    Bridge amplification and fragment size

    Hi All,
    I'm new to this whole NGS business. I will be using an Illumina GAII. In a recent article from Bio Techniques, I read that longer (about 600bp) fragments can increase the uniformity of coverage across a targeted region. In the paper, amplicons were fragmented to 200bp prior to adaptor ligation and subsequence sequencing. Does anyone know why this would increase uniformity? Also, does anyone know if it is possible to successfully generate clusters during bridge amplification with fragments larger than 200-300 bp? And if so, how large?
  • ScottC
    Senior Member
    • Jan 2008
    • 244

    #2
    Illumina tends to recommend up to 600b or thereabouts.

    Comment

    • aperera
      Member
      • Jan 2008
      • 23

      #3
      At a recent Illumina user meeting, we were told the ideal insert size is 150-200bp.

      Comment

      • ScottC
        Senior Member
        • Jan 2008
        • 244

        #4
        That's possibly the ideal size, but it works with larger fragments than that.

        Comment

        • huguesparri
          Member
          • May 2008
          • 97

          #5
          I used PE fragment as high as 800 pb (including adaptors) and it did work fine.

          Comment

          • aperera
            Member
            • Jan 2008
            • 23

            #6
            That's right, it is the ideal. We have used larger sizes as well.

            Comment

            • dmborek
              Junior Member
              • Jun 2010
              • 5

              #7
              Are there any negatives of using longer inserts e.g. reduced number of clusters due to their larger size?
              Anyone used even longer than 800 bp for PE (not mate pairs)?
              Dominika Borek, Ph.D.
              UT Southwestern Medical Center at Dallas
              5323 Harry Hines Blvd.
              Dallas, TX 75390
              Tel. 214-645-6378
              Fax. 214-645-6453

              Comment

              • konrad98
                Member
                • Jan 2009
                • 17

                #8
                Beyond 800bp you start getting cluster sizes which are almost guaranteed to merge with each other during the course of a run. You can can get around this by lowering the cluster density to around 100-200k/mm^2

                Comment

                • kmcarr
                  Senior Member
                  • May 2008
                  • 1181

                  #9
                  Besides cluster size as konrad noted there is also an issue of cluster shape. Due to the randomness of the direction which the "bridge" may form relative to the anchor point of the strand, clusters may start out being somewhat misshapen, e.g. oblong, jagged, etc. It takes a certain number of amplification cycles until the shape of the cluster becomes round. This shape effect is exaggerated with longer inserts, and for very long (800bp) inserts you may need to increase the number of amplification cycles. Of course additional cycles will make the clusters even larger.

                  The shape of the cluster matters because the image analysis algorithms look for round spots only. Odd, misshapen spots are rejected. The size of the cluster may also have an effect on cluster identification. Spots which are too large may be rejected by the software as potential artifacts. You really should discuss this with your Illumina FAS.

                  Comment

                  • dorix
                    Member
                    • Apr 2012
                    • 13

                    #10
                    Hi,
                    I have libraries of insert size of about 600bp. I'm quantifing them by qPCR. Those are amplification free libraries, so I gess I may have same fragments wihout adapters and so on. Unfortunatelly comparing molarity from bioanalyzer and from qPCR are incosistent: qPCR shows two times lower amount. I'm wondering which is more likely: I have samples with not fully constructed fragments or amplification efficiency is that much lower with longer fragments. I'm affraid of overclustering, using results from qPCR, however I do not want to loose money by less density.

                    Does anyone has similar expirience?

                    Comment

                    • nahmed
                      Junior Member
                      • Oct 2016
                      • 3

                      #11
                      Time required to sequence one base

                      Hello,
                      I am Nauman Ahmed. I am a new new member of SEQanswers.com. I am a PhD student in TU Delft.
                      My question is general about the time to sequence one base in Illuminia SBS technology, but let me give an example: In the specification sheet of Illuminia HiSeq 4000 , the run time is 3.5 days at maximum, generating 2 X 150bp reads. Does it mean that to sequence one base (one cycle), it takes about 17 minutes?

                      Comment

                      • misterc
                        Member
                        • Jan 2016
                        • 46

                        #12
                        Hi Nauman,

                        On HiSeq 4000, its about 16 minutes per cycle but this is limited by the imaging time to excite the incorporated fluorophores and image them simultaneously. The chemistry time is about 4 minutes but since it runs 2 flow cells simultaneously, the cycle time is about 8 minutes per side. If you ran 1 flow cell by itself, then the entire run time would be reduced for 2x150 cycles as you'd then have a 12 minute cycle time instead of 16. Hope this helps.

