    NovaSeq Experience

    Hi, I have seen a lot of the theoretical discussion around the NovaSeq and the resulting data. I was just wondering if there were any users that were actually running a NovaSeq that could comment on how it was performing in the lab in terms of loading, errors, general issues etc?
  • Brian Bushnell
    Super Moderator
    • Jan 2014
    • 2709

    #2
    I don't know anything about the lab issues, but the sequence quality is good. Coverage exhibits slightly more bias than HiSeq for the same libraries. Using unique dual barcodes and performing deduplication are both essential, because of the high crosstalk and duplicate rates.

    Comment

    • nucacidhunter
      Jafar Jabbari
      • Jan 2013
      • 1250

      #3
      It will be good instrument when Illumina comes up with real fix(es) for index hopping issue. Currently its use is limited to sequencing large whole genomes where available UDI combinations can be utilized and not many are willing to use it knowing that their discoveries or positive expected results could be due index hopping.

      Comment

      • kcchan
        Senior Member
        • Jul 2012
        • 186

        #4
        Honestly this index hopping issue is way overblown. For most sequencing applications it's a non-issue. From our experience a clean library (i.e. no visible primers or dimers in BA trace), the hopping rate is less than 1%. What's causing problems are those who got used to the fact that the MiSeqs and HiSeq 2000s were able to sequencing bad libraries without issues and then running into problems with pattern flow cells. Once you understand the nuances of the new instrument it's a great platform and much easier to use compared to the older instruments.

        Comment

        • Markiyan
          Senior Member
          • Sep 2010
          • 126

          #5
          We need to move index closer (start) of the sequencing read... - Like 454-MID.

          Originally posted by nucacidhunter View Post
          It will be good instrument when Illumina comes up with real fix(es) for index hopping issue. Currently its use is limited to sequencing large whole genomes where available UDI combinations can be utilized and not many are willing to use it knowing that their discoveries or positive expected results could be due index hopping.
          Since index hopping seems to be caused by recombinase, recombining over a homologus sequencing/index primer binding sequence, what about using 454-MID style indexing - devote first 6-8 basepairs of the sequencing read to the index?

          Than recombination events over flanking sequencing primers would not change much - the index would remain with its read.

          Also, it would be great if Illumina could make a 4-channell version of the NovaSeq system/kits, since a lot of researchers would sacrifice half of the throughput for 3-10 times lower error rate. - Esp. if they want to assemble de novo umapped reads.

          And one would be easilly able to run a 2 chanell sequencing kits on that system for NextSeq-style data.

          I hope that the quality of the sequencing reagents keeps up with the time (Unlike the NextSeq V2 story): http://seqanswers.com/forums/showpos...0&postcount=89 and does not fall under BGISeq500 or similar system.
          Last edited by Markiyan; 10-06-2017, 11:41 PM. Reason: Typo/clarification.

          Comment

          • nucacidhunter
            Jafar Jabbari
            • Jan 2013
            • 1250

            #6
            Originally posted by Markiyan View Post
            Since index hopping seems to be caused by recombinase, recombining over a homologus sequencing/index primer binding sequence, what about using 454-MID style indexing - devote first 6-8 basepairs of the sequencing read to the index?

            Than recombination events over flanking sequencing primers would not change much - the index would remain with its read.
            The issue with inline barcode/index in Illumina systems is base diversity. In some library preps such as GBS which is done in batches of multiple samples it is easy but for a single library they have to have at least four colour balanced barcodes. Also if the inline barcodes were added to adapters then both R1 and R2 will read the barcode wasting a bit of sequencing.

            Comment

            • pmiguel
              Senior Member
              • Aug 2008
              • 2328

              #7
              Originally posted by Brian Bushnell View Post
              I don't know anything about the lab issues, but the sequence quality is good. Coverage exhibits slightly more bias than HiSeq for the same libraries. Using unique dual barcodes and performing deduplication are both essential, because of the high crosstalk and duplicate rates.
              Brian,
              "HiSeq" doesn't denote a single kind of Illumina instrument. HiSeq 2000/2500 are quite different from HiSeq 3000/4000/X. Which type do you mean?

              --
              Phillip

              Comment

              • GenoMax
                Senior Member
                • Feb 2008
                • 7138

                #8
                I think @Brian is referring to 4- vs 2-color chemistry difference in that sentence.

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #9
                  Originally posted by nucacidhunter View Post
                  It will be good instrument when Illumina comes up with real fix(es) for index hopping issue. Currently its use is limited to sequencing large whole genomes where available UDI combinations can be utilized and not many are willing to use it knowing that their discoveries or positive expected results could be due index hopping.
                  Actually you can purchase full (96 or 384) UDI adapter sets from IDT. Although last I checked they were still a custom synthesis order. Which means you get more adapter than you are likely to use at a pretty high price. (Approaching $10K for the 96 UDI adapter set.)

                  Also, NuGen apparently has released a 96 UDI with some of their Illumina library construction kits.

                  --
                  Phillip

                  Comment

                  • pmiguel
                    Senior Member
                    • Aug 2008
                    • 2328

                    #10
                    Originally posted by GenoMax View Post
                    I think @Brian is referring to 4- vs 2-color chemistry difference in that sentence.
                    Seems unlikely. That aspect was only introduced to the thread after Brian's post.

