Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • cement_head
    Senior Member
    • Mar 2012
    • 265

    #1

    MiSeq | Updating the Windows7 Embedded System

    Hi,

    I'm having endless issues updating/patching the Windows7 Embedded system on the MiSeq. Constantly getting errors from Windows Update about it not being able to update, and of course, being M$, the error codes are next to useless.

    Is there anyway around this? I just don't like the idea of an un-patched system on the Internet running M$ Windows.

    Regards,
    Andor
  • kmcarr
    Senior Member
    • May 2008
    • 1181

    #2
    In my opinion it is unwise to install software patches which have not been tested for compatibility by Illumina. We do not patch any instrument computers with anything other than patches/upgrades provided by the instrument vendor.

    Your MiSeq and all networked laboratory computers should be protected by a firewall with very stringent access rules.

    Comment

    • GenoMax
      Senior Member
      • Feb 2008
      • 7142

      #3
      You should leave your MiSeq as is. Isolating instrument computers from open internet with a strong firewall (preferably via a defined/separate VLAN) is the best option for securing instrument computers.

      Comment

      • cement_head
        Senior Member
        • Mar 2012
        • 265

        #4
        Okay, that works in a "closed" facility. What happens if we have an "open" facility and any user can insert a USB drive? Do I just cross my fingers?

        I guess I'd like to know how to prevent usb spawned malware.

        Comment

        • misterc
          Member
          • Jan 2016
          • 46

          #5
          You could open it up and remove the internal USB cable that passes through to the motherboard for the USB port on the right hand side of the instrument. I forget how the USB port on the back of the MiSeq is plumbed, but that one might be part of the motherboard (and not a cable).

          Comment

          • GenoMax
            Senior Member
            • Feb 2008
            • 7142

            #6
            @misterc: Problem is a lot of times users (or even the facility techs) may create samplesheets using IEM elsewhere and then bring those to the machine at the time of run-start. People may also want to off-load data using USB external drives (if networking is not available).

            @cement_head: User education generally works well. You should put up signs (if users can walk in any hour of the day) about using good USB hygiene and absolute need to ensure a clean environment. If you can't trust certain users to follow this basic protocol then you would need to be a "bad cop" and make sure they don't access the sequencer unattended or first scan their USB's on a separate machine before they are allowed to plug those into sequencer.

            Comment

            • Mastersales1015
              Junior Member
              • Jan 2019
              • 3

              #7
              770-490-3265 we have one that is being retired. A illumina MiSeq DOM 2013.

              Comment

              • cement_head
                Senior Member
                • Mar 2012
                • 265

                #8
                Originally posted by GenoMax View Post
                @misterc: Problem is a lot of times users (or even the facility techs) may create samplesheets using IEM elsewhere and then bring those to the machine at the time of run-start. People may also want to off-load data using USB external drives (if networking is not available).

                @cement_head: User education generally works well. You should put up signs (if users can walk in any hour of the day) about using good USB hygiene and absolute need to ensure a clean environment. If you can't trust certain users to follow this basic protocol then you would need to be a "bad cop" and make sure they don't access the sequencer unattended or first scan their USB's on a separate machine before they are allowed to plug those into sequencer.
                Yep, you are correct - almost always USER ERROR is what we fight (constantly).

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                  by SEQadmin2




                  NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                  10-30-2026, 06:46 AM
                • SEQadmin2
                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                  by SEQadmin2



                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                  ...
                  10-14-2026, 09:55 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, 10-28-2026, 09:57 AM
                0 responses
                35 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-28-2026, 09:51 AM
                0 responses
                22 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-22-2026, 11:13 AM
                0 responses
                72 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-22-2026, 10:48 AM
                0 responses
                48 views
                0 reactions
                Last Post SEQadmin2  
                Working...