This run is borderline overclustered (as I recollect for NextSeq optimal cluster density is 170-220 K cluster/mm^2).
Since there is nothing to lose at this time contact tech support and see what they say. You may have to eat the cost on this one and re-run.
Leave a comment:
-
The cluster density for the four lanes range from 0-250k/mm2.
The lanes:
Lane 1: 235 ±41
Lane 2: 237 ±38
Lane 3: 212 ±56
Lane 4: 230 ±35
I hope this helps.
Leave a comment:
-
What was the cluster density for this run?
If the run was borderline overclustered to begin with, it is possible that sequencer lost the ability to distinguish the clusters over time (they become fatter). There is also the possibility of this being some sort of a hardware issue.
Leave a comment:
-
Thanks for your reply.
Yes I will do that - unfortunately since its a weekend they cannot reply.
I was thinking - if clusters are not formed in those two particular areas, would the cameras be disabled? My pooled library concentration was sufficient, so I did not anticipate any problems. Also, if the cluster density is low then I am not sure how QC would be affected..
Many thanksLast edited by ox_zoo; 11-19-2016, 07:39 AM.
Leave a comment:
-
Probably. You should contact Illumina tech support. They can take a look at your sequencer remotely to diagnose the problem.
If your instrument is not directly accessible over the network then you will need to send them some diagnostic files.
Leave a comment:
-
Yes, now it has finished and the QC is only 15.2.
The machine also has a warning message on two of the lanes:
camera disabled:failed to detect clusters
Does this mean that the cameras are not working properly?
Thanks
Leave a comment:
-
Low QC scores?
I have just set up my very first sequencing experiment on the Next-Seq 500. It is now 45 cycles into PE sequencing, however it seems that my QC scores are a bit low. I only have a % >= Q30%: 25.2%
I have around:
280 million reads in the QC 10 bin.
25 million in the QC 20 bin
8 million in the QC 25 bin
65 million in QC 30 bin
44 million in QC 35 bin.
Is this a bad result? Apologies for my stupidness, just was wondering whether data could be salvaged from this?
Latest Articles
Collapse
-
by SEQadmin2
NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...-
Channel: Articles
Today, 06:46 AM -
-
by SEQadmin2
Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?
Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating...-
Channel: Articles
10-14-2026, 09:55 AM -
ad_right_rmr
Collapse
News
Collapse
| Topics | Statistics | Last Post | ||
|---|---|---|---|---|
|
Genome Sequencing Reveals Large, Repeat-Rich DNA in Trapdoor Spider Named for Stephen Colbert
by SEQadmin2
Started by SEQadmin2, 10-28-2026, 09:57 AM
|
0 responses
24 views
0 reactions
|
Last Post
by SEQadmin2
10-28-2026, 09:57 AM
|
||
|
Started by SEQadmin2, 10-28-2026, 09:51 AM
|
0 responses
15 views
0 reactions
|
Last Post
by SEQadmin2
10-28-2026, 09:51 AM
|
||
|
Started by SEQadmin2, 10-22-2026, 11:13 AM
|
0 responses
71 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 11:13 AM
|
||
|
Started by SEQadmin2, 10-22-2026, 10:48 AM
|
0 responses
48 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 10:48 AM
|
Leave a comment: