    Problems with kit 600 cycles

    Hi,
    We were using kit 600 cycles v3 for library made in truseq RNA in Miseq, but the quality of the results was bad in the ends and the length of the reads was 240 pb in average. We saw in illumina that the quality was good in comparison with 500 cycles.
    Maybe the problem is that this kit isn’t compatible with truseq library or this kit is exclusive with nextera library, we don’t know this.
    Help me please.
    Thanks!
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #2
    The problem here may not be the kit but your library. It is possible that inserts in your library are shorter than you expect them to be.

    I assume "average length of reads 240 bp" is after trimming?

    Comment

    • Mario Moreno
      Junior Member
      • Nov 2013
      • 7

      #3
      Hi,
      Yes, the "average length of reads 240 bp" is after trimming.
      you know if the kits is compatible with truseq library or this kit is exclusive with nextera library?

      Thanks

      Comment

      • microgirl123
        Senior Member
        • Jun 2012
        • 199

        #4
        The 600-cycle kit is compatible with TruSeq - we've had great results with it on TruSeq libraries. What does your Bioanalyzer trace look like?

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          Which TruSeq library kit?
          With minor modifications, a 600 cycle kit would work for most TruSeq libraries. For some values of "minor"...

          --
          Phillip

          Comment

          • Mario Moreno
            Junior Member
            • Nov 2013
            • 7

            #6
            Thanks you for your prompt replies.
            I am wondering if you use the same protocol or perform variations in the step of fragmentation of the mRNA ( Elution 2-Frag-Prime: 94°C for 5 min.) or in other step.
            Thanks
            Last edited by Mario Moreno; 05-05-2014, 07:02 AM.

            Comment

            • pmiguel
              Senior Member
              • Aug 2008
              • 2328

              #7
              Originally posted by Mario Moreno View Post
              Thanks you for your prompt replies.
              I am wondering if you use the same protocol or perform variations in the step of fragmentation of the mRNA ( Elution 2-Frag-Prime: 94°C for 5 min.) or in other step.
              Thanks
              For RNA we use 1-2 minutes instead of 8 minutes for fragmentation. Also we use lower ratios of sample:ampure for clean-ups. But to figure the ratios out we calibrate each Ampure bottle. (We don't a gel cut purification of samples, if you do a gel cut, you might not need to modify the Ampure amounts.)

              But the trick is, you have to do both -- fragment shorter time and use less ampure.

              You can tell how you did by running the final, amplified library on a bioanalyzer chip.

              Look at the left-hand side of the peak.

              The very left-most size will seed the majority of your clusters. So if you see anything at all at 370bp or less, then you will get lots of reads that contain adapter and/or read all the way through into the flow cell oligo attachment. For best results you want the left side of your peak to be >720bases. That way your forward and reverse reads will both be providing novel sequence. But to get to 600 base inserts you may need to fragment less than 1 minute.

              Obviously, if you remove the shorter insert amplicons, then you bias the library away from short (>500 nucleotide) transcripts. So there is a cost...

              --
              Phillip
              Last edited by pmiguel; 05-05-2014, 07:42 AM.

              Comment

              • DJ Flowcell
                Member
                • Apr 2013
                • 13

                #8
                Hi Mario,

                Yeah you would definitely run into problems with such short libraries. As Phillip points out, if you are running 2x300bp runs, you will need your average total library size to be at least 420bp - 300bp for the actual library and 120bp for the adapters (60bp each). Otherwise, you will run into adapter sequences.

                Comment

                • Mario Moreno
                  Junior Member
                  • Nov 2013
                  • 7

                  #9
                  Thanks you for your prompt replies.
                  It has helped me a great help.
                  Soon will perform the modifications to make the libraries with larger means.
                  When you have the results back to comment
                  Thanks!

