agarose? , It will be anyone nice.
          Attached Files
          Last edited by induivain; 04-28-2013, 11:24 AM.

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
            by SEQadmin2




            NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
            10-30-2026, 06:46 AM
          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 10-28-2026, 09:57 AM
          0 responses
          33 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-28-2026, 09:51 AM
          0 responses
          21 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 11:13 AM
          0 responses
          72 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 10:48 AM
          0 responses
          48 views
          0 reactions
          Last Post SEQadmin2  
          Working...