  • dsobral
    Member
    • Jan 2012
    • 23

    #1

    MiSeq fastq output: 250-251 bp reads

    Hello,

    I'm getting reads from a MiSeq machine and noticing that many have 251bp instead of 250bp, and that the last base has a highly skewed base composition. From what I read of the manual, the sequencer always make read_length + 1 cycles of imaging, but only read_length are analyzed (for phasing etc...). Shouldn't the final fastq have all reads the same size and of only 250bp?? Is it ok to cut out the final base or is it a sign that there's something wrong?

    Also, a side note: some reads have less than 250bp, and I assume it is because of adapter trimming. If this is the case, then both read1 and read2 of a pair should be trimmed to the same size, otherwise it shouldn't be trusted, no?

    Thanks,
    Daniel
  • mcnelson.phd
    Senior Member
    • Jul 2011
    • 162

    #2
    Hi Daniel,

    You're correct in that the actual read length of an Illumina run is always N+1 because the extra base is used for phasing/pre-phasing analysis. Ideally that last base should be trimmed off because it's not properly quality checked.

    As for reads < 250bp, if you're using a Nextera kit and had Trim Adapters checked in the sample sheet, then you're correct about why you have shorter reads. If you're seeing that the two reads of a pair aren't the same length, then you're also probably seeing that read 1 is shorter than read 2. This would be an issue with the trimming where the base quality of read 2 dropped low enough that the adapter sequence wasn't properly called and thus couldn't be recognized to be trimmed. Some third-party apps can do a much better job of trimming so you may want to try those.

    Comment

    • dsobral
      Member
      • Jan 2012
      • 23

      #3
      Thanks for the reply. Very useful information.

      The only thing I'm still puzzled is why some reads have 250bp and others have 251bp.

      Daniel

      Comment

      • GenoMax
        Senior Member
        • Feb 2008
        • 7142

        #4
        Some facilities set up a run as (n+1) depending on the number of bases (n) you had asked to be sequenced.

        If you did not set this run up yourself then it is possible that the original run was set up as 250 x 251 bp (if one read is consistently 250 or less and other is 251 bp or less depending on trimming).

        Comment

        • dsobral
          Member
          • Jan 2012
          • 23

          #5
          I would understand if there was some obvious consistency.
          What I observe is that for the same run, read1 OR read2 can be either 250 or 251bp (and sometimes 249bp!) with no apparently consistent pattern. I'm suspicious that the behaviour is coming from adapter trimming.

          Counts | Read1 | Read2
          4223 | 250 | 248
          7940 | 250 | 249
          58517 | 250 | 250
          130842 | 250 | 251
          10571 | 251 | 248
          21321 | 251 | 249
          145959 | 251 | 250
          331396 | 251 | 251
          ...

          Comment

          • GenoMax
            Senior Member
            • Feb 2008
            • 7142

            #6
            Adapter trimming can't be the cause of it (unless this was set up as a longer run originally than 250 bp).

            Did you run this yourself (if not you should ask the facility that ran it to see how the original run was set up).

            Comment

            • dsobral
              Member
              • Jan 2012
              • 23

              #7
              I didn't run it myself, but it was using Nextera V2 250x250
              Adaptor trimming was on (I guess by default)

              Thanks,
              Daniel
              Last edited by dsobral; 12-10-2013, 06:16 AM.

              Comment

              • dsobral
                Member
                • Jan 2012
                • 23

                #8
                PS: although the data has these peculiarities, I used these for denovo assembly of a bacteria, and it gave good results...

                I just noticed because when I tried Edena on the full data, it complained about the sizes...

                I was just wandering what to think of it.

                Thanks

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                  by SEQadmin2




                  NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                  Yesterday, 06:46 AM
                • SEQadmin2
                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                  by SEQadmin2



                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                  ...
                  10-14-2026, 09:55 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, 10-28-2026, 09:57 AM
                0 responses
                29 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-28-2026, 09:51 AM
                0 responses
                18 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-22-2026, 11:13 AM
                0 responses
                71 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-22-2026, 10:48 AM
                0 responses
                48 views
                0 reactions
                Last Post SEQadmin2  
                Working...