    replied
    Two main reasons for overclustering:
    1- Underestimating library concentration
    2- pipetting errors

    Other causes for low PF :
    1- using non-optimal custom primers either by design or synthesis inefficiencies
    2- low diversity library

    Leave a comment:


  • GA-J
    replied
    @nucacidhunter: Thank you for your opinion. I am not able to post a screenshot now. I could try on next Monday. But there is no reason to be overloaded. Let's try to dig more.


    Originally posted by nucacidhunter View Post
    With your description it looks like overclustering and consequent low PF. Could you post a screenshot of flow cell chart (from run SAV) by selecting cluster from the top tab.

    Leave a comment:


  • nucacidhunter
    replied
    With your description it looks like overclustering and consequent low PF. Could you post a screenshot of flow cell chart (from run SAV) by selecting cluster from the top tab.

    Leave a comment:


  • GA-J
    replied
    Recently I had a bad run with overcluster density(over 1100 for V2), but we saw very very low passing filter and very low phix. For troubleshooting, we checked everything, libraries no problem, prep steps no problem. I just recall I left cartridge at bench for a while after it thawed. Could that be problem? No amplify but over 1000 cluster density? Anybody left cartridge at bench for how long still got good run?

    Help needed, any hint? Thanks.
    Last edited by GA-J; 09-25-2015, 04:55 PM.

    Leave a comment:


  • fanli
    replied
    1. We do 2x250 for the V4 region and routinely get 90+% pairs that can be joined together.

    2. 30% is overkill, 10% is probably fine like microgirl123 said. We honestly do even less (5% or none) because the MiSeq chemistry is pretty good with low diversity samples.

    3/4. You get some variation in #reads/sample. Fortunately, 16S tends to saturate very quickly (think 10-20k reads/sample), so even 100K with a broad distribution should cover most of your samples.

    Leave a comment:


  • kmcarr
    replied
    Originally posted by marcpavi View Post
    3-As the number of samples is quite large, we are not sure if we’ll be able to quantify all samples through qPCR. Does a QUBIT based normalization of individual amplicons and a qPCR based normalization of pooled amplicons work? How badly could this bias our run?
    At first we tried to quantify and normalize by hand every PCR sample; quickly realized this too much work. Now after PCR, either one step for 16S or two step for everything else, all reactions are normalized using Invitrogen (now Life Tech) SequalPrep Normalization plates. Very easy and fast. Is it perfect? No, it is definitely good enough to get decent representation of all sample in the pool.

    Leave a comment:


  • microgirl123
    replied
    1. A 460 bp amplicon with a 500-cycle kit is probably not enough overlap. I know Illumina claims a 50-bp overlap is enough, but all my users like a much greater overlap.

    2. I use a 10% phiX spike for amplicon runs and aim for ~800K clusters. This seems to generate good data.

    3. If you are using the 2-step PCR approach to generate your amplicons (or any approach where PCR, not ligation, is your last step, you should be able to quantify and pool based on Qubit readings and then just quantify your pooled library with qPCR. This is what I do and it works well - not perfectly but you are quantifying and pooling small volumes of sample (even with qPCR) so there's a fair amount of pipetting error introduced. Make sure you convert the ng/ul reading spit out by the Qubit to nM so that you are pooling equimolar amounts of DNA.

    4. If you do run a V2 kit, you should end up with ~100K reads per sample. You'd get more reads with a V3 kit.

    Leave a comment:


  • marcpavi
    started a topic 16S Miseq run with 96 indexed samples

    16S Miseq run with 96 indexed samples

    Hi everyone,
    First post here, sorry for the bad writing.
    Me and my colleagues are planning to do a Miseq run to examine the 16S RNA of several bacterial samples.
    We would like to sequence as many samples as possible in one run, and we are thinking to use the 96 index nextera kit coupled with de v2 chemistry to sequence the V3-V4 region.
    Our concerns are about 4 points:

    1-Combining 2x250 PE and V3-V4 (460bp) works allright?
    2-How much PhiX we should use to avoid overcluster? We were told that 30% should be enough.
    3-As the number of samples is quite large, we are not sure if we’ll be able to quantify all samples through qPCR. Does a QUBIT based normalization of individual amplicons and a qPCR based normalization of pooled amplicons work? How badly could this bias our run?
    4-If someone usually does this run, how many reads per sample are expected to be yielded?

    Any help would be appreciated.
    Cheers

Latest Articles

Collapse

  • seqadmin
    Non-Coding RNA Research and Technologies
    by seqadmin




    Non-coding RNAs (ncRNAs) do not code for proteins but play important roles in numerous cellular processes including gene silencing, developmental pathways, and more. There are numerous types including microRNA (miRNA), long ncRNA (lncRNA), circular RNA (circRNA), and more. In this article, we discuss innovative ncRNA research and explore recent technological advancements that improve the study of ncRNAs.

    Nobel Prize for MicroRNA Discovery
    This week,...
    10-07-2024, 08:07 AM
  • seqadmin
    Recent Developments in Metagenomics
    by seqadmin





    Metagenomics has improved the way researchers study microorganisms across diverse environments. Historically, studying microorganisms relied on culturing them in the lab, a method that limits the investigation of many species since most are unculturable1. Metagenomics overcomes these issues by allowing the study of microorganisms regardless of their ability to be cultured or the environments they inhabit. Over time, the field has evolved, especially with the advent...
    09-23-2024, 06:35 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by seqadmin, Today, 06:55 AM
0 responses
8 views
0 likes
Last Post seqadmin  
Started by seqadmin, 10-02-2024, 04:51 AM
0 responses
105 views
0 likes
Last Post seqadmin  
Started by seqadmin, 10-01-2024, 07:10 AM
0 responses
114 views
0 likes
Last Post seqadmin  
Started by seqadmin, 09-30-2024, 08:33 AM
1 response
117 views
0 likes
Last Post EmiTom
by EmiTom
 
Working...
X