  • Low QC scores?

    I have just set up my very first sequencing experiment on the Next-Seq 500. It is now 45 cycles into PE sequencing, however it seems that my QC scores are a bit low. I only have a % >= Q30%: 25.2%

    I have around:
    280 million reads in the QC 10 bin.
    25 million in the QC 20 bin
    8 million in the QC 25 bin
    65 million in QC 30 bin
    44 million in QC 35 bin.

    Is this a bad result? Apologies for my stupidness, just was wondering whether data could be salvaged from this?

  • #2
    Yes, now it has finished and the QC is only 15.2.

    The machine also has a warning message on two of the lanes:

    camera disabled:failed to detect clusters


    Does this mean that the cameras are not working properly?

    Thanks

    Comment


    • #3
      Probably. You should contact Illumina tech support. They can take a look at your sequencer remotely to diagnose the problem.

      If your instrument is not directly accessible over the network then you will need to send them some diagnostic files.

      Comment


      • #4
        Thanks for your reply.

        Yes I will do that - unfortunately since its a weekend they cannot reply.

        I was thinking - if clusters are not formed in those two particular areas, would the cameras be disabled? My pooled library concentration was sufficient, so I did not anticipate any problems. Also, if the cluster density is low then I am not sure how QC would be affected..

        Many thanks
        Last edited by ox_zoo; 11-19-2016, 07:39 AM.

        Comment


        • #5
          What was the cluster density for this run?

          If the run was borderline overclustered to begin with, it is possible that sequencer lost the ability to distinguish the clusters over time (they become fatter). There is also the possibility of this being some sort of a hardware issue.

          Comment


          • #6
            The cluster density for the four lanes range from 0-250k/mm2.

            The lanes:
            Lane 1: 235 ±41
            Lane 2: 237 ±38
            Lane 3: 212 ±56
            Lane 4: 230 ±35

            I hope this helps.

            Comment


            • #7
              This run is borderline overclustered (as I recollect for NextSeq optimal cluster density is 170-220 K cluster/mm^2).

              Since there is nothing to lose at this time contact tech support and see what they say. You may have to eat the cost on this one and re-run.

              Comment


              • #8
                Hi, How was this error fixed?

                Comment

                Latest Articles

                Collapse

                • seqadmin
                  Improved Targeted Sequencing: A Comprehensive Guide to Amplicon Sequencing
                  by seqadmin



                  Amplicon sequencing is a targeted approach that allows researchers to investigate specific regions of the genome. This technique is routinely used in applications such as variant identification, clinical research, and infectious disease surveillance. The amplicon sequencing process begins by designing primers that flank the regions of interest. The DNA sequences are then amplified through PCR (typically multiplex PCR) to produce amplicons complementary to the targets. RNA targets...
                  03-21-2023, 01:49 PM
                • seqadmin
                  Targeted Sequencing: Choosing Between Hybridization Capture and Amplicon Sequencing
                  by seqadmin




                  Targeted sequencing is an effective way to sequence and analyze specific genomic regions of interest. This method enables researchers to focus their efforts on their desired targets, as opposed to other methods like whole genome sequencing that involve the sequencing of total DNA. Utilizing targeted sequencing is an attractive option for many researchers because it is often faster, more cost-effective, and only generates applicable data. While there are many approaches...
                  03-10-2023, 05:31 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by seqadmin, 03-24-2023, 02:45 PM
                0 responses
                14 views
                0 likes
                Last Post seqadmin  
                Started by seqadmin, 03-22-2023, 12:26 PM
                0 responses
                15 views
                0 likes
                Last Post seqadmin  
                Started by seqadmin, 03-17-2023, 12:32 PM
                0 responses
                17 views
                0 likes
                Last Post seqadmin  
                Started by seqadmin, 03-15-2023, 12:42 PM
                0 responses
                24 views
                0 likes
                Last Post seqadmin  
                Working...
                X