Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • santorini
    Junior Member
    • Oct 2015
    • 8

    #1

    Liquid Handling Instruments

    Hello everyone,

    First time posting here in a long time...good to be back!

    My lab currently uses three of the Agilent Bravo option A instruments for our SureSelect capture protocols. From what I've read online, they're supposed to be very accurate and reliable instruments. However our experience hasn't been very positive, and for the 6 months we've had them, we've had at least one experiencing some kind of problem.

    Pros - 1-96 pipetting channel capability, good at bead purification, very tough/robust instrument, pipetting is accurate when working.

    Cons - Very small deck and limitations where the head can go/pipette from, pipetting accuracy can vary wildly causing constant troubleshooting to find what's wrong, only 1 service technician for whole US so bad tech support, software/programming extremely not user friendly.

    Essentially, these instruments are not going to fit with us long term and we are looking into other liquid handlers.

    So far on my list are: VERSA 1100, Hamilton NIMBUS/STAR, Perkin JANUS/Sciclone/Zephyr, and Tecan Fluent or Freedom Evo. I know that's a big list!

    We are already using the Eppendorf EpMotion for our library prep and love it, but it doesn't have the 96 well capability.

    Does anyone have any input on any these liquid handlers? Know of any others I'm leaving out? Our lab is rapidly growing/evolving, so we are looking for an instrument we can grow into that's versatile and relatively easy to customize and change protocols.

    Thanks!
  • luc
    Senior Member
    • Dec 2010
    • 469

    #2
    Hi Santorini,

    we have the Sciclone and "software/programming not user friendly" would certainly apply there, too. The good thing is that the Sciclone is supported by Kapa, Biooscientific, and Nugen. However, perhaps due to the programming difficulties, it even takes companies a long time to get the programs right.

    Which EpMotion do you have and what kind of library preps are you running?

    Comment

    • santorini
      Junior Member
      • Oct 2015
      • 8

      #3
      Thanks for the reply!

      We are using the EpMotion 5075 tc (thermomixer+cleancap) and are using NEBnext reagents for Library Prep and Ampure bead purifications.

      Since you picked the Sciclone, would you be able to tell me any differences between that and the Zephyr or JANUS? Any reason for choosing one over the others?

      Comment

      • luc
        Senior Member
        • Dec 2010
        • 469

        #4
        Hi,

        the Sciclone NGS was purchased before I joined the lab.

        I assume that Zephyr and Sciclone are the instruments with the 96-channel pipette heads; of which the Sciclones are the ones with the larger deck.
        The Janus works with 4 or 8 tip Varispan heads only.

        Did you write the scripts for the NEBNext kits on the EpMotion yourself?
        (DNA and RNA-seq kits? Do your scripts accommodate varying sampel numbers?)

        Comment

        • ScienceGrrl
          Junior Member
          • Jul 2015
          • 8

          #5
          Hello,

          I am the Scientific Leader, NGS for Hamilton Robotics. If you would like to speak with me before jumping into getting sales reps involved, feel free to contact me here or I can give you my contact info.

          Comment

          • Nighthawkrao77
            Member
            • Jan 2015
            • 11

            #6
            Look out for the BD GenCell CLiC!

            Comment

            • DRYTCYV
              Member
              • Apr 2011
              • 50

              #7
              Hello everyone,

              Does anyone have more information to update this post?

              We're in the middle of the same process to choose a Liquid Handling Robot, and our list is:

              - Hamilton NGS

              - Eppendorf 5075

              - Perkin Elmer Sciclone

              Pro and Cons ? Any other options that we should check?

              Thanks

              Comment

              • thermophile
                Senior Member
                • Apr 2015
                • 243

                #8
                I have the epmotion 5075 and the qiagility. They are obviously very different instruments. Our epmotion has the tmx and vac-but the vac hardly ever works, some clients have used it but for our lib prep we are all bead based. The software for the epmotion is frustrating, all point and click, but you can control everything that it does once you figure out how the software thinks. Expect to fine tune every plasticware that you use with the beads (adjust height, asp. speed, etc). The labware from eppendorf is just the starting point if you want to maximize your yields.