                    Brian actually praises the accuracy of the NovaSeq--which is what is claimed later is an issue.

                    I don't think that in actuality the two-color nature of the NovaSeq makes it less accurate than the HiSeq 3000/4000/X. I think it performs similarly or better. At least all Illumina's testing seems to show that.

                    --
                    Phillip

                    Comment

                    • Brian Bushnell
                      Super Moderator
                      • Jan 2014
                      • 2709

                      #11
                      Originally posted by pmiguel View Post
                      Brian,
                      "HiSeq" doesn't denote a single kind of Illumina instrument. HiSeq 2000/2500 are quite different from HiSeq 3000/4000/X. Which type do you mean?

                      --
                      Phillip
                      Oh, sorry, I meant OUR HiSeq machines Those are 2000/2500/1T. In this specific case I was comparing it to a 2500 run.

                      To clarify, from isolate random fragment data downsampled to the same number of reads in each case, after mapping, I observed slightly more genomic bases with very low (0x, 1x, etc.) coverage on the NovaSeq, and its primary genomic peak was shifted to slightly higher counts. We have not done a robust analysis of the effects on assembly, though I can't imagine coverage bias would give a better assembly. I will try to make some time to quantify this effect; it's of great interest to JGI as well. But we've only done 3 runs and I've only looked at 2 of them. Statistically it's not significant if you consider runs as the number of samples, but it is quite important if you consider reads as the number of samples. I'm not yet sure which is the case here.
                      Last edited by Brian Bushnell; 10-09-2017, 06:20 PM.

                      Comment

                      • nucacidhunter
                        Jafar Jabbari
                        • Jan 2013
                        • 1250

                        #12
                        Originally posted by pmiguel View Post
                        Actually you can purchase full (96 or 384) UDI adapter sets from IDT. Although last I checked they were still a custom synthesis order. Which means you get more adapter than you are likely to use at a pretty high price. (Approaching $10K for the 96 UDI adapter set.)

                        Also, NuGen apparently has released a 96 UDI with some of their Illumina library construction kits.

                        --
                        Phillip
                        Thanks. IDT has not released their UDI adapters that they are co-developing with Illumina. NuGEN adapters are not sold as a stand alone product and I am not sure if their adapters are specific to their kits or if they can be used with other kits as well.

                        Comment

                        • Brian Bushnell
                          Super Moderator
                          • Jan 2014
                          • 2709

                          #13
                          It's interesting to me that Illumina introduced NovaSeq without accompanying adapter kits to enable a high degree of multiplexing. Their current 24-unique-index kit seems targeted at human exome-capture on NovaSeq. From Illumina's perspective, higher degrees of multiplexing are probably seen as hurting profitability.

                          Comment

                          • nucacidhunter
                            Jafar Jabbari
                            • Jan 2013
                            • 1250

                            #14
                            Originally posted by Brian Bushnell View Post
                            It's interesting to me that Illumina introduced NovaSeq without accompanying adapter kits to enable a high degree of multiplexing. Their current 24-unique-index kit seems targeted at human exome-capture on NovaSeq.
                            Illumina's current indexing strategy (up to 384 samples for Nextera) can be used on NovaSeq but using current TruSeq adapters only 8 unique combinations of index 1 and 2 is possible which is not enough for lots of applications.

                            Comment

                            • Markiyan
                              Senior Member
                              • Sep 2010
                              • 126

                              #15
                              Inline indexes & patterned flowcells.

                              Originally posted by nucacidhunter View Post
                              The issue with inline barcode/index in Illumina systems is base diversity. In some library preps such as GBS which is done in batches of multiple samples it is easy but for a single library they have to have at least four colour balanced barcodes. Also if the inline barcodes were added to adapters then both R1 and R2 will read the barcode wasting a bit of sequencing.
                              In that case we need to have first 4-5 bases as a random sequence (for clusters ID), followed by the barcode (another 4-8 bases), than the insert...

                              This is really important for RNAseq & CHIPseq applications, where we want to keep the cross-talk bellow 10^-6...
                              Classical dual unique indexing would get us into 10^-4 - 10^-5 range.

                              BTW: Actually the spatial signal separation from the patterned flowcell clusters should resemble more the 454 signal, and the clusters should be much more easier to call, even if low diversity at the first few bases.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                The Factors Reshaping Genome Assembly
                                by SEQadmin2



                                When the Human Genome Project declared the human genome finished in 2003, portions of the sequence were still missing, including many of its most repetitive and structurally complex regions.1 Still, for the next two decades, researchers analyzed their sequencing data against the single linear reference that came out of this effort. Released in 2013 as GRCh38, that reference has powered countless discoveries, but it has gaps and cannot capture the genetic diversity of the world’s...
                                11-24-2026, 07:04 AM
                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2





                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification,...
                                10-30-2026, 06:46 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 11-23-2026, 11:11 AM
                              0 responses
                              39 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 11-17-2026, 06:27 AM
                              0 responses
                              47 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 11-12-2026, 07:59 AM
                              0 responses
                              68 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 11-05-2026, 09:25 AM
                              0 responses
                              68 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...