Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #16
    No I did not know that those tests even exist...
    I strongly suggest not running those tests, if you are not familiar with them (and before talking with Illumina tech support).

    Comment

    • GenoMax
      Senior Member
      • Feb 2008
      • 7142

      #17
      Originally posted by SarahAurora View Post
      It's not our MiSeq, we just use it from our cooperation partner. I'll check if they have a maintenance contract with Illumina. Nevertheless, I wrote tech support an email with all details.
      Illumina tech support will entertain questions from end-users (so you have done the right thing) but you would need your partner to be involved if they need remote access to the sequencer to check this run/require the partner to run the diagnostic tests.

      Comment

      • GW_OK
        Senior Member
        • Sep 2009
        • 411

        #18
        Onboard diagnostics are in the "System Check" icon under "Manage Instrument" from the main screen on the Miseq Control Software. The optics checks are fairly fast to run, the fluidics will take some time and babysitting (raising and lowering sippers as prompted).

        Has your partner had any successful runs since your failed run?

        Edit: Asking me to run these tests is usually what Tech Support has me do first when I call them with an issue. Enough so that I'll sometimes run them prior to calling...

        Comment

        • SarahAurora
          Member
          • Jun 2017
          • 17

          #19
          Ok that's all very good advice, thanks a lot! I hope tech support can help me!

          Comment

          • GenoMax
            Senior Member
            • Feb 2008
            • 7142

            #20
            If this failure is due to hardware/reagent problem (and the partner has a maintenance contract with Illumina) then Illumina generally will replace the reagents at no cost, so you should be able to re-run your samples for free (or at least no charge for the reagents).

            Comment

            • SarahAurora
              Member
              • Jun 2017
              • 17

              #21
              Originally posted by GW_OK View Post
              Onboard diagnostics are in the "System Check" icon under "Manage Instrument" from the main screen on the Miseq Control Software. The optics checks are fairly fast to run, the fluidics will take some time and babysitting (raising and lowering sippers as prompted).

              Has your partner had any successful runs since your failed run?
              No, we are the only one using the MiSeq. We already used it for 8 runs, but the ninth one was now the fatal one....

              Comment

              • SarahAurora
                Member
                • Jun 2017
                • 17

                #22
                Originally posted by GenoMax View Post
                If this failure is due to hardware/reagent problem (and the partner has a maintenance contract with Illumina) then Illumina generally will replace the reagents at no cost, so you should be able to re-run your samples for free (or at least no charge for the reagents).
                That would be great!

                Comment

                • thermophile
                  Senior Member
                  • Apr 2015
                  • 243

                  #23
                  I run primarily amplicons too so the metrics that I look for are a bit different than the standards that Illumina uses (our runs don't meet the base diversity that makes some of the standard number make sense).

                  From the %q30 it looks like something bad happened part way through R1-amplicons aren't really long enough, optics issue, fluidics issue. Whatever it was *seems* to be machine/reagent related because it didn't improve after the turnaround. Are you spiking in primers or do your amplicons use Illumina sequencing primers? If you make an error spiking in the primers, you won't see anything after the turn around.

                  My list to check (roughly in order)

                  1. %phiX aligned. is it close to what I thought I was spiking in?
                  2. %base does it look appropriately horrible (my 16s amplicons should be at 90% for the first ~50 bases, then decline to ~80% for the rest of the read)
                  3. Cluster density across flowcell, top and bottom
                  4. thumbnails, does the density seem about right
                  5. fwmh. The software is not made for 90% single base, focus issues are common, but fwmh shouldn't be too jagged
                  Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

                  Comment

                  Latest Articles

                  Collapse

                  • SEQadmin2
                    16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                    by SEQadmin2




                    NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                    10-30-2026, 06:46 AM
                  • SEQadmin2
                    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                    by SEQadmin2



                    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                    ...
                    10-14-2026, 09:55 AM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by SEQadmin2, 10-28-2026, 09:57 AM
                  0 responses
                  43 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-28-2026, 09:51 AM
                  0 responses
                  26 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-22-2026, 11:13 AM
                  0 responses
                  74 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-22-2026, 10:48 AM
                  0 responses
                  48 views
                  0 reactions
                  Last Post SEQadmin2  
                  Working...