    Failed Illumina lane

    Hi everyone
    Our group has recently sequenced 8 RADtag libraries on Illumina HiSeq 2500.
    7 of the lanes were very good but one of them was a complete failure. Even though the cluster density was good (~650k/mm2), most of them failed the filter. The 47 million reads we recieved were of very bad quality: 60% of them had an uncalled base (N) at the 7th position and after ~40 bp the quality drops below Phred score 10.
    The library had a good concentration and fragment length distribution.
    We have no clue what may have gone wrong and we would be very happy for any hint. I have attached some more information, FastQC and SAV plots.
    Thanks a lot in advance for any help!
    Best,
    Joana
    Attached Files
  • kmcarr
    Senior Member
    • May 2008
    • 1181

    #2
    I would suspect a fluidics problem (e.g. clog or bubble) in lane 2 during cycle 7.

    Comment

    • SNPsaurus
      Registered Vendor
      • May 2013
      • 525

      #3
      Just so I understand, the run wasn't done in rapid mode, but the traditional mode with 8 lanes? And it didn't have a control lane with phiX, but rather phiX was spiked into each lane? Were different libraries loaded into each lane?

      It very well could have been a fluidic problem, but the fact that the problem first appeared at the invariant nucleotide of the restriction site makes complexity of the read another possibility. Or a combination of the two. I was surprised that the other lanes with cluster densities approaching 1000 did as well as they did. We (in my academic side) found it worthwhile to aim a bit lower than suggested with RAD libraries. I'm also surprised using all 8 lanes without a control lane worked out as well as it did. Sorry to not have any real advice!
      Providing nextRAD genotyping and PacBio sequencing services. http://snpsaurus.com

      Comment

      • JoanaIEE
        Junior Member
        • Oct 2011
        • 6

        #4
        Hi SNPsaurus and kmcarr

        Thanks a lot for your comments.

        Originally posted by SNPsaurus View Post
        Just so I understand, the run wasn't done in rapid mode, but the traditional mode with 8 lanes? And it didn't have a control lane with phiX, but rather phiX was spiked into each lane? Were different libraries loaded into each lane?
        Yes, it was a traditional mode run with 8 lanes. Each library containing several barcodes was loaded on one lane and spiked with 20% PhiX.

        It very well could have been a fluidic problem, but the fact that the problem first appeared at the invariant nucleotide of the restriction site makes complexity of the read another possibility. Or a combination of the two. I was surprised that the other lanes with cluster densities approaching 1000 did as well as they did. We (in my academic side) found it worthwhile to aim a bit lower than suggested with RAD libraries. I'm also surprised using all 8 lanes without a control lane worked out as well as it did. Sorry to not have any real advice!
        Yes, it does not seem like a coincidence that the problematic first nucleotide is an invariant one in the library. But all other libraries also had the same base at position 7 and it was no problem for the other libraries even though most of them had a higher cluster density. At another sequencing center they also successfully use 20% PhiX for RAD libraries, that's why this amount was chosen.

        Comment

        • TonyBrooks
          Senior Member
          • Jun 2009
          • 303

          #5
          Filtering is done using data from the first 25 bases. One failed test is allowed for every cluster. If the same cluster fails a test across two different cycles, then the cluster fails filter. If you get BMS (bubble causing mis-focus across a swath) within the first 25 cycles this causes problems as every cluster will fail tests. It means you are inherently at risk of losing clusters to the filter. Can you post the thumbnails from the dodgy cycle (7).

          Comment

          • HGalvez
            Junior Member
            • Sep 2013
            • 1

            #6
            Had a similar problem here and we attributed to a BMS failure in the first cycles. As other people mentioned in this thread, it seems there was a problem in cycle 7 for lane 2. You should be able to see what it was in the SAV. If you saved the thumbnail pictures the problem should be even more evident.

            BMS are a major source of headaches for anyone working with the HiSeq platform and I can tell you from experience that the 2500 seems more likely to fail for some reason than the 2000. If you call illumina to tell them about your problem, they will tell you that you should change gaskets often, but it doesn't seem to be failsafe.

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
              by SEQadmin2



              NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection of antimicrobial...
              Today, 06:46 AM
            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-28-2026, 09:57 AM
            0 responses
            22 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-28-2026, 09:51 AM
            0 responses
            14 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            67 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            47 views
            0 reactions
            Last Post SEQadmin2  
            Working...