                The qiagility is made to be a pcr/qpcr robot. If you can figure out how to translate your task into a pcr like task it will do it, if not too bad. I wouldn't recommend that instrument for anything but PCR (we use it for library normalization and pooling)
                Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

                Comment

                • jteeee2
                  Member
                  • Oct 2014
                  • 42

                  #9
                  We have both an Eppendorf 5075 TMX and a Perkin Elmer Sciclone. In terms of NGS capability, the Sciclone is far superior to the Eppendorf 5075. As thermophile mentioned, the real headache is fine tuning all of the plasticware. Even when you get that worked out, the instrument is not as good as the Sciclone in when doing the steps required for library prep (e.g. bead cleanups). We use the Sciclone to run several different library prep workflows (up to 96 libraries generated at one time). The Eppendorf we use for simple dilution and pooling of libraries. One downside to the Sciclone is that unless you know programming language, it's much more difficult for the user to make small changes without having an engineer from Perkin Elmer come out and do it. With the Eppendorf, making tweaks to protocols is very straightforward for the average user.

                  Comment

                  • santorini
                    Junior Member
                    • Oct 2015
                    • 8

                    #10
                    I'll update with some info we've since learned about some of these instruments. We haven't decided on one yet either, so more feedback is appreciated!

                    - Epmotion 5075 - Agree with everything thermophile said. Easy enough to program with drag and drop, but lots of work needed to optimize your protocols, especially with beads/magnets. Also only has single and 8 channel pipettes, does not have the capability for 96 channel head.

                    - Sciclone - Only is capable of pipetting 200uL at a time, so if you need large volumes this may not be your instrument. The Agilent Bravo is similar, where you might need to pipette something multiple times to get the right volume. We ruled this one out for that reason.

                    So far we're pretty much between the Hamilton STAR and the BioMek FxP. We also looked into the Lynx platform from Dynamic Devices, but we're not sure about their pipetting accuracy.

                    Comment

                    • DRYTCYV
                      Member
                      • Apr 2011
                      • 50

                      #11
                      Hi,

                      Our list just increase with 1 more brand, but we still have some questions/concerns that we're trying to get in contact with the tech support.

                      - Epmotion 5075 : In the beginning, they said that the software was very user friendly and the possibilities to customize your protocols. Then, was not so easy to do that.

                      - Perkin Elmer Sciclone: The most expensive equipment, and looks like we are always dependent from the tech support to design the protocols.
                      - Hamilton NGS Start : Still waiting for a reply to some questions: basic accessories, the price for the extra.

                      - TECAN Evo NGS: We received a good proposal for this equipment, we'll have a meeting with them next week
                      Any one already design their own protocol in any of this equipments? More information?

                      Cheers

                      Comment

                      • huguesparri
                        Member
                        • May 2008
                        • 97

                        #12
                        We use a Tecan evo to build RNAseq libraries and RADseq libraries and we are happy with it.
                        We often process 96 samples at the same time.

                        For the RAseq part, we ends up with less material than when we do it handmade so we need to start with more material. But sequencing results are similar.

                        I don't know for other liquid handling instruments and maybe it's due to our machine workbench configuration but having 1 person near the machine during nearly all the library construction process is mandatory.

                        Comment

                        • DRYTCYV
                          Member
                          • Apr 2011
                          • 50

                          #13
                          Originally posted by huguesparri View Post
                          We use a Tecan evo to build RNAseq libraries and RADseq libraries and we are happy with it.
                          We often process 96 samples at the same time.

                          For the RAseq part, we ends up with less material than when we do it handmade so we need to start with more material. But sequencing results are similar.

                          I don't know for other liquid handling instruments and maybe it's due to our machine workbench configuration but having 1 person near the machine during nearly all the library construction process is mandatory.
                          Thank you huguesparri for your reply, it was very useful. Do you use official protocols or you design your owns?

