  • bini
    replied
    Hi, How was this error fixed?

    Leave a comment:


  • GenoMax
    replied
    This run is borderline overclustered (as I recollect for NextSeq optimal cluster density is 170-220 K cluster/mm^2).

    Since there is nothing to lose at this time contact tech support and see what they say. You may have to eat the cost on this one and re-run.

    Leave a comment:


  • ox_zoo
    replied
    The cluster density for the four lanes range from 0-250k/mm2.

    The lanes:
    Lane 1: 235 ±41
    Lane 2: 237 ±38
    Lane 3: 212 ±56
    Lane 4: 230 ±35

    I hope this helps.

    Leave a comment:


  • GenoMax
    replied
    What was the cluster density for this run?

    If the run was borderline overclustered to begin with, it is possible that sequencer lost the ability to distinguish the clusters over time (they become fatter). There is also the possibility of this being some sort of a hardware issue.

    Leave a comment:


  • ox_zoo
    replied
    Thanks for your reply.

    Yes I will do that - unfortunately since its a weekend they cannot reply.

    I was thinking - if clusters are not formed in those two particular areas, would the cameras be disabled? My pooled library concentration was sufficient, so I did not anticipate any problems. Also, if the cluster density is low then I am not sure how QC would be affected..

    Many thanks
    Last edited by ox_zoo; 11-19-2016, 07:39 AM.

    Leave a comment:


  • GenoMax
    replied
    Probably. You should contact Illumina tech support. They can take a look at your sequencer remotely to diagnose the problem.

    If your instrument is not directly accessible over the network then you will need to send them some diagnostic files.

    Leave a comment:


  • ox_zoo
    replied
    Yes, now it has finished and the QC is only 15.2.

    The machine also has a warning message on two of the lanes:

    camera disabled:failed to detect clusters


    Does this mean that the cameras are not working properly?

    Thanks

    Leave a comment:


  • ox_zoo
    started a topic Low QC scores?

    Low QC scores?

    I have just set up my very first sequencing experiment on the Next-Seq 500. It is now 45 cycles into PE sequencing, however it seems that my QC scores are a bit low. I only have a % >= Q30%: 25.2%

    I have around:
    280 million reads in the QC 10 bin.
    25 million in the QC 20 bin
    8 million in the QC 25 bin
    65 million in QC 30 bin
    44 million in QC 35 bin.

    Is this a bad result? Apologies for my stupidness, just was wondering whether data could be salvaged from this?

Latest Articles

Collapse

  • SEQadmin2
    16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
    by SEQadmin2




    NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
    Today, 06:46 AM
  • SEQadmin2
    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
    by SEQadmin2



    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
    ...
    10-14-2026, 09:55 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, 10-28-2026, 09:57 AM
0 responses
24 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-28-2026, 09:51 AM
0 responses
15 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-22-2026, 11:13 AM
0 responses
71 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-22-2026, 10:48 AM
0 responses
48 views
0 reactions
Last Post SEQadmin2  
Working...