  • firefly2280
    Member
    • Oct 2012
    • 28

    #1

    MiSeq run failure - inhibition?

    Hi all,

    I recently did a MiSeq run that failed before it had really begun. I am doing 16S custom sequencing with the barcodes.

    Everything was properly calculated, diluted and Qubited, so I am sure the right concentrations were loaded (went for 8pM) and 10% Phi X.

    The only thing we can think it might have been is that some chlorine or other inhibitors may have got through and ruined the run.... Does anyone have any ideas about this or any experience with inhibitory samples?

    thanks
  • barrmur
    Member
    • Nov 2009
    • 26

    #2
    What do you mean by failed? No clusters formed, low intensity data, fluidics check failed??
    Last edited by barrmur; 07-16-2013, 06:10 AM.

    Comment

    • firefly2280
      Member
      • Oct 2012
      • 28

      #3
      The error message

      "Object reference not set to an instance of an object”

      came up and that was it, so didnt even get as far as cluster densities.

      Illumina thought it was the concentration of my library but It was definately at 2nM to start with then denatured and diluted correctly as per their protocol

      Comment

      • AStretton
        Member
        • Jul 2012
        • 25

        #4
        When we had that error come up, it was because we had a pinched line. Try doing a fluidics check to see if there is a blockage.

        Comment

        • firefly2280
          Member
          • Oct 2012
          • 28

          #5
          Thanks. Other people are doing successful runs, so I think it is a sample issue sadly!

          Comment

          • mcnelson.phd
            Senior Member
            • Jul 2011
            • 162

            #6
            Originally posted by firefly2280 View Post
            The error message

            "Object reference not set to an instance of an object”

            came up and that was it, so didnt even get as far as cluster densities.

            Illumina thought it was the concentration of my library but It was definately at 2nM to start with then denatured and diluted correctly as per their protocol
            Unfortunately, that error message comes up for a large number of reasons on the MiSeq, and it's the most uninformative error ever because you can never figure out what actually caused a problem.

            We've had that specific message show up for a number of run failures, and only two or three could be traced back to the same cause. In that case, our optical table wasn't fully secured, so over the course of a run the images became progressively more out of focus until the system just gave up and quit. Other reasons for that error have been traced to bad reagents, bad libraries, and a few other random problems.

            Since you said people are successfully using the MiSeq now, best thing to do is get Illumina to replace your kits and try the samples again.

            Comment

            • firefly2280
              Member
              • Oct 2012
              • 28

              #7
              thanks for the input. Sadly Illumina are not going to replace the kits and I cant afford to trial and error things, so I dont really know what to do now. The only thing any of us can think of for failure of my library is some kind of inhibition...

              Comment

              • AStretton
                Member
                • Jul 2012
                • 25

                #8
                I'm surprised that Illumina are not going to replace the kit considering you had 10% PhiX in with your sample. The PhiX should have clustered even if your library didn't, I would argue that indicates a reagent or flow cell problem.

                Comment

                • mcnelson.phd
                  Senior Member
                  • Jul 2011
                  • 162

                  #9
                  I agree, Illumina should be replacing those reagents. Every time we've had tat object reference error Illumina has replaced the kits. I would pretty much demand that they send replacements unless they can prove that this was your fault, which given the nature of the error they most likely can't.

                  Comment

                  • microgirl123
                    Senior Member
                    • Jun 2012
                    • 199

                    #10
                    How do you think inhibitors could have gotten into your run? From your original sample DNA? I would think that over the library creation process, you would have done enough clean up steps to get rid of anything!

                    Comment

                    • rwinegar
                      Junior Member
                      • Nov 2009
                      • 6

                      #11
                      Unfortunately, we have encountered some similar situations recently where there seems to be inhibitory factors in our samples. We always get terrific results from pure samples isolated with DNeasy, but for environmental samples isolated with MOBio kits, we usually have to perform a follow-up MinElute clean-up. Since the MinElute produces losses and cuts out high MW DNA we are currently evaluating other alternatives. The insight we've gained is that removal of PCR inhibitors does not mean you won't have inhibition of Nextera tagmentation.

                      Comment

                      • fozrun
                        Junior Member
                        • Jan 2010
                        • 6

                        #12
                        Did you put the custom primers into the appropriate wells with the Illumina primers and make sure you didn't indicate in your worksheet that you used custom primers?

                        Comment

                        • firefly2280
                          Member
                          • Oct 2012
                          • 28

                          #13
                          Originally posted by microgirl123 View Post
                          How do you think inhibitors could have gotten into your run? From your original sample DNA? I would think that over the library creation process, you would have done enough clean up steps to get rid of anything!
                          Thats what I think too but maybe its something that associates with the DNA and co-purifies?

                          Comment

                          • firefly2280
                            Member
                            • Oct 2012
                            • 28

                            #14
                            Originally posted by rwinegar View Post
                            Unfortunately, we have encountered some similar situations recently where there seems to be inhibitory factors in our samples. We always get terrific results from pure samples isolated with DNeasy, but for environmental samples isolated with MOBio kits, we usually have to perform a follow-up MinElute clean-up. Since the MinElute produces losses and cuts out high MW DNA we are currently evaluating other alternatives. The insight we've gained is that removal of PCR inhibitors does not mean you won't have inhibition of Nextera tagmentation.
                            Thanks for this. Im not using Nextera but I suppose its 'good' to know that this is what might be causing a problem!

                            Comment

                            • firefly2280
                              Member
                              • Oct 2012
                              • 28

                              #15
                              I ran some samples on a collegues run which worked but my samples only produced 4, 10 and 43 reads...!! so something is for sure going on!

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                Today, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              22 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              15 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              68 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              48 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...