                          For RNAseq we are thinking to use this one : http://www.nature.com/nprot/journal/....2014.006.html

                          We want to put the robot in a room, more or less a clean room, just to prepare libraries, so no one will be there...should we install a camera ?

                          Cheers

                          Comment

                          • huguesparri
                            Member
                            • May 2008
                            • 97

                            #14
                            For the RNA part, we use the truseq RNA stranded sample prep kit and protocol.
                            For the RAD part, we use the Baird protocol. We just replaced the column purification steps with AMPure XP purification steps.

                            Basically, somebody is watching after the machine:
                            - to refill tips: we use a lot of them (and I mean a lot!).
                            - to empty the trash (due to tips).
                            - to seal plates: we didn't buy an automatic sealing machine as the price seemed to high regarding the benefits. It's still possible to automatize that step.
                            - to put plate in the PCR machine for incubations and PCR. Same as previous point: it's a choice we made but it can be automatized as well.
                            - and of course to be ready when something goes wrong: because Murphy's law is powerfull.

                            So if you can afford it, the machine can be self-sufficient and a camera may be enough (except for the last point ^^).

                            Here are what convinced us to use a TECAN EVO as Liquid Handling Instrument:
                            - the machine was available for free from another facility in our institute: I can't deny it's a BIG BIG pro-point
                            - we already had a TECAN Genesis so we knew how to program/manipulate such a device.
                            - programs are "easy" to create: if you change your protocols, you don't need to wait until the robot manufacturer implements it on your machine. You can do it yourself.

                            But as I didn't had the opportunity to manipulate another instrument, I can't tell you if it's better or worst than something else.
                            All I can say is that it's satisfying for our use.

                            Comment

                            • DRYTCYV
                              Member
                              • Apr 2011
                              • 50

                              #15
                              Hi everyone,

                              Originally posted by huguesparri View Post
                              For the RNA part, we use the truseq RNA stranded sample prep kit and protocol.
                              For the RAD part, we use the Baird protocol. We just replaced the column purification steps with AMPure XP purification steps.

                              Basically, somebody is watching after the machine:
                              - to refill tips: we use a lot of them (and I mean a lot!).
                              - to empty the trash (due to tips).
                              - to seal plates: we didn't buy an automatic sealing machine as the price seemed to high regarding the benefits. It's still possible to automatize that step.
                              - to put plate in the PCR machine for incubations and PCR. Same as previous point: it's a choice we made but it can be automatized as well.
                              - and of course to be ready when something goes wrong: because Murphy's law is powerfull.

                              So if you can afford it, the machine can be self-sufficient and a camera may be enough (except for the last point ^^).

                              Here are what convinced us to use a TECAN EVO as Liquid Handling Instrument:
                              - the machine was available for free from another facility in our institute: I can't deny it's a BIG BIG pro-point
                              - we already had a TECAN Genesis so we knew how to program/manipulate such a device.
                              - programs are "easy" to create: if you change your protocols, you don't need to wait until the robot manufacturer implements it on your machine. You can do it yourself.

                              But as I didn't had the opportunity to manipulate another instrument, I can't tell you if it's better or worst than something else.
                              All I can say is that it's satisfying for our use.
                              Thank you huguesparri for your input!

                              We had some meetings and my last informations are:

                              - Epmotion 5075 : The software looks easy to work, I thought it was more difficult. Huge error rate when pipetting 1ul (+- 15% in a special ISO).


                              - Perkin Elmer Sciclone: Still the most expensive equipment, and looks like we are always dependent from the tech support to design the protocols.

                              - Hamilton NGS Start : Still waiting for a reply to some questions: basic accessories, the price for the extra.

                              - TECAN Evo NGS:
                              • Very good impressions about everything.
                              • Completely customized and appear to be simple to design your own protocols.
                              • Like the Hamilton, it comes with the Dynamic Positioning System : All pipetting channels to move independently in the Y- and Z-axes.
                              • It's possible to upload files with all the details about how many ul to pipette, from where to where and you can give the concentration that you want to normalize your samples and the software/robot can do the rest

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                10-30-2026, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              40 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              24 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              73 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              48 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...