                  Comment

                  • snetmcom
                    Senior Member
                    • Oct 2008
                    • 159

                    #10
                    Originally posted by pmiguel View Post
                    For RNA we use 1-2 minutes instead of 8 minutes for fragmentation. Also we use lower ratios of sample:ampure for clean-ups. But to figure the ratios out we calibrate each Ampure bottle. (We don't a gel cut purification of samples, if you do a gel cut, you might not need to modify the Ampure amounts.)

                    But the trick is, you have to do both -- fragment shorter time and use less ampure.

                    You can tell how you did by running the final, amplified library on a bioanalyzer chip.

                    Look at the left-hand side of the peak.

                    The very left-most size will seed the majority of your clusters. So if you see anything at all at 370bp or less, then you will get lots of reads that contain adapter and/or read all the way through into the flow cell oligo attachment. For best results you want the left side of your peak to be >720bases. That way your forward and reverse reads will both be providing novel sequence. But to get to 600 base inserts you may need to fragment less than 1 minute.

                    Obviously, if you remove the shorter insert amplicons, then you bias the library away from short (>500 nucleotide) transcripts. So there is a cost...

                    --
                    Phillip
                    very helpful. I also recommend shorter fragmentation times.
                    I am curious what Ampure ratios do you typically test.

                    Comment

                    • nucacidhunter
                      Jafar Jabbari
                      • Jan 2013
                      • 1250

                      #11
                      In my experience size cut will be around 160 bp and 100 bp for 1x and 1.2x, respectively. It is easy to test the size cut without a library if you have a Covaris. Shear some DNA and size select (left size selection) 100 ng of shear with various ratios of beads and after elution, run size selected sample and input shear on TapeStation or Bioanlyzer.
                      I have not seen differences with different batches of AMPure beads in practice to justify calibration. I have only read about it in 454 DNA library prep kit manual which recommends titration with DNA ladder. Sheared DNA is much better than ladder for calibration to see the exact size cut because fragment size distribution is continuous like the libraries. If you are worried about titration, the alternative is SPRIselect which has already been factory calibrated and it only cost little more.
                      Last edited by nucacidhunter; 05-10-2014, 01:44 AM.

                      Comment

                      • JBKri
                        Member
                        • Jan 2014
                        • 80

                        #12
                        Originally posted by DJ Flowcell View Post
                        Hi Mario,

                        Yeah you would definitely run into problems with such short libraries. As Phillip points out, if you are running 2x300bp runs, you will need your average total library size to be at least 420bp - 300bp for the actual library and 120bp for the adapters (60bp each). Otherwise, you will run into adapter sequences.
                        If we do run a library with a short insert (~260 bp) with a 2x300 cycles kit, would it cause any problems other than reading into the adapters? Would it affect the quality of the index read, or read2?

                        Jon

                        Comment

                        • nucacidhunter
                          Jafar Jabbari
                          • Jan 2013
                          • 1250

                          #13
                          Originally posted by JBKri View Post
                          If we do run a library with a short insert (~260 bp) with a 2x300 cycles kit, would it cause any problems other than reading into the adapters? Would it affect the quality of the index read, or read2?

                          Jon
                          No, it won't affect any of those.

                          Comment

                          • nbogard
                            Member
                            • Apr 2011
                            • 10

                            #14
                            What if the number of cycles is longer than the insert + adapter + flow cell oligo?

                            Comment

                            • nucacidhunter
                              Jafar Jabbari
                              • Jan 2013
                              • 1250

                              #15
                              The overall quality displayed in SAV will be very low. FastQC plots also will show that quality is dropping as the cycle number increases. The quality of insert or index reads will not be affected and should be high after trimming adapters.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM
                              • SEQadmin2
                                New Genomics Technologies Take Aim at Long-Standing Limits
                                by SEQadmin2


                                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                ...
                                09-28-2026, 10:25 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              23 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              20 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-15-2026, 10:52 AM
                              0 responses
                              36 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-14-2026, 05:39 AM
                              0 responses
                              42 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...