    MiSeq: no data for A/C but G/T look OK?

    hi all,

    i'm wondering if anyone else has seen this kind of data in their thumbnails before and if so what the problem turned out to be. it was sequenced with a v3 150-cycle kit; the first MiSeq we tried has a chronic pumping problem that illumina can't seem to repair and i ultimately drove the flow cell and reagent box to another instrument nearby that wasn't already doing a run.

    thanks,
    steve
    Attached Files
  • dannyhi321
    Member
    • Jul 2014
    • 51

    #2
    Originally posted by sdbrown View Post
    hi all,

    i'm wondering if anyone else has seen this kind of data in their thumbnails before and if so what the problem turned out to be. it was sequenced with a v3 150-cycle kit; the first MiSeq we tried has a chronic pumping problem that illumina can't seem to repair and i ultimately drove the flow cell and reagent box to another instrument nearby that wasn't already doing a run.

    thanks,
    steve
    Looks like a focusing/imaging problem - did you get any usable data?

    Comment

    • sdbrown
      Junior Member
      • Feb 2014
      • 5

      #3
      Originally posted by dannyhi321 View Post
      Looks like a focusing/imaging problem - did you get any usable data?
      hi, thanks for replying.

      no usable data. there were a couple of tiles where A & C images looked OK too but of the ~9.9M reads reported by the SAV software there were zero reads passing the filter. i converted the .bcl files to .fastq files with bcl2fastq's --with-failed-reads option but still nothing.

      edit: sorry, i should have said that the --with-failed-reads option produced FASTQ files with lots of reads composed entirely of 'N'.
      Last edited by sdbrown; 03-08-2016, 08:12 AM. Reason: clarified what "nothing" meant.

      Comment

      • GenoMax
        Senior Member
        • Feb 2008
        • 7142

        #4
        In cases like this the first thing you should do is contact Illumina tech support. Have you done that? If they determine this to be a hardware/reagent failure then you could get a free replacement kit.

        That said I don't remember seeing anything similar to images you have posted above. I wonder if one of the lasers failed on this sequencer.
        Last edited by GenoMax; 03-08-2016, 08:20 AM.

        Comment

        • sdbrown
          Junior Member
          • Feb 2014
          • 5

          #5
          Originally posted by GenoMax View Post
          In cases like this the first thing you should do is contact Illumina tech support. Have you done that? If they determine this to be a hardware/reagent failure then you could get a free replacement kit.

          That said I don't remember seeing anything similar to images you have posted above. I wonder if one of the lasers failed on this sequencer.
          that's exactly what illumina tech support said; A & C apparently share a (red?) excitation laser. i'm currently going through the back-and-forth with them but since the instrument is not connected to the internet, it's not as quick as it might otherwise be.

          Comment

          • GenoMax
            Senior Member
            • Feb 2008
            • 7142

            #6
            That makes sense. As long as the instrument is under a maintenance contract an engineer visit should fix that problem and you will get the free replacement kit.

            Comment

            • sdbrown
              Junior Member
              • Feb 2014
              • 5

              #7
              is there a website somewhere that has an archive of instrument problems and the (diagnostic) data associated with those problems?

              edit: in hindsight, perhaps seqanswers is that website...
              Last edited by sdbrown; 03-08-2016, 09:07 AM.

              Comment

              • Jessica_L
                Senior Member
                • Feb 2010
                • 117

                #8
                Illumina has previously hosted user group meetings with sessions for run monitoring and troubleshooting that discussed some of the diagnostic information you'd be interested in for run/hardware failures, but I don't know that there's an archive of any of that.

                One thing I've noticed is that the grainy lines/static you see in the A/C images are some kind of an artifact exclusive to thumbnail images-- on older instruments (like GAs) where you could manually control lasers and nucleotide filters, if you went in and took some focus shots, you'd see a completely blank image, but when looking at the jpeg thumbnails they'd have a very static-y kind of look to them. I had asked Illumina at one point if it was related to the jpeg compression of the original tiff image file, but no one seemed to have an answer for that.

                Comment

                • GenoMax
                  Senior Member
                  • Feb 2008
                  • 7142

                  #9
                  @Jessica_L: If I remember it right these images are saved with a non-standard bit depth and require a capable viewing program (e.g. Irfanview) that can decode them correctly.

                  Comment

                  • Jessica_L
                    Senior Member
                    • Feb 2010
                    • 117

                    #10
                    GenoMax-- the bit depth would make sense. But the jpegs from blank images have always looked strange, whether they were viewed in SAV or IrfanView-- I actually still have IrfanView on my computer, but maybe there's a setting I'm missing in order for the image to be displayed accurately.

                    Comment

                    • mgangel
                      Junior Member
                      • Apr 2015
                      • 2

                      #11
                      Dead red module

                      Just had the same problem. After a month of troubleshooting with tech support, we found that the red LED module was dead. I had runs fail at cycles 1, 16, 32, 176, and then I finally saw it die completely on the last run. They first blamed them software and downgraded to the previous version. They then replaced the hard drive. Finally they replaced the LED module for the A/C channel.

                      Comment

                      Latest Articles

                      Collapse

                      • SEQadmin2
                        16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                        by SEQadmin2




                        NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                        Today, 06:46 AM
                      • SEQadmin2
                        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                        by SEQadmin2



                        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                        ...
                        10-14-2026, 09:55 AM

                      ad_right_rmr

                      Collapse

                      News

                      Collapse

                      Topics Statistics Last Post
                      Started by SEQadmin2, 10-28-2026, 09:57 AM
                      0 responses
                      23 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-28-2026, 09:51 AM
                      0 responses
                      15 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-22-2026, 11:13 AM
                      0 responses
                      68 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-22-2026, 10:48 AM
                      0 responses
                      48 views
                      0 reactions
                      Last Post SEQadmin2  
                      Working...