new posts

  • bini
    replied
    Hi, How was this error fixed?

    Leave a comment:


  • GenoMax
    replied
    This run is borderline overclustered (as I recollect for NextSeq optimal cluster density is 170-220 K cluster/mm^2).

    Since there is nothing to lose at this time contact tech support and see what they say. You may have to eat the cost on this one and re-run.

    Leave a comment:


  • ox_zoo
    replied
    The cluster density for the four lanes range from 0-250k/mm2.

    The lanes:
    Lane 1: 235 ±41
    Lane 2: 237 ±38
    Lane 3: 212 ±56
    Lane 4: 230 ±35

    I hope this helps.

    Leave a comment:


  • GenoMax
    replied
    What was the cluster density for this run?

    If the run was borderline overclustered to begin with, it is possible that sequencer lost the ability to distinguish the clusters over time (they become fatter). There is also the possibility of this being some sort of a hardware issue.

    Leave a comment:


  • ox_zoo
    replied
    Thanks for your reply.

    Yes I will do that - unfortunately since its a weekend they cannot reply.

    I was thinking - if clusters are not formed in those two particular areas, would the cameras be disabled? My pooled library concentration was sufficient, so I did not anticipate any problems. Also, if the cluster density is low then I am not sure how QC would be affected..

    Many thanks
    Last edited by ox_zoo; 11-19-2016, 07:39 AM.

    Leave a comment:


  • GenoMax
    replied
    Probably. You should contact Illumina tech support. They can take a look at your sequencer remotely to diagnose the problem.

    If your instrument is not directly accessible over the network then you will need to send them some diagnostic files.

    Leave a comment:


  • ox_zoo
    replied
    Yes, now it has finished and the QC is only 15.2.

    The machine also has a warning message on two of the lanes:

    camera disabled:failed to detect clusters


    Does this mean that the cameras are not working properly?

    Thanks

    Leave a comment:


  • ox_zoo
    started a topic Low QC scores?

    Low QC scores?

    I have just set up my very first sequencing experiment on the Next-Seq 500. It is now 45 cycles into PE sequencing, however it seems that my QC scores are a bit low. I only have a % >= Q30%: 25.2%

    I have around:
    280 million reads in the QC 10 bin.
    25 million in the QC 20 bin
    8 million in the QC 25 bin
    65 million in QC 30 bin
    44 million in QC 35 bin.

    Is this a bad result? Apologies for my stupidness, just was wondering whether data could be salvaged from this?

Latest Articles

Collapse

  • seqadmin
    Essential Discoveries and Tools in Epitranscriptomics
    by seqadmin




    The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...
    04-22-2024, 07:01 AM
  • seqadmin
    Current Approaches to Protein Sequencing
    by seqadmin


    Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
    04-04-2024, 04:25 PM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by seqadmin, Yesterday, 08:47 AM
0 responses
16 views
0 likes
Last Post seqadmin  
Started by seqadmin, 04-11-2024, 12:08 PM
0 responses
60 views
0 likes
Last Post seqadmin  
Started by seqadmin, 04-10-2024, 10:19 PM
0 responses
60 views
0 likes
Last Post seqadmin  
Started by seqadmin, 04-10-2024, 09:21 AM
0 responses
54 views
0 likes
Last Post seqadmin  
Working...
X