                        Best,
                        Charles

                        Comment

                        • Jessica_L
                          Senior Member
                          • Feb 2010
                          • 117

                          #13
                          Originally posted by dorix View Post
                          Hi,
                          I have libraries of insert size of about 600bp. I'm quantifing them by qPCR. Those are amplification free libraries, so I gess I may have same fragments wihout adapters and so on. Unfortunatelly comparing molarity from bioanalyzer and from qPCR are incosistent: qPCR shows two times lower amount. I'm wondering which is more likely: I have samples with not fully constructed fragments or amplification efficiency is that much lower with longer fragments. I'm affraid of overclustering, using results from qPCR, however I do not want to loose money by less density.

                          Does anyone has similar expirience?
                          Hi dorix,

                          The issue you're having is, I think, rather common. It helps to look at the overall shape of the electropherogram in the bioanalyzer data and make sure that it looks like appropriate for your library type. Illumina manuals usually have examples of what good/passing bioanalyzer data looks like. See page 25 of the PCR-free manual for an example of what I mean.

                          The other thing to consider is the quality of the qPCR data. If the standard curve looks good and passes all the necessary quality criteria (i.e. difference in Ct values between data points on the standard curve) and the efficiency and R^2 values are within the acceptance criteria, then I'd probably lean toward using the qPCR data over the bioanalyzer data. As you've mentioned already, there's a good chance that some of the material in the bioanalyzer traces will have adapters in an orientation that will not lead to successful sequencing, so the qPCR data gives you a better idea of how much material will bridge amplify and produce clusters.

                          I hope that helps.

                          Comment

                          • nahmed
                            Junior Member
                            • Oct 2016
                            • 3

                            #14
                            Originally posted by misterc View Post
                            Hi Nauman,

                            On HiSeq 4000, its about 16 minutes per cycle but this is limited by the imaging time to excite the incorporated fluorophores and image them simultaneously. The chemistry time is about 4 minutes but since it runs 2 flow cells simultaneously, the cycle time is about 8 minutes per side. If you ran 1 flow cell by itself, then the entire run time would be reduced for 2x150 cycles as you'd then have a 12 minute cycle time instead of 16. Hope this helps.

                            Best,
                            Charles

                            Thank you for the reply . If I look at the specifications of HiSeq 2500, the sequence runtime in upto 6 days (in high output mode) which means that it has much longer cycle time as compared to HiSeq 4000. It generates smaller reads i.e. 2x125. Is it due to older technology or something else?
                            http://www.illumina.com/content/dam/...-portfolio.pdf

                            Comment

                            • dorix
                              Member
                              • Apr 2012
                              • 13

                              #15
                              Originally posted by Jessica_L View Post
                              Hi dorix,

                              The issue you're having is, I think, rather common. It helps to look at the overall shape of the electropherogram in the bioanalyzer data and make sure that it looks like appropriate for your library type. Illumina manuals usually have examples of what good/passing bioanalyzer data looks like. See page 25 of the PCR-free manual for an example of what I mean.

                              The other thing to consider is the quality of the qPCR data. If the standard curve looks good and passes all the necessary quality criteria (i.e. difference in Ct values between data points on the standard curve) and the efficiency and R^2 values are within the acceptance criteria, then I'd probably lean toward using the qPCR data over the bioanalyzer data. As you've mentioned already, there's a good chance that some of the material in the bioanalyzer traces will have adapters in an orientation that will not lead to successful sequencing, so the qPCR data gives you a better idea of how much material will bridge amplify and produce clusters.

                              I hope that helps.
                              Thank you for your response, Jessica. The shape seems fine to me, Gaussian type shifted towards higher lengths. I have also seen amplicon libraries without amplification, which has given me a sense of libraries construction yield (that it is not very efficient). Nevertheless, so happend I've tested other qPCR kit with shorter amplification steps and it has given me even lower results, with comparable for short libraries. That warned me even harder :/

                              I wander if anybody have an expieriece of overclustering with that kind of observation?

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM
                              • SEQadmin2
                                New Genomics Technologies Take Aim at Long-Standing Limits
                                by SEQadmin2


                                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                ...
                                09-28-2026, 10:25 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-15-2026, 10:52 AM
                              0 responses
                              13 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-14-2026, 05:39 AM
                              0 responses
                              21 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-13-2026, 09:56 AM
                              0 responses
                              19 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-09-2026, 11:14 AM
                              0 responses
                              27 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...